Flow Proportional Counter Backflow Gas Regulation
Started by DavidAdams, July 19, 2018, 09:10:27 AM
Hi Dawn,
I still have the Alcatel oil in the bubbler and it seems very stable (no discolouring, no change in the level, constant bubble rates...). I don't have a dataset similar to what Brian has done to be able to verify if and how much it reduces drift in the bias settings. I'm hoping it does not only prevent air (with changing humidity) backstreaming into the detectors but also to have at least some backpressure regulating effect.
Karsten
Below is a more or less final version of my plot of GFPC anode bias versus atmospheric pressure:
In contrast to previous versions of the plot, I've now contoured it (by hand/eye) for dew point temperature, and I've also color-coded the data.  The curves that I've drawn on the plot are given by the following equations, which work well for interpolation:
Channel 1:  bias [V] = 7 V/kPa * Patm [kPa] + 0.52 V/°C * Tdew [°C] + 909.4 V
Channel 4:  bias [V] = 7 V/kPa * Patm [kPa] + 0.76 V/°C * Tdew [°C] + 929.7 V
Some unexplained scatter is still present on the plot, and so additional variables are likely significant in addition to atmospheric pressure and dew point temperature.  For instance, I don't really have a good handle on time required for equilibration, which obviously would be particularly important when atmospheric conditions are changing rapidly.
When determining the appropriate anode bias, I should note that I used the JEOL "base level" scan rather than the "high voltage" scan and then adjusted the bias in 2 V increments until I got a distribution centered at/near 4 V.  I made the final scan using a step of 0.1 V and dwell time of 1 s; generally I repeated the slow scan at least once in order to assure reproducibility.  This is a tedious process, but it produces better results than the "high voltage" scan, which tends to have a broad, gently sloping "peak."
So I guess now I need to do something about this problem.  I'm reluctant to add liquid to the bubbler at the exhaust, as I worry about backstreaming in the event of counter window failure.
DavidAdams
Wow, Brian!! This is a fantastic data set! I'm really happy that you took the time to do such intensive testing and compile all this. This definitely shows VERY similar behaviour to what I have observed for a lot of years. As to a fix, I'm still scratching my head on that myself.
I don't think you should be too concerned about backstreaming. I've had fluid in all of my instruments and have also had window failures. I've never had any fluid pulled back into the system. The key seems to be to only fill the bubbler to just where the bubbles start and no more.
-dave
David Adams
The University of Auckland
I've attached a copy of the measurements in an Excel file in case anyone wants to examine them more carefully and/or plot them differently.
I have no experience or knowledge with Jeol probe construction, but I am very familiar with Cameca probes. I want to point to some very important physical behaviors/issues which can influence the observed optimal bias drift and is more plausible to proposed water diffusion against gas flow.
At first, It is not enough to be sure that your room temperature is rock stable. Can gas bottle be heated by floor or wall? (Or more precisely how different room air temperature gets in contrast to wall and floor temperatures (in day cycles, in seasonal cycles)? Hopefully the bottle is not outside of lab - that would be the fatal flaw). Air humidity is very important in air <-> object temperature exchange rate. The higher humidity - the faster object's temperature equalizes to the temperature of the air. Temperature of the bottle influences the temperature of P10, and so that influences the density of the gas in the proportional counter which will shift optimal bias value. Same counts for the probe, just differently to static bottle, probe has lots of heat sources and many passive and active cooling solutions. So even if Air temperature in the room is kept extremely stable, the humidity variations will change the temperature of the objects. BTW, pressure also affects the cooling rates.
I want also to point out that between setting bias value (or scanning bias values with digital interface that is Your computer and software, or values which You are displayed on the screen) there is the complicated electronic system which in different parts is actively and passively cooled. So again, any change in humidity will swing lazily/carelessly designed analog signals and that will result in observable PHA shifts and optimal bias shifts. On some instruments there are such design atrocities as analog signal squeezed-down (down-voltaged) with primitive voltage divider (pair of resistors) which will toss the PHA left and right depending from humidity... And changing proportional counters with SDD's wont change a lot. The EDS systems (where electronics part has much better engineering) has exactly same issues (spectrum shift after humidity change), at least I am aware of these in our labs.
Perhaps you could present some actual data to support your argument??
Quote from: sem-geologist on September 10, 2020, 03:56:49 AM
Quote from: Brian Joy on September 10, 2020, 06:12:29 PM
I should apologize at first. I forgot to say thank You a lot for this extensive excellent measurements and data. I wish I could had this at hand 6 years ago and could show that under nose of our management, so they would not cheapskate on air conditioning (AC) system in our lab.
