Software => Probe for EPMA => Topic started by: pgopon on July 24, 2013, 02:41:34 PM

Title: Standard/ unknown different coating?
Post by: pgopon on July 24, 2013, 02:41:34 PM
Hi all,

I have been playing with the feature in PfEPMA where you can designate different coatings for standards/unknowns and was wondering if anyone knew how well or if this feature worked.

I have attached a excel file that shows the results of the PfEPMA matrix correction with and without the coating correction.  In this case my carbon standard was un-coated, the rest of the standards were C coated, with the exception of the Fe and Si standards were Pt coated at the same time as the unknown.  I would have thought add/removing this correction would have a larger effect on the wt %, does anyone have any insight as to why the correction is so small.  I have done some of the monte carlo modeling and we should be seeing more counts on the platinum coated sample.

Cheers,

phil

Edit by JJD to add Excel sheet as attachment
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on July 24, 2013, 04:46:01 PM
The coating correction in PFE is quite a crude correction.  One part is an absorption correction for low energy x-rays emitted through the coating and the second, is a correction for electron energy loss from primary electrons going through the coating.

It does not account for additional emitted x-rays from the coating.
Title: Using a Different Coating on standard and unknown
Post by: John Donovan on May 19, 2014, 04:35:09 PM
An improved coating correction has been implemented in Probe for EPMA (and CalcZAF) based on a modified coating correction by Kerrick, et. al., Amer. Min. 58, 920-925 (1973). You'll need v. 10.3.6 to utilize these new methods.

There are two parts to this coating correction:

1. A correction for "electron absorption" which compensates for the decrease in over voltage as the beam decelerates through the coating thickness.

' Intensity correction from electron energy loss due to coating
ratio! = (100# - ((8.3 * sample(1).CoatingDensity! * sample(1).CoatingThickness! / ANGPERNM&) / (sample(1).KilovoltsArray!(chan%) ^ 2 - (sample(1).LineEdge!(chan%) / EVPERKEV#) ^ 2))) / 100#

2. A correction for x-ray attenuation as the emitted x-ray transmits through the coating layer (coating thickness / Sin (theta)).

' Calculate thickness based on takeoff angle
radians! = sample(1).takeoff! * PI! / 180#
thickness! = sample(1).CoatingThickness! / Sin(radians!)

' Calculate x-ray transmission based on Sin thickness in angstroms
transmission! = NATURALE# ^ (-1# * atotal! * sample(1).CoatingDensity! * thickness! * CMPERANGSTROM#)


A test sample was prepared of an ultra pure Ti metal coated with pure Ag. The thickness was determined by XRR as 120 nm as shown in the following diffraction fringes:

(https://smf.probesoftware.com/oldpics/i59.tinypic.com/30wbdb4.jpg)

A separate determination was performed using StrataGEM using 6, 8 and 15 keV which gave 124.7 nm assuming a density of 10.5 for Ag and no carbon coat on the pure Ti or the Ag coated Ti.

(https://smf.probesoftware.com/oldpics/i57.tinypic.com/s2acuc.jpg)

(https://smf.probesoftware.com/oldpics/i57.tinypic.com/35jcmye.jpg)

Results from initial EPMA measurements produce the following k-ratios:

6 keV       0.0298
8 keV       0.0795
15 keV     0.463

Results from MC modeling and PFE coating corrections will be posted later this week and thanks to John Armstrong for his assistance.
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on May 20, 2014, 02:49:42 PM
The coating correction is specified for both standards and unknowns within each probe run database. This way, your coating specifications cannot be modified unless you want to. For example to specify the coating parameters for your standards separately or all at once using the Standard | Edit Standard Parameters (coating) menu as seen here:

(https://smf.probesoftware.com/oldpics/i62.tinypic.com/2eg7cdg.jpg)

For your unknowns the coating parameters can also be assign to each unknown or all selected as seen here:

(https://smf.probesoftware.com/oldpics/i60.tinypic.com/14e7tyo.jpg)

These flags are always defaulted to the coating parameters defined in the Probewin.ini file, but can be modified by editing these keywords:

[standards]
StandardCoatingFlag=1       ; 0 = not coated, 1 = coated
StandardCoatingElement=6     ; assume carbon
StandardCoatingDensity=2.1
StandardCoatingThickness=200   ; in angstroms

and the coating flags for electron absorption and x-ray transmission can be turned on from the Analytical | Analysis Options as seen here:

(https://smf.probesoftware.com/oldpics/i59.tinypic.com/vmsqi8.jpg)
Title: Re: Standard/ unknown different coating?
Post by: Probeman on May 27, 2014, 01:10:18 PM
I re-ran the Ag coated Ti sample at 8, 9, 10, 11, 12, 13, 14, 15, 20 and 25 keV on my SX50 and after processing with STRATAGem* I obtained 119.6 nm. This compares quite well to the XRR measurement of 119.7 nm!

I call this "spurious accuracy" (credit to Joe Michael for the concept)!  That is, too good to be true!   ;)

In fact the STRATAGem* keV vs. K-ratio plot seems to show a small but systematic bias in the data. The student that performed the deposition and XRR thinks this is due to the Ag density at the surface of the Ti not being exactly 10.5.  But for the purposes of a coating correction let us assume that the Ag film is 120 nm.

See attached documents below.

* © Copyright 1993-2016 SAMx
Title: Re: Standard/ unknown different coating?
Post by: Probeman on June 03, 2014, 12:23:02 PM
Here are some preliminary results comparing 6, 8 and 15 keV measurements of Ti Ka on a Ti sample coated with 120 nm of Ag (see attached spreadsheet).

(https://smf.probesoftware.com/oldpics/i60.tinypic.com/1glyz8.jpg)

As you can see the MC calculations with an assumed coating thickness of 120 nm are close, but the MC k-ratios seem to indicate a coating thickness of a little over 100 nm thickness for the Ag.

Be that as it may, we can now start to test the coating correction in Probe for EPMA... but remember, this coating correction was calibrated for 20 nm of carbon, not 120 nm of Ag!  So we can assume that it will fail, but how bad will it fail and can this coating correction be improved further?
Title: Re: Standard/ unknown different coating?
Post by: Probeman on June 03, 2014, 04:58:56 PM
Let's start with what should be the easiest coating correction for this sample, Ti Ka underneath 120 nm of Ag, at 25 keV. Without the coating correction, we obtain the following:

Un   68 coated Ti (MVA)
TakeOff = 40.0  KiloVolt = 25.0  Beam Current = 30.0  Beam Size =    0
(Magnification (analytical) =  20000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =    400)
Image Shift (X,Y):                                          .00,   .00
Pre Acquire String :                                        mag 100000
Post Acquire String :                                          mag 400
Number of Data Lines:   5             Number of 'Good' Data Lines:   5
First/Last Date-Time: 05/22/2014 03:19:44 AM to 05/22/2014 03:32:22 AM

Average Total Oxygen:         .000     Average Total Weight%:   68.318
Average Calculated Oxygen:    .000     Average Atomic Number:   22.000
Average Excess Oxygen:        .000     Average Atomic Weight:   47.900
Average ZAF Iteration:        1.00     Average Quant Iterate:     2.00

Un   68 coated Ti (MVA), Results in Elemental Weight Percents

ELEM:       Ti
BGDS:      LIN
TIME:    80.00
BEAM:    30.14

ELEM:       Ti   SUM 
   305  68.423  68.423
   306  68.206  68.206
   307  68.639  68.639
   308  68.538  68.538
   309  67.784  67.784

AVER:   68.318  68.318
SDEV:     .339    .339
SERR:     .152
%RSD:      .50
STDS:     2522


With a 68% total (relative to pure Ti) we can see that 120 nm of Ag significantly affects the production and emission of Ti Ka x-rays. Now we'll turn the correction on as discussed here above:

http://smf.probesoftware.com/index.php?topic=23.msg1258#msg1258

and now we obtain the following results corrected for the 120 nm of Ag on Ti:

Un   68 coated Ti (MVA)
TakeOff = 40.0  KiloVolt = 25.0  Beam Current = 30.0  Beam Size =    0
(Magnification (analytical) =  20000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =    400)
Image Shift (X,Y):                                          .00,   .00
Pre Acquire String :                                        mag 100000
Post Acquire String :                                          mag 400
Number of Data Lines:   5             Number of 'Good' Data Lines:   5
First/Last Date-Time: 05/22/2014 03:19:44 AM to 05/22/2014 03:32:22 AM

Average Total Oxygen:         .000     Average Total Weight%:   99.101
Average Calculated Oxygen:    .000     Average Atomic Number:   22.000
Average Excess Oxygen:        .000     Average Atomic Weight:   47.900
Average ZAF Iteration:        1.00     Average Quant Iterate:     2.00

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=Ag, Density=10.5 gm/cm3, Thickness=1200 angstroms, Sin(Thickness)=1866.87 angstroms

Un   68 coated Ti (MVA), Results in Elemental Weight Percents

ELEM:       Ti
BGDS:      LIN
TIME:    80.00
BEAM:    30.14

ELEM:       Ti   SUM 
   305  99.253  99.253
   306  98.938  98.938
   307  99.567  99.567
   308  99.420  99.420
   309  98.326  98.326

AVER:   99.101  99.101
SDEV:     .492    .492
SERR:     .220
%RSD:      .50
STDS:     2522


Not bad at all, in fact it looks like "spurious accuracy" too good to be true, but let's look at the other lower voltages here in this table, which we expect to be more significant coating corrections:

(https://smf.probesoftware.com/oldpics/i59.tinypic.com/htut7a.jpg)

Clearly we got lucky on the 25 keV correction, but the lower voltages increasingly produce an over correction in the Ti concentrations, at least until the electron absorption correction goes negative!

