Software => DTSA II => Topic started by: Amber Coles on July 29, 2025, 02:37:04 PM

Title: Morphology which detector when doing APA, EDS raster?
Post by: Amber Coles on July 29, 2025, 02:37:04 PM
I have two questions: I am working on quantifying uncertainty and resolution in APA using a traceable pitch standard (metro-chip for the moment). We rely almost exclusively on the BSE detector for APA except when the sample material is same or lighter than the substrate (as with C fibers). In order to be consistent, then, I need to measure the metrochip with the same detector as the morphology. I assume if we set APA to identify particles using BSE then the morphology measurements are also BSE?

Also, when collecting EDS for a particle during APA, is the beam in a fixed position or does it "raster" or move similar to the RCA?

Thanks!
Title: Re: Morphology which detector when doing APA, EDS raster?
Post by: Nicholas Ritchie on July 29, 2025, 04:41:56 PM
Amber,
    The search and measure detectors are the same.
    The default APA EDS beam position is point mode.  It can also be configured for chord raster.

Nicholas