I am not questioning credibility of data, I am questioning only the proposed mechanism for observed bias drifts.
So lets look to my claims step by step.
I have no humidity logs in our lab, albeit there is plan for those. The hardware is sitting and waiting for configuration, but I had no time to configure it. I have at the moment only extensive temperature data (logged every 1 minute for a bit more than a year with some brakes) from different places such as outside wall near window (old site for AC control panel (and internal thermometer used for T control), inner build wall where P10 gas bottle is presently placed by (also a new site for AC control panel), P-10 bottle metal case, backside of EPMA (air at 20 cm from floor, 10 cm from microprobe), microprobe chasis at two points (column isolation valve screw (EP6), junction of second ion pump and column).
Initial idea for thermometer network came from some weird behavior of one of spectrometers, so I wanted to see if air streams from AC are not effecting more one of the spectrometers (the one with weird behavior) and so wanted to measure separate temperatures for every spectrometer (using DS18B20, digital 1-wire thermometers connected in network). Before I started long term logging I had re-modified the network to troubleshoot vacuum leak problems (I was suspecting either metal gasket of that valve or metal gasket of second ion pump, thus the weird positions given below). Also I had expanded that network with other points in lab to prove that AC control panel was initially badly placed, and P10 Gas bottle had to be moved from external wall to internal wall (I was getting critique from management for moving stuff, and needed hard proof). After P10 gas movement we saw a lot of improvement in calibration stability, but that was not enough. Unfortunately I started logging temperatures with that network only after AC and P10 Gas placement modifications, so I can't compare directly with old setup. You can see in above picture that at winter the temperature of outer wall gets much cooler, and gas bottle (when it was standing by outer wall) was also surely affected. Also AC logic within c/p in old place was often getting very biased temperature readings. I.e. in cold winters we were getting very warm room (i.e. when -20 C outside), while I have no data record from those times, it was very obvious for human senses.
I should mention that our AC panel is set to keep in between 23 (start cooling if above) and 22 (heat up if below) degrees, and integrated thermometer reading on AC c/p is showing 22 or 23 whole year. I also should mention that previously we had in room installed some USB temperature humidity logger which was also showing very stable temperatures (Multimetrix DL 53), basically lulling us into believing that there is no issues. That useless piece of USB logger was installed by Cameca with our new probe. The "PRO" thermometers/loggers have one significant drawback - they are encased in plastic packaging and measure highly averaged temperature - they measure something from alternative reality - or like going to listen for opera but putting the bucket on the head. That is why I made network of thermometer chips (DS18B20) which are bare-exposed to the stuff which it is supposed to measure (i.e. direct contact to gas bottle with thermal paste in between chip and bottle, but well isolated from direct air blow; or other example - directly exposed/hanged in the air for air temperature measurement - so that room humidity and air streams would affect the thermometer in exactly same manner as the machine). Before assembling the network of thermometers I had tested them all by simultaneously measuring temperatures while keeping them immersed in same liquid container to see that they get exactly same (+/- 0.0625 C) temperatures in range of 30-15 degrees. So the temperature differences showed in charts are not artificial but real. DS18B20 is able to differentiate temperatures with 0.0625 C steps, which is pretty good and it is able to capture not only daily and seasonal temperature fluctuations but also individual AC cooling/heating cycles. (In the end this precision helped me to locate and fix vacuum leak).
The temperatures are pretty stable during autumn, winter and spring. As an example 4 days during stable period:
However it is pretty terrible during hot summer days:
Here comes so waited autumn, which begins the stable cycle again (the figure below depicts logged temperatures with transition from hot summer weather changed with atmospheric front going through location and lowering outside temperatures):
This I think summarizes currently gathered all temperature measurements:
So why during summer our lab gets so bad variations? I think (I still need to install humidity measurements hardware to be 100% sure) that it is so at summer as lab is coolest place compared to the surrounding rooms and halls in the building and gathers/traps humidity, that is outside from lab (inside building) and outside of building temperatures gets significantly higher during hot day than inside the lab. The increased humidity significantly increases the rate of feedback between air conditioning injected cold air streams and AC c/p (where we can see very steep negative peaks recorded on all temperature readings). As control panel logic recognizes that AC is over-cooling it switches off AC very rapidly and waits for threshold to be crossed again. Thus there is much larger cycles than during stable period, where AC works in PID mode (like i.e. water chiller; it is integrating temperature measurement and controlling AC air stream strength and temperature not so violently - thus on chart we see smaller cycles which switches much more often). And so these are the data which I base my question about if being sure that room temperature is rock stable, and being sure about all heat fluxes.