Clearly this 20 nm carbon coating correction can be improved. Suggestions...?
Title: Re: Standard/ unknown different coating?
Post by: Gseward on June 26, 2014, 11:26:44 AM
John,

I am a little confused about when the modifications to the coating parameters are applied. I open the Standards coating parameters dialog from the Standard menu, modify my standard coatings (some coated with 20nm C, some uncoated), close the dialog and then recalculate the analysis, but the modified coating parameters don't seem to have been applied? What am I doing wrong?

Having modified the coatings, if I select all samples (unkn and Sts) in Analyze! open the calculations options window, then close it again, the modifications are applied, but this seems like a round about way of updating the changes.

Please enlighten me!

gareth
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on June 26, 2014, 11:42:52 AM
Happy to do so!

Basically PFE loads defaults for the standard and unknown coating parameters based on the values in the Probewin.ini file.

Then those options are disabled at a global level (you'll see why in a minute), so they are loaded but not utilized by default in the matrix corrections.

To turn on the coating parameters explicitly you need to go to the Analytical | Analysis Option menu dialog and check these options:

(https://smf.probesoftware.com/oldpics/i57.tinypic.com/2mxpctl.jpg)

Because the user will generally only be modifying their unknown coating parameters, this global flag is turned on automatically if the user edits the coating parameters in the Calculation Options dialog for unknown samples. This is explained in the post here above:

http://smf.probesoftware.com/index.php?topic=23.msg1258#msg1258

This action is similar to how specifying a spectral interference correction in the Standard Assignments dialog also turns on the global interference correction flag in Analysis Options.

Why do it like this?  So every correction the software performs can be toggled on/off without having to re-specify all the sample level parameters!

Question for you:  does your version have the new purple interactive Help button visible?  That is in the latest version and takes you right here in the forum!
Title: Re: Standard/ unknown different coating?
Post by: Gseward on June 26, 2014, 08:37:59 PM
Thanks for the reply, John, but this doesn't really answer my question. When I modify the standard coatings, the new parameters are not applied, unless I go through the route I described previously. 

I am running the current version, purple buttons and all!

I don't have to modify standard coatings very often, so I'll not worry about it any further.

Gareth
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on June 26, 2014, 08:46:05 PM
Quote from: Gseward on June 26, 2014, 08:37:59 PM
Thanks for the reply, John, but this doesn't really answer my question. When I modify the standard coatings, the new parameters are not applied, unless I go through the route I described previously. 

Right, exactly. That is a feature (as opposed a bug)- because one doesn't often modify the standard coatings!

In other words- I want the user to specifically want the coating correction applied when simply browsing the standard coating parameters.
john
Title: Re: Standard/ unknown different coating?
Post by: JohnF on September 03, 2017, 09:47:18 AM
JohnD:

Can you please clarify: if the two boxes for Coating Corrections (electron energy,  X-ray absorption) are NOT checked in Analysis Calculations Options, while meanwhile there are default settings for Carbon coating on both std and unknowns, is NO correction done for the carbon coating? And ONLY when the two boxes are checked, and only then, is there a correction (albeit small) made?

thanks.

JohnF
Title: Re: Standard/ unknown different coating?
Post by: Probeman on September 03, 2017, 11:23:35 AM
Quote from: JohnF on September 03, 2017, 09:47:18 AM
JohnD:

Can you please clarify: if the two boxes for Coating Corrections (electron energy,  X-ray absorption) are NOT checked in Analysis Calculations Options, while meanwhile there are default settings for Carbon coating on both std and unknowns, is NO correction done for the carbon coating? And ONLY when the two boxes are checked, and only then, is there a correction (albeit small) made?

thanks.

JohnF

JohnF,
That is correct.

The coating element, thickness, on/off flags in the Analyze! | Calculation Options dialog are specified on a sample basis.  The Analytical | Analysis Options menu dialog are a way to turn all these sample coating calculations on/off on a global basis.
JohnD
Title: Re: Standard/ unknown different coating?
Post by: Probeman on September 04, 2017, 11:12:58 AM
I should add that you will notice that the coating options in the Analyze! | Calculation Options dialog are disabled for standard samples. 

This is because coating parameters for standards are specified from the Standard | Edit Standard Coating Parameters menu dialog as seen here:

(https://smf.probesoftware.com/gallery/395_04_09_17_11_09_36.png)

john
Title: Re: Standard/ unknown different coating?
Post by: UofO EPMA Lab on August 13, 2019, 04:45:57 PM
Probeman here.

While at M&M last week I was asked a question about analyzing standards in Probe for EPMA. That is one acquires data on a standard and then from the Analyze! window one clicks the Analyze button to get a quantitative analysis of that standard *as though it were an unknown*.

Of course if the standard is the primary (assigned) standard for that element, the element in question will iterate to the correct concentration, because the standard is analyzing itself as the standard. That is why the %VAR (percent variance) value for that element is displayed in parentheses- because it isn't really a test of the analytical quality. 

More interesting is when one analyzes a standard (as an unknown from the Analyze! window), and that standard contains a non-zero concentration of an element (or a zero concentration to check for spectral interferences!), and that standard is *not* the primary standard for the element in question. Then we have the so-called secondary standard situation which is the best way to check for accuracy in EPMA.

And of course because the program knows that this sample is a standard, it can also print out the published analysis of the standard from the standard database for comparison.

But it is important to remember that for the quantification of unknowns, the only elements utilized in the primary standard are the elements that are actually assigned as the primary standard for that element.  All the other elements in that standard are only useful for accuracy evaluation of major elements, or zero concentrations for interference corrections, or of course as standards for the MAN background correction.

So, when you have the situation where the standards are coated with, for example carbon, and the unknowns are not coated, say for carbon analysis of a steel, you will be analyzing the coating. So, the standard for iron might be pure Fe, and one would get an analysis (from the Analyze! window) like this:

St  526 Set   3 Iron metal

St  526 Set   3 Iron metal
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 50.0  Beam Size =    0
(Magnification (analytical) =  40000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =   1572)
Image Shift (X,Y):                                         .00,    .00

From Johnson-Matthey, Vacuum remelted, Batch BM1664
Optical emission: Al < 1ppm, Ca < 1 ppm,
Cr 2 ppm, Co 20 ppm, Cu 3 ppm, Ni 3 ppm
Si 60 ppm, Sn 10 ppm, Ag < 1 ppm
Oxygen 310 ppm, Nitrogen 10 ppm
Number of Data Lines:   3             Number of 'Good' Data Lines:   3
First/Last Date-Time: 08/13/2019 01:58:46 PM to 08/13/2019 02:07:32 PM
WARNING- Using Exponential Off-Peak correction for n ka
WARNING- Using Exponential Off-Peak correction for c ka
WARNING- Using Exponential Off-Peak correction for si ka
WARNING- Using Exponential Off-Peak correction for cu ka
WARNING- Using Exponential Off-Peak correction for o ka

Average Total Oxygen:         .000     Average Total Weight%:  113.447
Average Calculated Oxygen:    .000     Average Atomic Number:   24.060
Average Excess Oxygen:        .000     Average Atomic Weight:   41.316
Average ZAF Iteration:        5.00     Average Quant Iterate:     2.00

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=C, Density=2.1 gm/cm3, Thickness=200 angstroms, Sin(Thickness)=311.145 angstroms

St  526 Set   3 Iron metal, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    40.00   40.00   40.00   40.00   20.00   20.00   20.00   40.00   20.00   20.00   20.00
BEAM:    49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   156   -.160  10.904   -.002    .011   -.006    .002   -.016    .357 102.982    .011    .008 114.091
   157   -.085  10.732   -.020    .007   -.019   -.013   -.003    .372 101.793   -.003   -.007 112.755
   158   -.285  10.792   -.025    .006    .011   -.002    .033    .432 102.560   -.019   -.009 113.495

AVER:    -.177  10.809   -.016    .008   -.004   -.004    .005    .387 102.445   -.003   -.003 113.447
SDEV:     .101    .087    .012    .002    .015    .007    .025    .040    .603    .015    .009    .669
SERR:     .058    .050    .007    .001    .008    .004    .015    .023    .348    .009    .005
%RSD:   -57.10     .81  -78.83   30.67 -328.34 -169.36  544.36   10.29     .59 -433.32 -346.16

PUBL:     n.a.    n.a.    n.a.    n.a.    n.a.    n.a.    n.a.    n.a. 100.000    n.a.    n.a. 100.000
%VAR:      ---     ---     ---     ---     ---     ---     ---     ---  (2.45)     ---     ---
DIFF:      ---     ---     ---     ---     ---     ---     ---     ---  (2.45)     ---     ---
STDS:      604     506     542     514     528     523     529     913     526     524     525

STKF:    .1637   .9635   .9910  1.0000  1.0000  1.0000   .9974   .2509  1.0000   .9988  1.0000
STCT:    28.99  853.49  258.45 1148.84  609.87  416.77  562.01  195.43  210.20  154.13  188.74

UNKF:   -.0006   .0285  -.0001   .0001   .0000  -.0001   .0000   .0021  1.0000   .0000   .0000
UNCT:     -.11   25.22    -.03     .06    -.03    -.02     .02    1.64  210.20    -.01     .00
UNBG:     1.31     .99     .39     .22    2.37     .64    2.90    2.40     .68     .28     .39

ZCOR:   2.8933  3.7968  1.1943  1.4322  1.0712   .8660  1.1135  1.8413  1.0245   .7501  1.0443
KRAW:   -.0037   .0295  -.0001   .0001   .0000  -.0001   .0000   .0084  1.0000   .0000   .0000
PKBG:      .92   26.40     .91    1.29     .99     .97    1.01    1.68  310.00     .98     .99
INT%:     ----    ----    ----    ----    ----    ----    ----    ----    ----    ----    -.04

Note that the total is quite high because the standard is carbon coated, and we are analyzing for carbon. And note also that the Fe concentration is wrong because we are adding 10 percent or so of carbon from the carbon coating.

