Cyclic patterns in quant data

Started by Ken Orr, August 19, 2019, 08:25:57 PM

Previous topic - Next topic

Ken Orr

Hi,

I'm using the polygon extraction tool to extract the quant data from my maps. When I look into the data and start plotting elements up, a cyclic pattern emerges (see attachments). I'm plotting around 40000 points from a single pyroxene grain.
This pattern occurs in all the elements. Does anyone have any ideas as to why this is happening?

Thanks.

Probeman

#1
Hi Ken,
Just a wild guess here.

Could the raw counts per pixel be close to small integers, e.g., 1, 2, 3, etc. photons?   In other words were these pixels acquired with a very short dwell time per pixel, so only a few photons were captured?  Take a look at the net intensity maps, and look at the count rate...

By the way, what was the pixel dwell time and beam current?
john
The only stupid question is the one not asked!

Ken Orr

The dwell time was 40 ms with 15 kv and 40 amps. And a 2 µm pixel size. They were 16 hour maps. I don't have net intensity maps as yet. Does 40 ms sound like that might be the issue?

Probeman

40 millisec is a pretty short time for that beam current (I assume you meant 40 nano-amps!). It could be simply due to counting statistics, once the background is subtracted.

One will see the same effect on an EDS spectrum when zoomed in (see attached).

Take a look at the net intensity maps and let us know what you see.
john
The only stupid question is the one not asked!