Talking about P10 gas temperature and its influence to counting, I have rather "anecdotal" experience without real measurements (sorry, no data). When we change P10 gas, old calibrations gets instantly outdated; But then after some time the newly made calibrations gets outdated, and old calibrations gets again to be good. The effect is clearly correlative with temperature differences between lab room, that is: if weather is mild (spring or autumn) and temperature of transported bottle is similar to our lab, then there is no or very week calibration misalignment observed. But if gas bottle is transported and changed at cold winter or hot summer, we get 1-2 weeks of constantly out-dating calibrations until gas temperature stabilizes with room temperature. Then we can use again old calibrations. I know that gas bottles are stored in some not-heated or cooled shed by gas producer, and so its temperature equilibrates with environment/weather before being transported to our lab.
Now talking about electronics, most of datasheets of electronic components present in one or other form how the values of described component react to temperature changes and (optionally) humidity. Resistors are particularly sensitive to temperature and so rating can change (and so voltage values, i.e. which are fed into ADC, or other way around - the digital values which are converted from digital to analog and amplified (there are always some resistors in control loops of OPAMPs in most cases)). I don't think I can give any data for that.
JEOL uses metal film resistors where necessary; these typically have tempco = 50 ppm/°C and also low voltage coefficient.  As far as op amps, even the humble 741 is specified by TI to have an average input offset voltage drift of 15 µV/°C; for variation of a few degrees C, this drift is at least an order of magnitude lower than the input offset voltage specified at 25°C (typical maximum = 1 mV, 5 mV guaranteed [for +/-15 V supplies]).  For a precision op amp, the corresponding values are at least an order of magnitude lower.
Can you present data to show that the centroid of the pulse amplitude distribution shifts position significantly (perhaps a few hundred millivolts) as a function of temperature (in any part of the lab or instrument)?  Of course, as you pointed out, the water content of the air will tend to rise as temperature of the air rises, and so the issue becomes more complicated.
I'm not trying to say that I have all the answers, but I can show conclusively that the PHA centroid varies systematically with variation in dew point temperature as well as atmospheric pressure.  Though I presented a suggestion to explain the effect, I can't prove a cause-and-effect relationship -- only a correlation.  I collected data over a period of more than a year and found the results to be largely reproducible, even when comparing data from different seasons.  Of course it's difficult to compare values from summer and winter here, as the winter dew point T in the lab rarely exceeds 5°C, and summer values typically fall between 10 and 15°C.  I never write in terms of relative humidity since a given value only applies at a specified temperature.
I also obsess over temperature in the lab (though maybe not quite as much as you), and I also have to deal with people who think that temperature variation is of no consequence.  I see effects on peak positions (for all diffracting crystals, not just PET) and count rates (including sealed Xe counters), and I attribute this to expansion and contraction of metal parts as temperature varies.  Obviously this will change the geometric relationship between sample surface, diffracting crystal, and counter window.  JEOL spectrometers are constructed almost entirely of brass, which has a relatively large coefficient of thermal expansion.
Quote from: sem-geologist on September 11, 2020, 07:38:50 AM
Quote from: Brian Joy on September 11, 2020, 02:30:20 PM
I also have obsessed about temperature stability in the lab, though now that we have that under control in our new facility it never enters my mind any longer.
I agree with Brian that PET crystals show the largest changes in intensity with respect to temperature, but I note this previous post from him on this question:
https://smf.probesoftware.com/index.php?topic=854.msg5422#msg5422
Brian: how much of the intensity change on PET crystals do you now think is due to temperature changes, and how much to exposure to the beam (electron trap issues)?  On Cameca instruments we don't need (or have) electron traps because all our crystals are behind "column separation" windows.  But I still saw significant intensity changes, especially on my PET crystals, correlated with temperature though I can't place my hands on the data at the moment.
Quote from: Probeman on September 13, 2020, 09:59:54 AM
Ah yes, fond memories...
The effects of failure of the static filter depended on Bragg angle.  I routinely monitor Ca Ka count rate, and, at this peak position, the decrease in count rate due to failure of the static filter was small (2 or 3 per-cent).  Although it pains me to do this, below I've presented a plot of Si Ka count rate on PETL at a point in time before failure of the filter and also after; both scans were collected under identical conditions.  The variation in count rate that I attribute to variation in lab temperature is in the range of 1%.  I haven't investigated the effect systematically because I'm not willing to subject the lab to abnormal temperature variation intentionally.  The changes in peak position with varying temperature are obvious, especially on PET.  At any rate, I think the plot below will answer your question.  By the way, I now monitor Si Ka count rates on the PET crystals, and I also check the static filter periodically with an electron mirror.