But this extra carbon seen when a carbon coated standard is analyzed as an unknown, has zero effect on the standardization of unknowns because only the Fe intensities (in this standard) are being utilized for the unknown analysis. And of course we have specified in Probe for EPMA that the standards are carbon coated and the unknowns are not.

When we do this, we can analyze a uncoated sample using a standard that is carbon coated and the differences in the coating are compensated for in the matrix correction:

Un    6 H13 trav
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 50.0  Beam Size =    0
(Magnification (analytical) =  40000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =   1572)
Image Shift (X,Y):                                         .00,    .00
Number of Data Lines:   5             Number of 'Good' Data Lines:   1
First/Last Date-Time: 08/13/2019 04:01:21 PM to 08/13/2019 04:27:59 PM
WARNING- Using Exponential Off-Peak correction for n ka
WARNING- Using Exponential Off-Peak correction for c ka
WARNING- Using Exponential Off-Peak correction for si ka
WARNING- Using Exponential Off-Peak correction for cu ka
WARNING- Using Exponential Off-Peak correction for o ka
WARNING- Using Time Dependent Intensity (TDI) Element Correction

Average Total Oxygen:         .000     Average Total Weight%:  100.766
Average Calculated Oxygen:    .000     Average Atomic Number:   24.805
Average Excess Oxygen:        .000     Average Atomic Weight:   46.435
Average ZAF Iteration:        4.00     Average Quant Iterate:     3.00

No Sample Coating and/or No Sample Coating Correction

Un    6 H13 trav, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    60.00   60.00   60.00   60.00   36.00   40.00   36.00   90.00   20.00   40.00   40.00
BEAM:    48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   165   3.406   2.406   1.224    .872    .194    .953    .092    .284  86.449   4.550    .334 100.766

AVER:    3.406   2.406   1.224    .872    .194    .953    .092    .284  86.449   4.550    .334 100.766
SDEV:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
SERR:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
%RSD:      .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00
STDS:      604     506     542     514     528     523     529     913     526     524     525

STKF:    .1637   .9635   .9910  1.0000  1.0000  1.0000   .9974   .2509  1.0000   .9988  1.0000
STCT:    29.07  853.49  258.45 1148.84  609.87  416.77  562.01  195.43  210.20  154.13  188.74

UNKF:    .0142   .0061   .0103   .0061   .0018   .0107   .0008   .0015   .8480   .0573   .0032
UNCT:     3.02    5.58    2.72    7.15    1.14    4.53     .48    1.28  181.90    9.01     .62
UNBG:     1.25     .74     .32     .28    2.39     .68    2.82    2.54     .64     .29     .39

ZCOR:   2.4014  3.9407  1.1922  1.4274  1.0606   .8939  1.1019  1.8968  1.0194   .7941  1.0359
KRAW:    .1040   .0065   .0105   .0062   .0019   .0109   .0009   .0066   .8654   .0585   .0033
PKBG:     3.41    8.50    9.52   26.38    1.48    7.63    1.17    1.51  283.94   32.20    2.60
INT%:     ----    ----    ----    ----    ----    ----    ----    ----     .00    ----   -1.01

TDI%:     .000   3.192   -.735    .000    .851    .000    .000  -4.725   2.630    .000    .000
DEV%:       .0     4.6     6.6      .0     4.9      .0      .0     3.8      .4      .0      .0
TDIF:     ---- LOG-LIN LOG-LIN    ---- LOG-LIN    ----    ---- LOG-LIN LOG-LIN    ----    ----
TDIT:      .00  111.00  112.00     .00  103.00     .00     .00  151.00   74.00     .00     .00
TDII:     ----    6.32    3.03    ----    3.55    ----    ----    3.69    183.    ----    ----
TDIL:     ----    1.84    1.11    ----    1.27    ----    ----    1.31    5.21    ----    ----
Title: Re: Standard/ unknown different coating?
Post by: UofO EPMA Lab on August 14, 2019, 12:46:14 PM
Quote from: UofO EPMA Lab on August 13, 2019, 04:45:57 PM
While at M&M last week I was asked a question about analyzing standards in Probe for EPMA. That is one acquires data on a standard and then from the Analyze! window one clicks the Analyze button to get a quantitative analysis of that standard *as though it were an unknown*.

Of course if the standard is the primary (assigned) standard for that element, the element in question will iterate to the correct concentration, because the standard is analyzing itself as the standard. That is why the %VAR (percent variance) value for that element is displayed in parentheses- because it isn't really a test of the analytical quality. 

Of course it should be pointed out that even though Probe for EPMA only utilizes the intensities of the elements actually assigned as the primary standard for that element, if one is analyzing a standard sample as an unknown in the Analyze! window, the matrix corrections will not be able to exactly converge correctly on the (ideal or published) concentration of the assigned element for this primary standard, if there is something significantly wrong with the other (unassigned) elements in that standard sample.

So in the case of our Fe (pure metal) standard shown in the previous post, because we have analyzed carbon on a carbon coated standard, of course we detect a significant amount (~10 wt%) of carbon as being in the Fe metal standard. Now the matrix iteration correctly calculates the std k-factor for our pure metal standard (the one we are analyzing as an unknown!) as being 1.0, but when it calculates the matrix correction physics for our Fe standard as containing ~10 wt% carbon, of course it cannot converge to the 100% Fe as one might expect:

St  526 Set   3 Iron metal, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    40.00   40.00   40.00   40.00   20.00   20.00   20.00   40.00   20.00   20.00   20.00
BEAM:    49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   156   -.160  10.904   -.002    .011   -.006    .002   -.016    .357 102.982    .011    .008 114.091
   157   -.085  10.732   -.020    .007   -.019   -.013   -.003    .372 101.793   -.003   -.007 112.755
   158   -.285  10.792   -.025    .006    .011   -.002    .033    .432 102.560   -.019   -.009 113.495

AVER:    -.177  10.809   -.016    .008   -.004   -.004    .005    .387 102.445   -.003   -.003 113.447
SDEV:     .101    .087    .012    .002    .015    .007    .025    .040    .603    .015    .009    .669
SERR:     .058    .050    .007    .001    .008    .004    .015    .023    .348    .009    .005
%RSD:   -57.10     .81  -78.83   30.67 -328.34 -169.36  544.36   10.29     .59 -433.32 -346.16

PUBL:     n.a.    n.a.    n.a.    n.a.    n.a.    n.a.    n.a.    n.a. 100.000    n.a.    n.a. 100.000
%VAR:      ---     ---     ---     ---     ---     ---     ---     ---  (2.45)     ---     ---
DIFF:      ---     ---     ---     ---     ---     ---     ---     ---  (2.45)     ---     ---
STDS:      604     506     542     514     528     523     529     913     526     524     525


And again, this doesn't matter for any other sample that uses this Fe standard, because only the measured Fe intensities (from the standard sample), and the (ideal) composition from the standard database is being utilized for the std k-factor calculation, but if one really wants to see the composition of any carbon coated standards analyzed as though they were actual unknowns, one can simply select all the standards (except of course the carbon standard!), then click the Elements/Cations button and then click the carbon row and check the "Disable Quant" checkbox and click OK and OK. Then we will get output like this when analyzing our carbon coated standards as unknowns:

St  526 Set   3 Iron metal, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    40.00     ---   40.00   40.00   20.00   20.00   20.00   40.00   20.00   20.00   20.00
BEAM:    49.56     ---   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56   49.56

ELEM:        N     C-D      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   156   -.124     ---   -.002    .011   -.006    .002   -.016    .299 100.558    .011    .008 100.741
   157   -.066     ---   -.020    .007   -.018   -.012   -.003    .311  99.405   -.003   -.007  99.595
   158   -.221     ---   -.025    .007    .011   -.002    .032    .362 100.161   -.018   -.009 100.297

AVER:    -.137     ---   -.016    .008   -.004   -.004    .005    .324 100.041   -.003   -.003 100.211
SDEV:     .078     ---    .012    .002    .014    .007    .025    .033    .586    .014    .009    .578
SERR:     .045     ---    .007    .001    .008    .004    .014    .019    .338    .008    .005
%RSD:   -57.14     ---  -78.83   30.68 -328.37 -169.35  544.25   10.33     .59 -433.25 -346.11

PUBL:     n.a.    n.a.    n.a.    n.a.    n.a.    n.a.    n.a.    n.a. 100.000    n.a.    n.a. 100.000
%VAR:      ---     ---     ---     ---     ---     ---     ---     ---   (.04)     ---     ---
DIFF:      ---     ---     ---     ---     ---     ---     ---     ---   (.04)     ---     ---
STDS:      604     ---     542     514     528     523     529     913     526     524     525
Title: Re: Standard/ unknown different coating?
Post by: Philipp Poeml on September 18, 2019, 02:46:39 AM
Quote from: UofO EPMA Lab on August 13, 2019, 04:45:57 PM

When we do this, we can analyze a uncoated sample using a standard that is carbon coated and the differences in the coating are compensated for in the matrix correction:

Un    6 H13 trav
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 50.0  Beam Size =    0
(Magnification (analytical) =  40000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =   1572)
Image Shift (X,Y):                                         .00,    .00
Number of Data Lines:   5             Number of 'Good' Data Lines:   1
First/Last Date-Time: 08/13/2019 04:01:21 PM to 08/13/2019 04:27:59 PM
WARNING- Using Exponential Off-Peak correction for n ka
WARNING- Using Exponential Off-Peak correction for c ka
WARNING- Using Exponential Off-Peak correction for si ka
WARNING- Using Exponential Off-Peak correction for cu ka
WARNING- Using Exponential Off-Peak correction for o ka
WARNING- Using Time Dependent Intensity (TDI) Element Correction

