I want to see if my TAP spectrometer positions are drifting over a full day. I suspect they may be drifting over a few hours. How can I check spectrometer drift?
Quote from: Joe Boesenberg on February 25, 2025, 01:18:46 PMI want to see if my TAP spectrometer positions are drifting over a full day. I suspect they may be drifting over a few hours. How can I check spectrometer drift?
Here are some links on spectrometer reproducibility:
https://smf.probesoftware.com/index.php?topic=437.msg2365#msg2365
https://smf.probesoftware.com/index.php?topic=1555.msg11988#msg11988
https://smf.probesoftware.com/index.php?topic=210.msg10691#msg10691
https://smf.probesoftware.com/index.php?topic=369.msg1948#msg1948
The following link discusses using the Remote Automation Excel scripts to perform extensive tests on your instrument. You can modify these scripts with any sort of delays you want:
https://smf.probesoftware.com/index.php?topic=1038.msg12771#msg12771
There's also the Drift utility once you have a lot of data to sift through:
https://smf.probesoftware.com/index.php?topic=1156.0
If you don;t need to test the reproducibility of the spectrometer movement, one can simply run the StartWin application, peak the 5 spectrometers on some elements, and then just run Startwin to acquire counts in re-cycle mode from the Modes menu.
(https://smf.probesoftware.com/gallery/395_25_02_25_2_45_11.png)
Then you select the data lines and from the Output menu, click this menu:
(https://smf.probesoftware.com/gallery/395_25_02_25_2_50_12.png)
Then select the columns you want to plot...
You can also set the "cycle time" between counting cycles from the Count Times button:
(https://smf.probesoftware.com/gallery/395_25_02_25_3_04_07.png)