Automated image acquisition parameters

Started by glennpoirier, November 15, 2016, 08:26:07 AM

Previous topic - Next topic

glennpoirier

Hi All
When I request automated image acquisition during a quant run, where are the parameters coming from? I know that the mag is taken from the analytical conditions, but where is the program getting the values for signal, dwell time and number of pixels. I can usually get what I want using the imaging button in acquisition or in digitize, but it usually takes a couple of tries.

Cheers

Glenn

Probeman

Quote from: glennpoirier on November 15, 2016, 08:26:07 AM
When I request automated image acquisition during a quant run, where are the parameters coming from? I know that the mag is taken from the analytical conditions, but where is the program getting the values for signal, dwell time and number of pixels. I can usually get what I want using the imaging button in acquisition or in digitize, but it usually takes a couple of tries.

Hi Glenn,
Right now the automated imaging before and/or after each automated sample acquisition utilizes the last imaging parameters saved when the Imaging window is closed (image channel, imaging scan speed and image size). It does however, utilize the Imaging Magnification value from the digitized conditions or digitized sample setup defined in each Automate! position sample.
john
The only stupid question is the one not asked!