Average Total Oxygen:         .000     Average Total Weight%:  100.766
Average Calculated Oxygen:    .000     Average Atomic Number:   24.805
Average Excess Oxygen:        .000     Average Atomic Weight:   46.435
Average ZAF Iteration:        4.00     Average Quant Iterate:     3.00

No Sample Coating and/or No Sample Coating Correction

Un    6 H13 trav, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    60.00   60.00   60.00   60.00   36.00   40.00   36.00   90.00   20.00   40.00   40.00
BEAM:    48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   165   3.406   2.406   1.224    .872    .194    .953    .092    .284  86.449   4.550    .334 100.766

AVER:    3.406   2.406   1.224    .872    .194    .953    .092    .284  86.449   4.550    .334 100.766
SDEV:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
SERR:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
%RSD:      .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00
STDS:      604     506     542     514     528     523     529     913     526     524     525

STKF:    .1637   .9635   .9910  1.0000  1.0000  1.0000   .9974   .2509  1.0000   .9988  1.0000
STCT:    29.07  853.49  258.45 1148.84  609.87  416.77  562.01  195.43  210.20  154.13  188.74

UNKF:    .0142   .0061   .0103   .0061   .0018   .0107   .0008   .0015   .8480   .0573   .0032
UNCT:     3.02    5.58    2.72    7.15    1.14    4.53     .48    1.28  181.90    9.01     .62
UNBG:     1.25     .74     .32     .28    2.39     .68    2.82    2.54     .64     .29     .39

ZCOR:   2.4014  3.9407  1.1922  1.4274  1.0606   .8939  1.1019  1.8968  1.0194   .7941  1.0359
KRAW:    .1040   .0065   .0105   .0062   .0019   .0109   .0009   .0066   .8654   .0585   .0033
PKBG:     3.41    8.50    9.52   26.38    1.48    7.63    1.17    1.51  283.94   32.20    2.60
INT%:     ----    ----    ----    ----    ----    ----    ----    ----     .00    ----   -1.01

TDI%:     .000   3.192   -.735    .000    .851    .000    .000  -4.725   2.630    .000    .000
DEV%:       .0     4.6     6.6      .0     4.9      .0      .0     3.8      .4      .0      .0
TDIF:     ---- LOG-LIN LOG-LIN    ---- LOG-LIN    ----    ---- LOG-LIN LOG-LIN    ----    ----
TDIT:      .00  111.00  112.00     .00  103.00     .00     .00  151.00   74.00     .00     .00
TDII:     ----    6.32    3.03    ----    3.55    ----    ----    3.69    183.    ----    ----
TDIL:     ----    1.84    1.11    ----    1.27    ----    ----    1.31    5.21    ----    ----


And what happens if you switch on the C coating correction here? Does it change something significantly?
Title: Re: Standard/ unknown different coating?
Post by: Probeman on September 18, 2019, 09:02:16 AM
Hi Philipp,
Well of course if the sample was actually carbon coated, the measured carbon concentration would be much higher since we would be measuring that additional carbon, but if you're asking if the physics calculations would be different, the answer is yes.

The presence of a coating does two things. It first reduces the effective landing energy of the incident electrons, which mostly affects the calculation for low overvoltage elements such as Fe. And second it decreases the transmission of x-rays out of the sample, mostly low energy emission lines such as oxygen for example. But this latter effect can be offset by the increase in ionization efficiency due to less overvoltage for these low energy edges.

The only way to know is to do the physics calculation. The coating correction calculation in PFE is rather simple, and includes a calculation for electron energy loss from the coating material and thickness, and a calculation for x-ray absorption for the emitted x-rays.  The code is on GitHub.

For the previous example I had to relocate the specific file and found that I had re-standardized since the post you quoted, so the uncoated calculation numbers are slightly different. So I will show both the uncoated and coated calculations here, first the calculation assuming the standards are coated and the sample is uncoated (the actual situation):

Un    6 H13 trav
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 50.0  Beam Size =    0
(Magnification (analytical) =  40000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =   1572)
Image Shift (X,Y):                                         .00,    .00
Number of Data Lines:   5             Number of 'Good' Data Lines:   1
First/Last Date-Time: 08/13/2019 04:01:21 PM to 08/13/2019 04:27:59 PM
WARNING- Using Exponential Off-Peak correction for n ka
WARNING- Using Exponential Off-Peak correction for c ka
WARNING- Using Exponential Off-Peak correction for si ka
WARNING- Using Exponential Off-Peak correction for cu ka
WARNING- Using Exponential Off-Peak correction for o ka
WARNING- Using Time Dependent Intensity (TDI) Element Correction

Average Total Oxygen:         .000     Average Total Weight%:  100.818
Average Calculated Oxygen:    .000     Average Atomic Number:   24.804
Average Excess Oxygen:        .000     Average Atomic Weight:   46.430
Average ZAF Iteration:        4.00     Average Quant Iterate:     3.00

No Sample Coating and/or No Sample Coating Correction

Un    6 H13 trav, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    60.00   60.00   60.00   60.00   36.00   40.00   36.00   90.00   20.00   40.00   40.00
BEAM:    48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   165   3.412   2.401   1.244    .901    .197    .965    .093    .284  86.432   4.555    .335 100.818

AVER:    3.412   2.401   1.244    .901    .197    .965    .093    .284  86.432   4.555    .335 100.818
SDEV:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
SERR:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
%RSD:      .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00
STDS:      604     506     542     514     528     523     529     913     526     524     525

STKF:    .1637   .9635   .9910  1.0000  1.0000  1.0000   .9974   .2509  1.0000   .9988  1.0000
STCT:    29.03  855.51  254.38 1112.32  602.82  412.03  558.75  195.85  210.25  154.02  188.04

UNKF:    .0142   .0061   .0104   .0063   .0019   .0108   .0008   .0015   .8478   .0573   .0032
UNCT:     3.02    5.58    2.72    7.15    1.14    4.53     .48    1.28  181.90    9.01     .62
UNBG:     1.25     .74     .32     .28    2.39     .68    2.82    2.54     .64     .29     .39

ZCOR:   2.4026  3.9431  1.1922  1.4271  1.0606   .8942  1.1019  1.8988  1.0195   .7944  1.0360
KRAW:    .1041   .0065   .0107   .0064   .0019   .0110   .0009   .0066   .8652   .0585   .0033
PKBG:     3.41    8.50    9.52   26.38    1.48    7.63    1.17    1.51  283.94   32.20    2.60
INT%:     ----    ----    ----    ----    ----    ----    ----    ----     .00    ----   -1.01

TDI%:     .000   3.192   -.735    .000    .851    .000    .000  -4.725   2.630    .000    .000
DEV%:       .0     4.6     6.6      .0     4.9      .0      .0     3.8      .4      .0      .0
TDIF:     ---- LOG-LIN LOG-LIN    ---- LOG-LIN    ----    ---- LOG-LIN LOG-LIN    ----    ----
TDIT:      .00  111.00  112.00     .00  103.00     .00     .00  151.00   74.00     .00     .00
TDII:     ----    6.32    3.03    ----    3.55    ----    ----    3.69    183.    ----    ----
TDIL:     ----    1.84    1.11    ----    1.27    ----    ----    1.31    5.21    ----    ----

And here the same calculation assuming that both the standards and samples are coated (and as stated above, the sample was actually not coated, so this is a physically unreal situation):

Un    6 H13 trav
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 50.0  Beam Size =    0
(Magnification (analytical) =  40000),        Beam Mode = Analog  Spot
(Magnification (default) =      400, Magnification (imaging) =   1572)
Image Shift (X,Y):                                         .00,    .00
Number of Data Lines:   5             Number of 'Good' Data Lines:   1
First/Last Date-Time: 08/13/2019 04:01:21 PM to 08/13/2019 04:27:59 PM
WARNING- Using Exponential Off-Peak correction for n ka
WARNING- Using Exponential Off-Peak correction for c ka
WARNING- Using Exponential Off-Peak correction for si ka
WARNING- Using Exponential Off-Peak correction for cu ka
WARNING- Using Exponential Off-Peak correction for o ka
WARNING- Using Time Dependent Intensity (TDI) Element Correction

Average Total Oxygen:         .000     Average Total Weight%:  103.635
Average Calculated Oxygen:    .000     Average Atomic Number:   24.695
Average Excess Oxygen:        .000     Average Atomic Weight:   45.780
Average ZAF Iteration:        4.00     Average Quant Iterate:     3.00

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=C, Density=2.1 gm/cm3, Thickness=200 angstroms, Sin(Thickness)=311.145 angstroms

Un    6 H13 trav, Results in Elemental Weight Percents

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn
BGDS:      EXP     EXP     LIN     EXP     LIN     LIN     EXP     EXP     LIN     LIN     LIN
TIME:    60.00   60.00   60.00   60.00   36.00   40.00   36.00   90.00   20.00   40.00   40.00
BEAM:    48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49   48.49

ELEM:        N       C      Mo      Si      Ni       V      Cu       O      Fe      Cr      Mn   SUM 
   165   4.082   2.468   1.265    .917    .201    .984    .095    .317  88.315   4.648    .342 103.635

AVER:    4.082   2.468   1.265    .917    .201    .984    .095    .317  88.315   4.648    .342 103.635
SDEV:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
SERR:     .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
%RSD:      .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00
STDS:      604     506     542     514     528     523     529     913     526     524     525

STKF:    .1637   .9635   .9910  1.0000  1.0000  1.0000   .9974   .2509  1.0000   .9988  1.0000
STCT:    29.03  855.51  254.38 1112.32  602.82  412.03  558.75  195.85  210.25  154.02  188.04

UNKF:    .0170   .0063   .0106   .0064   .0019   .0110   .0009   .0016   .8652   .0584   .0033
UNCT:     3.02    5.58    2.72    7.15    1.14    4.53     .48    1.28  181.90    9.01     .62
UNBG:     1.25     .74     .32     .28    2.39     .68    2.82    2.54     .64     .29     .39

