Title: Blank Values Outside Min/Max Analytical Total?
Post by: Scott B. on January 04, 2024, 03:30:41 PM
Greetings,

In the CalcImage "Specify Quantitative Parameters" dialogue for setting up a quant map calculation, there's this checkbox.

(https://smf.probesoftware.com/gallery/2871_04_01_24_3_26_47.jpeg)

Where does one input their preferred values for this feature?

Cheers,

Scott B.
Title: Re: Blank Values Outside Min/Max Analytical Total?
Post by: John Donovan on January 04, 2024, 03:50:45 PM
The analytical min/max totals are defined in the [software] section of the probewin.ini file as shown here:

lpAppName$ = "Software"
lpKeyName$ = "AnalyticalTotalMinimum"
lpDefault$ = "95"

lpAppName$ = "Software"
lpKeyName$ = "AnalyticalTotalMaximum"
lpDefault$ = "105"

We will also add a "tool tip" popup text to help in the future.
Title: Re: Blank Values Outside Min/Max Analytical Total?
Post by: John Donovan on January 05, 2024, 08:53:03 AM
In the [software] section of the Probewin.ini file (usually found in the C:\ProgramData\Probe Software\Probe for EPMA folder), the keywords are:

AnalyticalTotalMinimum="95"
AnalyticalTotalMaximum="105"

The AnalyticalTotalMinimum value must be greater than 50 and the AnalyticalTotalMaximum value must be less than 150.