Hardware => Thermo => Topic started by: DavidAdams on April 12, 2019, 10:09:54 AM

Title: Pathfinder 200 eV cutoff
Post by: DavidAdams on April 12, 2019, 10:09:54 AM
Does anyone know of a way to have Thermo Pathfinder not use a default 200 eV cutoff value when triggered by Probe for EPMA? No matter what I do when I change the Low Energy Cutoff value in Pathfinder and save it once an analysis starts from PfE the 200 eV default cutoff is automatically selected.

Thanks,
DAve
Title: Re: Pathfinder 200 eV cutoff
Post by: SteveSeddio on April 12, 2019, 11:53:14 AM
Hi Dave,

Maybe  you have done these things, but here are two ideas:
1. Before beginning the acquisition with PfE, acquire an EDS spectrum from Pathfinder with the desired low energy cutoff. Then use PfE to acquire.
2. Create a Pathfinder project that has acquired spectra with the desired low energy cutoff. Then create a new project in that project and acquire data with PfE.

Let me know if either of these options help.
-Steve
Title: Re: Pathfinder 200 eV cutoff
Post by: DavidAdams on April 15, 2019, 05:15:42 AM
Thanks Steve!

I'll give those things a try