ZCOR:   2.3957  3.9291  1.1921  1.4258  1.0617   .8953  1.1032  1.9250  1.0208   .7955  1.0373
KRAW:    .1041   .0065   .0107   .0064   .0019   .0110   .0009   .0066   .8652   .0585   .0033
PKBG:     3.41    8.50    9.52   26.38    1.48    7.63    1.17    1.51  283.94   32.20    2.60
INT%:     ----    ----    ----    ----    ----    ----    ----    ----     .00    ----   -1.03

TDI%:     .000   3.192   -.735    .000    .851    .000    .000  -4.725   2.630    .000    .000
DEV%:       .0     4.6     6.6      .0     4.9      .0      .0     3.8      .4      .0      .0
TDIF:     ---- LOG-LIN LOG-LIN    ---- LOG-LIN    ----    ---- LOG-LIN LOG-LIN    ----    ----
TDIT:      .00  111.00  112.00     .00  103.00     .00     .00  151.00   74.00     .00     .00
TDII:     ----    6.32    3.03    ----    3.55    ----    ----    3.69    183.    ----    ----
TDIL:     ----    1.84    1.11    ----    1.27    ----    ----    1.31    5.21    ----    ----


As we can see, the assumption of a coated sample, when it was not actually coated, provides a poor result.
Title: Re: Standard/ unknown different coating?
Post by: Philipp Poeml on September 20, 2019, 02:27:16 PM
Hi John,

Thanks for replying! I think what I meant was:
You Write
Quote from: Probeman on September 18, 2019, 09:02:16 AM

No Sample Coating and/or No Sample Coating Correction


So, does it mean all coating correction is off?
Why not switch it on and define the standards as carbon coated and the unknown not?
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on September 20, 2019, 02:55:02 PM
Quote from: Philipp Poeml on September 20, 2019, 02:27:16 PM
Hi John,

Thanks for replying! I think what I meant was:
You Write
Quote from: Probeman on September 18, 2019, 09:02:16 AM

No Sample Coating and/or No Sample Coating Correction


So, does it mean all coating correction is off?
Why not switch it on and define the standards as carbon coated and the unknown not?

No, it just means that the (global) coating corrections were turned on in the Analytical | Analysis Options dialog, but that no sample coating was specified for this particular sample.

I think if you read this topic from the beginning it explains all this. For example, see here:

https://smf.probesoftware.com/index.php?topic=23.msg1402#msg1402
Title: Re: Standard/ unknown different coating?
Post by: Philipp Poeml on September 20, 2019, 03:07:54 PM
I see. Then I got it wrong. So coating correction was on for standards, but the sample was specified as uncoated. That makes total sense. Sorry for the confusion.
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on September 20, 2019, 03:17:28 PM
Quote from: Philipp Poeml on September 20, 2019, 03:07:54 PM
I see. Then I got it wrong. So coating correction was on for standards, but the sample was specified as uncoated. That makes total sense. Sorry for the confusion.

No worries, it's good to ask questions! 

Yes.  The global coating corrections (for standards and unknowns) in the Analytical | Analysis Options dialog were both turned on (energy loss and x-ray absorption from the coating), the standard sample coatings were specified in the Standard | Edit Standard Coating Parameters menu (the standard coatings are usually specified by default in the Probewin.ini file, but they are ignored if the Analysis Options parameters are not checked), and the unknown sample did not have a coating specified in the Analyze! Calculation Options dialog.

The reason the coating parameters are specified in different places for standards and unknowns is that the standards coating parameters are global, and the unknown coating parameters are specified on a sample by sample basis.
Title: Re: Standard/ unknown different coating?
Post by: Probeman on January 21, 2020, 04:46:01 PM
We recently were running some magnetite analyses for a customer and after standardizing we analyzed a few unknown grains and found surprisingly low totals, even after we specified calculating excess oxygen by applying the method of Droop (1987) as described in this post:

https://smf.probesoftware.com/index.php?topic=92.msg8593#msg8593

Un   41 Chill-9 Mgt-Ilm 1_MGT Replace, Results in Elemental Weight Percents

ELEM:       Fe      Mg      Ti       V      Mn      Cr      Al      Si      Ca       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    CALC
BGDS:      MAN     MAN     MAN     MAN     MAN     MAN     MAN     MAN     MAN
TIME:    50.00  120.00   80.00   60.00   60.00   60.00   60.00   60.00   40.00     ---
BEAM:    50.35   50.35   50.35   50.35   50.35   50.35   50.35   50.35   50.35     ---

ELEM:       Fe      Mg      Ti       V      Mn      Cr      Al      Si      Ca       O   SUM 
   469  53.826   1.046   9.557    .275    .423   -.002   1.061    .038    .013  26.413  92.650
   470  53.689   1.039   9.748    .265    .419    .001   1.057    .039    .012  26.434  92.702
   471  53.788   1.046   9.764    .274    .419    .003   1.067    .043    .021  26.500  92.925
   472  53.684   1.040   9.588    .260    .423   -.007   1.064    .051    .044  26.390  92.535
   473  53.646   1.042   9.752    .264    .427   -.003   1.079    .044    .057  26.462  92.770
   474  53.809   1.048   9.729    .267    .432   -.008   1.047    .036    .055  26.500  92.913
   475  53.318   1.058   9.764    .272    .424   -.013   1.027    .092    .066  26.375  92.385
   476  53.640   1.056  10.168    .270    .433   -.005    .989    .026    .037  26.600  93.214
   477  53.530   1.076  10.094    .270    .428   -.005    .987    .029    .035  26.541  92.986
   478  53.220   1.084  10.430    .262    .437    .003    .962    .036    .045  26.583  93.061
   479  52.999   1.083  10.291    .243    .438    .000    .952    .041    .024  26.421  92.492

AVER:   53.559   1.056   9.899    .266    .428   -.003   1.026    .043    .037  26.474  92.785
SDEV:     .268    .017    .295    .009    .007    .005    .046    .018    .018    .077    .259
SERR:     .081    .005    .089    .003    .002    .001    .014    .005    .005    .023
%RSD:      .50    1.61    2.98    3.43    1.58 -154.27    4.47   40.66   48.92     .29
STDS:      395      12      22      23      25      24     306      14     306     ---

So of course we first tried to analyzed some standards as unknowns and they came out just fine. Well it turned out that this user had carbon coated their thin sections themselves and because they weren't quite sure of the procedure they apparently laid on a *very* thick coat of carbon. After we removed the thin section from the probe it did indeed appear to be quite dark, so we walked over to the stylus profilometer and measured the carbon coat, after scratching some areas clean with a razor blade. And lo and behold, indeed we had close to 70 nm of carbon on these thin sections!   :o

So we went back to the probe and turned on the coating correction in the Analysis Options menu by checking the two boxes as described here:

https://smf.probesoftware.com/index.php?topic=23.msg1402#msg1402

and then we specified this extra thick carbon coat in the Calculation Options as shown here:

(https://smf.probesoftware.com/gallery/395_21_01_20_4_41_32.png)

and now we analyze the sample again and now it's much better:

Un   41 Chill-9 Mgt-Ilm 1_MGT Replace, Results in Elemental Weight Percents

ELEM:       Fe      Mg      Ti       V      Mn      Cr      Al      Si      Ca       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    CALC
BGDS:      MAN     MAN     MAN     MAN     MAN     MAN     MAN     MAN     MAN
TIME:    50.00  120.00   80.00   60.00   60.00   60.00   60.00   60.00   40.00     ---
BEAM:    50.35   50.35   50.35   50.35   50.35   50.35   50.35   50.35   50.35     ---

ELEM:       Fe      Mg      Ti       V      Mn      Cr      Al      Si      Ca       O   SUM 
   469  57.884   1.137  10.181    .287    .453   -.003   1.142    .040    .014  28.369  99.504
   470  57.736   1.129  10.384    .276    .449    .001   1.137    .042    .013  28.391  99.556
   471  57.842   1.137  10.401    .286    .449    .003   1.148    .046    .022  28.462  99.796
   472  57.731   1.130  10.214    .270    .453   -.008   1.144    .054    .047  28.343  99.379
   473  57.690   1.133  10.389    .275    .458   -.004   1.161    .047    .061  28.420  99.629
   474  57.865   1.139  10.364    .278    .463   -.009   1.126    .039    .058  28.461  99.784
   475  57.338   1.150  10.402    .283    .455   -.014   1.105    .098    .070  28.327  99.214
   476  57.683   1.148  10.832    .281    .464   -.005   1.064    .028    .039  28.567 100.101
   477  57.565   1.169  10.753    .281    .459   -.006   1.062    .031    .037  28.504  99.857
   478  57.232   1.178  11.111    .273    .468    .003   1.035    .038    .048  28.547  99.933
   479  56.994   1.177  10.963    .252    .470    .000   1.024    .044    .026  28.374  99.324

AVER:   57.596   1.148  10.545    .277    .458   -.004   1.104    .046    .040  28.433  99.643
SDEV:     .289    .018    .314    .010    .007    .005    .049    .019    .019    .082    .277
SERR:     .087    .006    .095    .003    .002    .002    .015    .006    .006    .025
%RSD:      .50    1.61    2.98    3.54    1.58 -143.60    4.47   40.64   48.93     .29
STDS:      395      12      22      23      25      24     306      14     306     ---

Note that the reason for Fe being so affected by this extra thick carbon coat is not due to x-ray absorption. The major effect is from electron energy loss from this extra thick carbon coat and hence a significantly lower landing energy for the incident electrons. This lowers the overvoltage (at 15 keV) significantly enough to cause the previously observed low totals.

