Hi

Is there anyway to read the peak and PHA from spectrometer, when 'Move' doesn't recognise the element?

Thanks

Ben

John Donovan

#1
Quote from: Ben Buse on October 30, 2025, 06:12:01 AMIs there anyway to read the peak and PHA from spectrometer, when 'Move' doesn't recognise the element?

To read all the PHA parameters for all spectrometers, you can use this handy new button:

https://smf.probesoftware.com/index.php?topic=1559.msg13512#msg13512

To read the PHA settings for a single spectrometer, use the Get PHA button in the PHA window. To read the current spectrometer position as the on-peak position for an element, use this button:

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Ben Buse

That looks nice, thanks for those options, I'll have to give them a try.

I found I can also do it through move, by selecting the element for each spectrometer first using the periodic table button, and then clicking update - to update the default positions, with the current positions. Then in elements window load from the move positions.