Just looking back through this as I'm once again having fun with fluctuating standards as weather fronts come through. We dropped about 20 mbars in 24 hours whilst I was doing low overvoltage for major elements and light elements and started seeing ever-worse totals! I think I must be running the standards more than the unknowns at this point.
The plots of barometric pressure vs detector bias chime well with what I see on both our EPMAs (SX100 and 8530F, in very different buildings). I've plotted up temperature (as measured in lab and external) vs raw count rate for our standards as part of longer runs (same beam conditions etc) and see little correlation, but a strong correlation with ambient pressure – and obviously much worse for our P10 than Xe counters (although still visible on those).
Looking at these plots:
I'd say you've got an air con unit set at 22C with a +/-1C accuracy. Of all the coloured lines, only the dark green one really tells you what is going on.
In winter, its gently warming the room to within minimum temperature (21C), which takes very little effort and results in the stable temperature profile you're seeing.
In summer, the air con only seems to be active during daytime hours, when there is a much greater rate of temperature rise (probably the Sun!). The temperature in your lab shoots up quickly above 23C, the air con switches the compressor on and blasts cold air into the lab, resulting in the rapid cooling of the air temperature. This is repeated during the day. What's interesting in this case is that overnight, your air con doesn't come on at all (or if it does, in a very low power mode), and what is probably the drop in ambient temperature brings the lab temperature back down. It seems to switch on about 7am and off again at maybe 6 or 7pm?
Jon, thanks for comment,
There are few problems causing this kind of fluctuations at summer.
(1) our external units of split AC are situated on the sunny roof top. Roof is tar type (black). We had once AC outer unit failure when outside was 35 C (in the shadow, no wind, quite high humidity), now try to imagine what a hellish conditions would be there on the roof (I guess surface temperature easily up to 60 C if not more at normal sunny day). It is badly planned, such stuff happens then institutions buy instruments and place it in some "free" space completely not designed bottom-up for that specific function.
(2) our lab is at second floor, fortunately at least from all-day-shadow side. However room has windows! and there is some reflected light (and heat) from other building. But most important thing is that the room is 3D island of cool 😎 from all sides and from bottom and up. That makes it perfect humidity sink as outside is hot, and corridor and rooms at sides and bottom and top are very warm. Humidity can raise up during day up to 70 %. The higher humidity - the worse AC can cope with keeping the set temperature.
This summer was better, as I found usage case for our second AC. Due to previous AC failure we had installed side-wall simple cheap AC as emergency AC. We normally had not run it, as pretty soon after installation I found out that its maximally-cooled wind-blow would make metal gaskets of our SXFiveFE leak. However, this summer I found out that it can work as dehumidifier and keep the humidity in check (< 60 %), which makes our main AC work more stable. In Dehumidifying mode it does not blow air, thus it does not influence our SXFiveFE directly.
Quote from: sem-geologist on October 01, 2021, 02:42:29 AM
...such stuff happens then institutions buy instruments and place it in some "free" space completely not designed bottom-up for that specific function.
I was saying something similar to a colleague of mine earlier today. Perhaps we should all start looking at running microprobes in pressure controlled fridges!
Your air con seems to be doing a decent job of cooling the lab though, it just does it all at once and then nothing. It must be like being sat in an arctic gale when it activates!
Where we are in the UK, I don't think there's a dehumidifier in existence that could get our humidity below 60%.
Quote from: JonF on October 01, 2021, 07:53:15 AM
Here is a nice schematic of "gauge" vs. "absolute" pressure:
Here's an example that goes to 30 psi, so maybe would take care of our barometric pressure induced P-10 changes from weather fronts:
https://store.mathesongas.com/3396-single-stage-high-purity-absolute-pressure-regulator-brass/
Has any tried using one of these so called "sub atmospheric" (absolute pressure) gas regulators for their P-10 flow?
The problem is that with sub-atmospheric pressure You will have no flow, as gas from lower pressure wont move to higher, but contrary. The only way for this to work would be to release sub-atmospheric pressure P10 into lower pressure (i.e. vacuum buffer). Of course we don't want to release it into current vacuum system (even primary) as that could contaminate vac system; however I think using a small low power pump (most important something not loud, as labs are already buzzing annoyingly) for such buffer for releasing gas from low-pressure counters could be very financially feasible and quite simple to achieve. This is really interesting concept, it could at last put some end to endless redoing of standard measurements forced by weather change... unless gas is not the only weakness (but probably it is as we see no such fluctuations on high-pressure spectrometers despite weather changes).