Just for fun, here is the excess oxygen calculation along with the oxide and cation output:

Ferrous/Ferric Calculation Results:
        Ferric/TotalFe   FeO        Fe2O3    Oxygen from Fe2O3 
   469        .350      48.405      28.964       2.902
   470        .344      48.734      28.388       2.844
   471        .343      48.877      28.381       2.843
   472        .349      48.386      28.767       2.882
   473        .343      48.764      28.288       2.834
   474        .347      48.585      28.737       2.879
   475        .342      48.516      28.060       2.811
   476        .338      49.150      27.850       2.790
   477        .340      48.911      27.946       2.800
   478        .330      49.348      26.984       2.704
   479        .333      48.892      27.151       2.720

AVER:         .342      48.779      28.138       2.819

Un   41 Chill-9 Mgt-Ilm 1_MGT Replace, Results in Oxide Weight Percents

ELEM:      FeO     MgO    TiO2    V2O3     MnO   Cr2O3   Al2O3    SiO2     CaO       O   SUM 
   469  74.467   1.885  16.982    .423    .585   -.004   2.157    .087    .020   2.902  99.504
   470  74.278   1.872  17.322    .406    .579    .001   2.148    .089    .018   2.844  99.556
   471  74.414   1.886  17.350    .420    .580    .004   2.169    .099    .031   2.843  99.796
   472  74.271   1.874  17.038    .398    .585   -.012   2.162    .116    .065   2.882  99.379
   473  74.218   1.878  17.329    .405    .591   -.005   2.193    .101    .085   2.834  99.629
   474  74.443   1.889  17.288    .409    .597   -.013   2.127    .083    .082   2.879  99.784
   475  73.765   1.907  17.351    .417    .587   -.020   2.087    .210    .099   2.811  99.214
   476  74.210   1.903  18.068    .414    .600   -.008   2.009    .060    .055   2.790 100.101
   477  74.057   1.939  17.937    .414    .592   -.008   2.007    .067    .052   2.800  99.857
   478  73.629   1.953  18.534    .401    .604    .005   1.955    .081    .067   2.704  99.933
   479  73.322   1.951  18.288    .370    .607    .000   1.934    .095    .036   2.720  99.324

AVER:   74.097   1.903  17.590    .407    .592   -.005   2.086    .099    .055   2.819  99.643
SDEV:     .371    .031    .524    .014    .009    .008    .093    .040    .027    .063    .277
SERR:     .112    .009    .158    .004    .003    .002    .028    .012    .008    .019
%RSD:      .50    1.61    2.98    3.54    1.58 -143.60    4.47   40.64   48.93    2.25
STDS:      395      12      22      23      25      24     306      14     306     ---

Un   41 Chill-9 Mgt-Ilm 1_MGT Replace, Results Based on Sum of 3 Cations

ELEM:       Fe      Mg      Ti       V      Mn      Cr      Al      Si      Ca       O   SUM 
   469   2.297    .104    .471    .013    .018    .000    .094    .003    .001   3.929   6.929
   470   2.290    .103    .480    .012    .018    .000    .093    .003    .001   3.930   6.930
   471   2.288    .103    .480    .012    .018    .000    .094    .004    .001   3.929   6.929
   472   2.293    .103    .473    .012    .018    .000    .094    .004    .003   3.930   6.930
   473   2.285    .103    .480    .012    .018    .000    .095    .004    .003   3.929   6.929
   474   2.290    .104    .478    .012    .019    .000    .092    .003    .003   3.931   6.931
   475   2.280    .105    .482    .012    .018   -.001    .091    .008    .004   3.932   6.932
   476   2.276    .104    .498    .012    .019    .000    .087    .002    .002   3.934   6.934
   477   2.276    .106    .496    .012    .018    .000    .087    .002    .002   3.934   6.934
   478   2.261    .107    .512    .012    .019    .000    .085    .003    .003   3.936   6.936
   479   2.265    .107    .508    .011    .019    .000    .084    .003    .001   3.937   6.937

AVER:    2.282    .104    .487    .012    .018    .000    .091    .004    .002   3.932   6.932
SDEV:     .011    .002    .014    .000    .000    .000    .004    .001    .001    .003    .003
SERR:     .003    .001    .004    .000    .000    .000    .001    .000    .000    .001
%RSD:      .50    1.59    2.88    3.47    1.53 -143.63    4.54   40.88   48.99     .07
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 17, 2023, 10:00:03 PM
Greetings!
I try to learn use the coating thickness option and made some tests.
I used 2 standards of Au in different blocks with different thickness of carbon coating:
Standard 1 - main standard, standard 2 - standard as unknown.
My results (% wt Au):

1. options do not "ON" (Analytical/Analysis options) (wt%): Standard 1 - 100.011 / Standard 2 - 101.757

2. options "ON" (Analytical/Analysis options), (wt%):           Standard 1 (100 A) - 102.205 / Standard 2 (200 A) - 103.989
                                                                                       Standard 1 (200 A) - 100.011 / Standard 2 (200 A) - 101.757
                                                                                       Standard 1 (300 A) - 97.817  /  Standard 2 (200 A) - 99.525

3. Also options "ON" (Analytical/Analysis options), (wt%):    Standard 1 (200 A) - 100.011 / Standard 2 (100 A) - 101.757
                                                                                       Standard 1 (200 A) - 100.011 / Standard 2 (200 A) - 101.757
                                                                                       Standard 1 (200 A) - 100.011 / Standard 2 (300 A) - 101.757

So, it means (please, correct me if I wrong):
-if options "OFF" (Analytical/Analysis options), the system uses the correction with default (200 A) coating thickness of main standard.
-the system do not use the correction for standards which are not using as main standards. Because these standards are not "unknowns" and and these standards do not standardize.

Is it correct?

Thank you.

Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 18, 2023, 08:47:31 AM
Please read carefully this whole topic.

It is difficult to understand what you are doing without more information. For example, in addition to turning on the coating corrections for electron energy loss and x-ray absorption in the Analytical | Analysis Options dialog, did you specify the standard coating in the Standard menu and also the unknown coating for the standard as an unknown in the Analyze! | Calculation Options dialog?
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 18, 2023, 04:59:45 PM
Quote from: John Donovan on January 18, 2023, 08:47:31 AM
Please read carefully this whole topic.

Of course, many times ))

Quote from: John Donovan on January 18, 2023, 08:47:31 AM
It is difficult to understand what you are doing without more information. For example, in addition to turning on the coating corrections for electron energy loss and x-ray absorption in the Analytical | Analysis Options dialog, did you specify the standard coating in the Standard menu

Of course, I choose in "Standard/Edit Standard coating parameters" "use coating" checkbox and thickness 100, 200 or 300 A which I noted in my question.
I can see that system uses "coating correction" with correct Thickness in main window of PFE with all other calculation results.

Quote from: John Donovan on January 18, 2023, 08:47:31 AM
... and also the unknown coating for the standard as an unknown in the Analyze! | Calculation Options dialog?

It is impossible because I work only with standards:
Standard 1 - is a main standard
Standard 2 - is also a standard but it is not using for standardize and its composition calculated based on Standard 1
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 18, 2023, 05:26:37 PM
Quote from: Rom on January 18, 2023, 04:59:45 PM
Quote from: John Donovan on January 18, 2023, 08:47:31 AM
... and also the unknown coating for the standard as an unknown in the Analyze! | Calculation Options dialog?

It is impossible because I work only with standards:
Standard 1 - is a main standard
Standard 2 - is also a standard but it is not using for standardize and its composition calculated based on Standard 1

OK, I think that is the problem.

Try running the unassigned standard (with the different coating thickness) as an unknown, and specify the coating on that material using the Calculation Options dialog.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 18, 2023, 07:21:11 PM
Yes, of course it works good. I did it many times.
But in the case like I asked I see
1. System uses "coating correction" (coating thickness by default, 200A) even if my checkboxes in Analytical/Analysis options are OFF.
2. System does not use "coating correction" with coating thickness different of default for unassigned standard.
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 18, 2023, 07:28:47 PM
Quote from: Rom on January 18, 2023, 07:21:11 PM
Yes, of course it works good. I did it many times.
But in the case like I asked I see
1. System uses "coating correction" (coating thickness by default, 200A) even if my checkboxes in Analytical/Analysis options are OFF.

Yes. The system applies the coating correction based on the flags in the Analytical | Analysis Options flags.

Quote from: Rom on January 18, 2023, 07:21:11 PM
2. System does not use "coating correction" with coating thickness different of default for unassigned standard.

The standards would need to have different standard numbers if you do not acquire them as unknowns.  Try acquiring them as unknowns and let me know how that works.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 19, 2023, 02:59:10 AM
1. Unfortunately even if  flags OFF, the system uses the correction. I showed upper the same result with and without flags in Analytical/Analysis Options.
2. Of course my standards 1 & 2 have different Standard numbers because they are different ))).

If I'll use standard 2 like Unknown - everything will good.
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 19, 2023, 08:39:04 AM
Quote from: Rom on January 19, 2023, 02:59:10 AM
1. Unfortunately even if  flags OFF, the system uses the correction. I showed upper the same result with and without flags in Analytical/Analysis Options.
2. Of course my standards 1 & 2 have different Standard numbers because they are different ))).

If I'll use standard 2 like Unknown - everything will good.

Is that last sentence a question?

