WDS Detector Dead Time calibration acquisition now in StartWin application
Started by John Donovan, December 16, 2018, 04:14:26 PM
Quote from: Brian Joy on March 27, 2022, 07:32:02 PM
I'd encourage everyone to measure deadtimes for multiple X-ray lines per spectrometer and not to rely on just Si Ka and Ti Ka.  It's a time-consuming but worthwhile exercise.  I see variation in deadtime from one X-ray line to another, but I can't see any obvious pattern to it -- it is not a simple function of X-ray energy.
Brian,
Had You done those measurements in differential mode or in integral mode?
I don't know how Jeol (and if) had cut corners in pulse processing... Are Your tests done one by one spectrometers or all at once?
I as well try to stay in low count range (for Cameca probes that is about 10kcps) where pulse-pileup probability is below 0.5% in integral mode. But I am looking forward to improve this and go up to reliable 50kcps with differential mode and modified equation of deadtime for diff mode (which needs to take into account counts of integral mode). At first I want to finish gathering the evidences and fact-checking, before presenting the full picture (This is going to take some time). I want to emphasize that deadtimes on WDS are partly from different sources than those on EDS. In coming separate post I will go through pulse pipeline and discuss the parts which majorly or only minorly can influence the dead time and how (and why) it differs from EDS counting. As for Cameca probes, currently I had done live oscilloscope based observations of low pressure (on SX100, last year) and high pressure (on SXFiveFE, revisiting this issue) GPC generated signals, and results (at first surprising a bit) are quite different, but well explainable taking all pieces together. There are many factors which could influence or for sure influence the miss events of some pulses.
Quote from: sem-geologist on March 28, 2022, 12:10:57 AM
Brian,
I use integral mode and only use results up to about 30 kcps.  In this way, I avoid large shifts in the pulse amplitude distribution and stay within the roughly linear region on a plot of cps/nA versus cps.  For a given X-ray line, I use as many spectrometers simultaneously as possible.
If someone using EDS is looking to this thread and think "oh those WDS-dudes, we on EDS have none of these dead time problems" - I would like to curb Your enthusiasm, especially if Your machine is running Bruker/Rontec EDS SDD detectors.
I had just discovered that dead time (live time) estimations can have very huge uncertainties in case the beam is blanked before running the analysis or mapping. Long counting (>30s) is not so affected, but short counting (<1s, few hundred of ms of pixel  dwell time) i.e. for mappings will be a lot affected. Live time of EDS is counted with pulser method (0eV pulse at set frequency, if it miss - it means the counting electronics is busy). Frequency of pulser on Bruker signal processing unit is set (absolutely automatically, with no programmable access) depending from input counts which are constantly monitored and value just before launching the EDS acquisition is used for automatic setup of pulser frequency. In case it is 0cps (beam is blanked) it decides it is very low count rate and set that to (on SX100 tailored Bruker Nano XFlash) 1000Hz. In case input count rates would be near 1Mcps it would set 50000Hz (and other frequencies in between scaling to the input count rate). It does not depend from selected throughput. For mapping, where pixel dwell time is few hundred ms it will produce severely shortened live time and net counts from such short dwell time maps will be not comparable with long EDS measurements of standard material.
The fix is to add 2s waiting time before the start of acquisition (beam-on), so that SPU would get adequate estimation of input count rate and would select adequate zero peak frequency. There is caveat. if mapping starts in the epoxy (left-top corner) and beam would wait those 2 s there, SPU would get quite underrated count rate and would select underestimated 0eV peak frequency.
Quote from: sem-geologist on April 12, 2022, 05:07:14 AM
... dead time (live time) estimations can have very huge uncertainties in case the beam is blanked before running the analysis or mapping...
This is exactly correct.
We have noticed this also on the Thermo EDS (NSS or Pathfinder) systems as well:
https://smf.probesoftware.com/index.php?topic=939.msg8735#msg8735
One solution is to select a fixed dead time constant for the system (do not use the Auto dead time mode) based on a typical spot on the sample, or as suggested, allow the system to "equilibrate" by adding a delay of several seconds after the cup is removed but before the EDS counting starts.
Dan R
Hi all, since my instrument (JEOL 8530F) is ~10years old and i had some extra time to play with newer PFE features, I decided to use Startwin to automate deadtime measurements.
Using Paul's xls sheet, two deadtimes are given, one fitting all of the data (DT us All) and one fitting only the higher beam currents (DT Last)-- do you have a rule of thumb as to which I should input into the scalars.dat file?
I updated my scalars.dat file in line ~77, so that should override the parameters specified in Line 13 correct?
Here are my values (micro-sec) for comparison over time:
Initial DT (Kremser, 2013)           2023 Values (full-fit, Si Ka on TAP/PETJ)
Sp1   1.72                                  1.85
Sp2   1.49                                  1.25
Sp3   1.70                                  1.43
Sp4   1.65                                  1.42
Sp5   1.68                                  1.70
Do these changes make sense?
Quote from: Dan Ruscitto on January 04, 2023, 11:46:36 AM
Hi all, since my instrument (JEOL 8530F) is ~10years old and i had some extra time to play with new features, I decided to use Startwin to automate deadtime measurements.
Sp1   1.72                                  1.85
Sp2   1.49                                  1.25
Sp3   1.70                                  1.43
Sp4   1.65                                  1.42
Sp5   1.68                                  1.70
The values don't seem too unreasonable though some are maybe a little high, but the instrument is pretty old. Have you ever had any detectors replaced?  The Xenon detectors can age particularly fast and should be replaced about every 5 years or so.
But I quickly want to point out that this dead time calibration method is now considered obsolete as it is only good to around 50 kcps. Instead you should be utilizing the much discussed "constant k-ratio" method (combined with the new logarithmic dead time expression), which allows one to perform quantitative analyses at count rates up to 300 or 400 kcps. More information on the "constant k-ratio" can be found in this topic and you might want to start with this post here, though the entire topic is worth a read:
https://smf.probesoftware.com/index.php?topic=1466.msg11416#msg11416
In fact, if you have a fairly recent version of Probe for EPMA, the step by step directions for this "constant k-ratio" method are provided as a pdf from the Help menu, but I've also attached the pdf to this post as well.
Please be sure to ask if you have any questions at all.