adding oxford instruments Ultim Max detector

Started by Ben Buse, July 09, 2024, 08:28:32 AM

Previous topic - Next topic

Ben Buse

Hi

Has anyone added an Oxford Ultim Max detector to DTSA-II and what parameters did then use?

Thanks

ZhaiQ

I've added the Ultim Max 170 detector befoer, here're parameters I used:

1.Elevation angle
2.Working distance
3.Sample-to-detector distance
4.Detector area
5.Thickness
6.Window

I think you may want to check "Configuring instruments and detectors" from DTSA documentation, there're more details https://www.cstl.nist.gov/div837/837.02/epq/dtsa2/DTSA-II_Configuration.pdf

Nicholas Ritchie

The most important parameters are listed in the document that ZhaiQ references.

1. Take-off angle
2. Optima working distance
3. Window type
4. Detector Area
5. Distance from sample to detector

The first two are important for quant.  The later ones are more important for simulation.   The remainder are less critical.
"Do what you can, with what you have, where you are"
  - Teddy Roosevelt

Ben Buse

That's good because have no idea the detector information for some of the others - like window and detector coatings.

Nicholas Ritchie

I'm sure someone at the detector vendor does but I've never been able to pry this information out of them.  They seem to view coatings and dead-layers as trade secrets.
"Do what you can, with what you have, where you are"
  - Teddy Roosevelt