I checked the coating corrections just now and they appear to be working fine.  For example, look at the Ca concentration in this feldspar standard without a coating correction (using anorthite as a primary Ca standard):

St  305 Set   1 Labradorite (Lake Co.), Results in Elemental Weight Percents

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC    SPEC
BGDS:      EXP     LIN     EXP     LIN     LIN
TIME:    10.00   10.00   10.00   10.00   10.00     ---     ---     ---
BEAM:    30.02   30.02   30.02   30.02   30.02     ---     ---     ---

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O   SUM 
   131  16.524    .526  24.125   2.988   9.695    .100    .084  46.823 100.866
   132  16.566    .468  23.898   3.044   9.698    .100    .084  46.823 100.680
   133  16.566    .586  23.851   2.923   9.719    .100    .084  46.823 100.653
   134  16.459    .516  24.216   3.022   9.552    .100    .084  46.823 100.773
   135  16.587    .518  24.262   2.929   9.541    .100    .084  46.823 100.843

AVER:   16.540    .523  24.070   2.981   9.641    .100    .084  46.823 100.763
SDEV:     .051    .042    .186    .054    .087    .000    .000    .000    .095
SERR:     .023    .019    .083    .024    .039    .000    .000    .000
%RSD:      .31    8.05     .77    1.82     .90     .00     .00     .00

PUBL:   16.359    .319  23.957   2.841   9.577    .100    .084  46.823 100.060
%VAR:     1.11   63.89     .47    4.94     .67     .00     .00     .00
DIFF:     .181    .204    .113    .140    .064    .000    .000    .000
STDS:       13      26      14     301     306     ---     ---     ---

I now set the carbon coating thickness for the anorthite standard from 200 angstroms to 400 angstroms (making sure to click the Apply Parameters to Selected Standard button from the Standard | Edit Standard Coating Parameters menu:

St  305 Set   1 Labradorite (Lake Co.), Results in Elemental Weight Percents

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC    SPEC
BGDS:      EXP     LIN     EXP     LIN     LIN
TIME:    10.00   10.00   10.00   10.00   10.00     ---     ---     ---
BEAM:    30.02   30.02   30.02   30.02   30.02     ---     ---     ---

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O   SUM 
   131  16.524    .526  24.125   2.988   9.695    .100    .084  46.823 100.866
   132  16.566    .468  23.898   3.044   9.698    .100    .084  46.823 100.680
   133  16.566    .586  23.851   2.923   9.719    .100    .084  46.823 100.653
   134  16.459    .516  24.216   3.022   9.552    .100    .084  46.823 100.773
   135  16.587    .518  24.262   2.929   9.541    .100    .084  46.823 100.843

AVER:   16.540    .523  24.070   2.981   9.641    .100    .084  46.823 100.763
SDEV:     .051    .042    .186    .054    .087    .000    .000    .000    .095
SERR:     .023    .019    .083    .024    .039    .000    .000    .000
%RSD:      .31    8.05     .77    1.82     .90     .00     .00     .00

PUBL:   16.359    .319  23.957   2.841   9.577    .100    .084  46.823 100.060
%VAR:     1.11   63.89     .47    4.94     .67     .00     .00     .00
DIFF:     .181    .204    .113    .140    .064    .000    .000    .000
STDS:       13      26      14     301     306     ---     ---     ---

No change, right?  Because I didn't yet turn on the *two* coating parameter flags from the Analytical | Analysis Options dialog, as I do now (that is, Use Conductive Coating Correction for Electron Absorption and Use Conductive Coating Correction for X-Ray Transmission):

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=C, Density=1.35 gm/cm3, Thickness=200 angstroms, Sin(Thickness)=311.145 angstroms

St  305 Set   1 Labradorite (Lake Co.), Results in Elemental Weight Percents

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC    SPEC
BGDS:      EXP     LIN     EXP     LIN     LIN
TIME:    10.00   10.00   10.00   10.00   10.00     ---     ---     ---
BEAM:    30.02   30.02   30.02   30.02   30.02     ---     ---     ---

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O   SUM 
   131  16.523    .526  24.127   2.987   9.589    .100    .084  46.823 100.760
   132  16.564    .468  23.900   3.043   9.592    .100    .084  46.823 100.574
   133  16.564    .586  23.854   2.923   9.613    .100    .084  46.823 100.546
   134  16.458    .516  24.219   3.022   9.447    .100    .084  46.823 100.669
   135  16.585    .518  24.264   2.928   9.436    .100    .084  46.823 100.739

AVER:   16.539    .523  24.073   2.980   9.536    .100    .084  46.823 100.658
SDEV:     .051    .042    .186    .054    .086    .000    .000    .000    .095
SERR:     .023    .019    .083    .024    .039    .000    .000    .000
%RSD:      .31    8.05     .77    1.82     .90     .00     .00     .00

PUBL:   16.359    .319  23.957   2.841   9.577    .100    .084  46.823 100.060
%VAR:     1.10   63.89     .48    4.91    -.43     .00     .00     .00
DIFF:     .180    .204    .116    .139   -.041    .000    .000    .000
STDS:       13      26      14     301     306     ---     ---     ---

Note the two statements just before the analysis results showing the coating parameters for this secondary standard are being utilized and now the results are different for Ca. Now I turn off those those conductive coating corrections in the Analytical | Analysis Options dialog and the results are the same as originally:

St  305 Set   1 Labradorite (Lake Co.), Results in Elemental Weight Percents

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC    SPEC
BGDS:      EXP     LIN     EXP     LIN     LIN
TIME:    10.00   10.00   10.00   10.00   10.00     ---     ---     ---
BEAM:    30.02   30.02   30.02   30.02   30.02     ---     ---     ---

ELEM:       Al      Fe      Si      Na      Ca       K      Mg       O   SUM 
   131  16.524    .526  24.125   2.988   9.695    .100    .084  46.823 100.866
   132  16.566    .468  23.898   3.044   9.698    .100    .084  46.823 100.680
   133  16.566    .586  23.851   2.923   9.719    .100    .084  46.823 100.653
   134  16.459    .516  24.216   3.022   9.552    .100    .084  46.823 100.773
   135  16.587    .518  24.262   2.929   9.541    .100    .084  46.823 100.843

AVER:   16.540    .523  24.070   2.981   9.641    .100    .084  46.823 100.763
SDEV:     .051    .042    .186    .054    .087    .000    .000    .000    .095
SERR:     .023    .019    .083    .024    .039    .000    .000    .000
%RSD:      .31    8.05     .77    1.82     .90     .00     .00     .00

PUBL:   16.359    .319  23.957   2.841   9.577    .100    .084  46.823 100.060
%VAR:     1.11   63.89     .47    4.94     .67     .00     .00     .00
DIFF:     .181    .204    .113    .140    .064    .000    .000    .000
STDS:       13      26      14     301     306     ---     ---     ---

I don't know what you are doing wrong, but again without more information it's difficult to tell.  I suggest re-reading this topic more carefully.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 19, 2023, 04:20:06 PM
Ok. Lets look at my 2 questions separately. So, firstly only question #1.

If you take your "start" results and turn on the *two* coating parameter flags from the Analytical | Analysis Options dialog WITHOUT changing carbon coating thickness for the anorthite standard from 200 angstroms to 400 angstroms in Standard menu
nothing will change - right?
It means coating parameters by default (200A) used in any case: OFF or ON *two* coating parameter flags from the Analytical | Analysis Options dialog - right?
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 19, 2023, 05:20:31 PM
Quote from: Rom on January 19, 2023, 04:20:06 PM
Ok. Lets look at my 2 questions separately. So, firstly only question #1.

If you take your "start" results and turn on the *two* coating parameter flags from the Analytical | Analysis Options dialog WITHOUT changing carbon coating thickness for the anorthite standard from 200 angstroms to 400 angstroms in Standard menu
nothing will change - right?
It means coating parameters by default (200A) used in any case: OFF or ON *two* coating parameter flags from the Analytical | Analysis Options dialog - right?

If the coating material and thickness on the two samples are the same, then yes, the results will be (essentially) the same whether you assume a coating correction or not.

If the coating material and thickness are not the same, then again yes, you must declare the coating material and thickness for each sample and turn on the coating correction flags.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 19, 2023, 05:59:08 PM
Yes, you are right.
If we turn on the *two* coating parameter flags from the Analytical | Analysis Options dialog - both: Standard and Unknown will use the correction. So if the coating thickness for both: Standard and Unknown will the same, the results before and after turn on the *two* coating parameter flags will the same.
It is clear.

But why  composition of Standard is not changing independent turn on or turn off the *two* coating parameter flags from the Analytical | Analysis Options dialog?
Composition of Standard will change ONLY if we change default 200A thickness (doesn't matter decrease or increase: 201 and more or 199 and less).
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 19, 2023, 09:43:38 PM
Quote from: Rom on January 19, 2023, 05:59:08 PM
Yes, you are right.
If we turn on the *two* coating parameter flags from the Analytical | Analysis Options dialog - both: Standard and Unknown will use the correction. So if the coating thickness for both: Standard and Unknown will the same, the results before and after turn on the *two* coating parameter flags will the same.
It is clear.

But why  composition of Standard is not changing independent turn on or turn off the *two* coating parameter flags from the Analytical | Analysis Options dialog?
Composition of Standard will change ONLY if we change default 200A thickness (doesn't matter decrease or increase: 201 and more or 199 and less).

Normally we do not utilize the coating correction because if both the standard and secondary standard (or unknown) have the same coating material and thickness, then as I stated in my previous post, the effects of the coatings essentially normalize out. Remember, everything in microanalysis is a ratio between the primary standard and the secondary standard (or unknown).

However there are two situations where the effects of the coatings can be slightly larger, even if the coatings are the same for both samples because these effects are non-linear. First, when we have a situation where the emitted (measured) x-ray is highly absorbed by the coating material, for example O Ka absorbed by carbon (MAC = ~12,000). And second where there is a low over-voltage situation, such as Fe Ka at say 8 keV, due to the energy loss of the electron beam.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 19, 2023, 10:42:47 PM
Yes, thank you. I expected this affect.

Do you have any ideas which can help me to understand, why composition of Standard is not changing independent turn on or turn off the *two* coating parameter flags from the Analytical | Analysis Options dialog?
Composition of Standard will change ONLY if we change default 200A thickness (doesn't matter decrease or increase: 201 and more or 199 and less).
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 20, 2023, 08:21:22 AM
Quote from: Rom on January 19, 2023, 10:42:47 PM
Yes, thank you. I expected this affect.

