Probe Software Users Forum

Software => Probe for EPMA => Topic started by: Malcolm Roberts on October 14, 2015, 06:47:51 PM

Title: Image acquisition mode during automated analyses
Post by: Malcolm Roberts on October 14, 2015, 06:47:51 PM
Hi John
Currently automated image acquisition uses SE. I would like to get BS images instead. Where can this change be facilitated in Probewin.ini? If the default cannot be changed, how can I get PfEPMA to acquire BS images? I've not had much luck! I've had a look through the manual at the layout of probewin.ini, but cannot see anything obvious.
Cheers,
Malc.
Title: Re: Image acquisition mode during automated analyses
Post by: John Donovan on October 14, 2015, 07:57:11 PM
Quote from: Malcolm Roberts on October 14, 2015, 06:47:51 PM
Hi John
Currently automated image acquisition uses SE. I would like to get BS images instead. Where can this change be facilitated in Probewin.ini? If the default cannot be changed, how can I get PfEPMA to acquire BS images? I've not had much luck! I've had a look through the manual at the layout of probewin.ini, but cannot see anything obvious.
Cheers,
Malc.

You can select BSE image acquisition from the Imaging dialog.  So just acquire a BSE image manually and then close the Imaging window, and the default signal should be saved for the automated imaging.