Do you have any ideas which can help me to understand, why composition of Standard is not changing independent turn on or turn off the *two* coating parameter flags from the Analytical | Analysis Options dialog?
Composition of Standard will change ONLY if we change default 200A thickness (doesn't matter decrease or increase: 201 and more or 199 and less).

Yes, the composition of the secondary standard (or unknown) will not change if the primary standard and the secondary standard (or unknown) have the same coating material and thickness, regardless of the Conductive Coating settings in the Analytical |Analysis Options dialog.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 22, 2023, 04:08:35 PM
Yes, but why composition of the main (not secondary) doesn't change without default cover thickness changing?
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 22, 2023, 04:32:04 PM
Quote from: Rom on January 22, 2023, 04:08:35 PM
Yes, but why composition of the main (not secondary) doesn't change without default cover thickness changing?

Because the primary standard always has the composition of the published value.  That's what it means to be a primary standard.
Title: Re: Standard/ unknown different coating?
Post by: Rom on January 22, 2023, 06:45:14 PM
But if the primary standard will have coating with thickness (chosen in PFE) 199 and less or 201 and more A, the calculated composition will change. Why the calculated composition will not change with thickness (chosen in PFE) exactly 200A?
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on January 24, 2023, 08:42:55 AM
Quote from: Rom on January 22, 2023, 06:45:14 PM
But if the primary standard will have coating with thickness (chosen in PFE) 199 and less or 201 and more A, it calculated composition will change. Why the calculated composition will not change with thickness (chosen in PFE) exactly 200A?

Hi Rom,
You are exactly correct!

Apparently I've only tested situations where the standard was uncoated and the unknown was coated with various coatings, or both the standard and the unknown were coated the same (various) coatings, or the standards were coated with the default value of 200 angstroms of carbon and the unknowns coated with (various) coatings  (generally the situation in my lab!), and of course for all of these scenarios the coating corrections were turned on.

In the situation you described where the primary standard was coated with a value different than the default coating (200 angstroms of carbon), *and* the standard was analyzed as an unknown *and* the coating correction flags in the Analytical | Analysis Options dialog were turned on, then a problem can be seen.  Though it's possible that the difference in the coatings thicknesses were small enough to not be noticed, so "good catch"!    ;D

Of course normally we use the same coating on both standards and unknowns and usually we never turn on the coating corrections, so all this time no one (until you) has noticed this issue.

Update Probe for EPMA from the Help menu and you will get this coating bug fix for standards analyzed as unknowns with non-default coatings specified.

Basically, the problem was in the calculation of the "primary" (pure element intensities in routine ZAFSetZAF) when the sample was a standard with a non-standard coating thickness and analyzed as an unknown. Here is what the old code showed when the standard was coated with 200 angstroms of carbon and analyzed with the coating corrections turned on:

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=C, Density=1.35 gm/cm3, Thickness=200 angstroms, Sin(Thickness)=311.145 angstroms

St  522 Set   2 Titanium metal, Results in Elemental Weight Percents

ELEM:       Ti       O
TYPE:     ANAL    ANAL
BGDS:      LIN     LIN
TIME:    10.00   10.00
BEAM:    30.01   30.01

ELEM:       Ti       O   SUM 
   106  99.276    .711  99.988
   107  99.842    .679 100.521
   108  99.300    .609  99.908
   109  99.230    .554  99.784
   110  99.987    .639 100.626

AVER:   99.527    .638 100.165
SDEV:     .359    .061    .381
SERR:     .160    .027
%RSD:      .36    9.60

PUBL:   99.500    .500 100.000
%VAR:    (.03)   27.63
DIFF:    (.03)    .138
STDS:      522      12

No problems with this analysis. And here with the standard coating set to 190 angstroms (again with the old code):

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=C, Density=1.35 gm/cm3, Thickness=190 angstroms, Sin(Thickness)=295.5877 angstroms

St  522 Set   2 Titanium metal, Results in Elemental Weight Percents

ELEM:       Ti       O
TYPE:     ANAL    ANAL
BGDS:      LIN     LIN
TIME:    10.00   10.00
BEAM:    30.01   30.01

ELEM:       Ti       O   SUM 
   106  99.333    .713 100.047
   107  99.900    .681 100.580
   108  99.357    .611  99.967
   109  99.287    .555  99.842
   110 100.045    .641 100.685

AVER:   99.584    .640 100.224
SDEV:     .359    .061    .382
SERR:     .160    .027
%RSD:      .36    9.60

PUBL:   99.500    .500 100.000
%VAR:    (.08)   28.02
DIFF:    (.08)    .140
STDS:      522      12

OK, it is a little different, but perhaps we can see why this was never noticed previously!

Now the same analysis but with the corrected code where the actual standard coating parameters are now loaded in the "primary intensity" calculation when the sample is a standard:

Using Conductive Coating Correction For Electron Absorption and X-Ray Transmission:
Sample Coating=C, Density=1.35 gm/cm3, Thickness=190 angstroms, Sin(Thickness)=295.5877 angstroms

St  522 Set   2 Titanium metal, Results in Elemental Weight Percents

ELEM:       Ti       O
TYPE:     ANAL    ANAL
BGDS:      LIN     LIN
TIME:    10.00   10.00
BEAM:    30.01   30.01

ELEM:       Ti       O   SUM 
   106  99.276    .711  99.988
   107  99.842    .679 100.521
   108  99.300    .609  99.908
   109  99.230    .554  99.784
   110  99.987    .639 100.626

AVER:   99.527    .638 100.165
SDEV:     .359    .061    .381
SERR:     .160    .027
%RSD:      .36    9.60

PUBL:   99.500    .500 100.000
%VAR:    (.03)   27.63
DIFF:    (.03)    .138
STDS:      522      12

Now we get the same result as before when all standards are coated with 200 angstroms of carbon!  And finally here again, but without any coating corrections turned on:

No Sample Coating and/or No Sample Coating Correction

St  522 Set   2 Titanium metal, Results in Elemental Weight Percents

ELEM:       Ti       O
TYPE:     ANAL    ANAL
BGDS:      LIN     LIN
TIME:    10.00   10.00
BEAM:    30.01   30.01

ELEM:       Ti       O   SUM 
   106  99.276    .711  99.988
   107  99.842    .679 100.521
   108  99.300    .609  99.908
   109  99.230    .554  99.784
   110  99.987    .639 100.626

AVER:   99.527    .638 100.165
SDEV:     .359    .061    .381
SERR:     .160    .027
%RSD:      .36    9.60

PUBL:   99.500    .500 100.000
%VAR:    (.03)   27.63
DIFF:    (.03)    .138
STDS:      522      12

Again, the same result because all standards have the same coating!

Remember, in situations of very high x-ray absorption by the coating, or in very low over voltage situations, the elemental compositions may be somewhat different due to differences in the coating physics between the primary and secondary (or unknown) compositions.
Title: Re: Standard/ unknown different coating?
Post by: Rom on October 19, 2023, 01:45:05 AM
Hello!
I met with some accounting issue with carbon coating thickness again.

Could you explain why changing carbon coating thickness on one standard doesn't change composition of this and other standards.

For instance, we use 2 standards: MgO and Magnetite.
We set MgO like oxygen standard for both of them (for MgO and for Magnetite).
Initially carbon coating thickness for both of them 200A.

But measured oxygen composition in MgO and Magnetite standards will not change if we'll change carbon coating thickness on MgO or on Magnetite standards.
At the same time if we change carbon coating thickness on MgO standard, oxygen composition in unknowns will change logically.
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on October 19, 2023, 07:57:25 AM
Quote from: Rom on October 19, 2023, 01:45:05 AM
Hello!
I met with some accounting issue with carbon coating thickness again.

Could you explain why changing carbon coating thickness on one standard doesn't change composition of this and other standards.

For instance, we use 2 standards: MgO and Magnetite.
We set MgO like oxygen standard for both of them (for MgO and for Magnetite).
Initially carbon coating thickness for both of them 200A.

But measured oxygen composition in MgO and Magnetite standards will not change if we'll change carbon coating thickness on MgO or on Magnetite standards.
At the same time if we change carbon coating thickness on MgO standard, oxygen composition in unknowns will change logically.

If one of the standards is the primary standard for oxygen, the oxygen composition should not change.  Did you select the two "use coating correction" options in the Analytical | Analysis Options dialog? 
Title: Re: Standard/ unknown different coating?
Post by: John Donovan on October 19, 2023, 03:03:55 PM
Quote from: John Donovan on October 19, 2023, 07:57:25 AM
Quote from: Rom on October 19, 2023, 01:45:05 AM
Hello!
I met with some accounting issue with carbon coating thickness again.

Could you explain why changing carbon coating thickness on one standard doesn't change composition of this and other standards.

For instance, we use 2 standards: MgO and Magnetite.
We set MgO like oxygen standard for both of them (for MgO and for Magnetite).
Initially carbon coating thickness for both of them 200A.

But measured oxygen composition in MgO and Magnetite standards will not change if we'll change carbon coating thickness on MgO or on Magnetite standards.
At the same time if we change carbon coating thickness on MgO standard, oxygen composition in unknowns will change logically.

If one of the standards is the primary standard for oxygen, the oxygen composition should not change.  Did you select the two "use coating correction" options in the Analytical | Analysis Options dialog?

OK, this was caused by a silly mistake in the code and is now fixed.  Nice catch!

Please update using the Help menu to the latest 13.6.0 of PFE and it will now work as expected.
Title: Re: Standard/ unknown different coating?
Post by: Rom on October 19, 2023, 06:21:22 PM
Thank you very much, it works right.