New method for calibration of dead times (and picoammeter)

Started by Probeman, May 26, 2022, 09:50:09 AM

Previous topic - Next topic

John Donovan

#180
With regards to Probeman's post above we've updated our Constant K-Ratio method document which is attached below (remember to log in to see attachments), to emphasize the importance of constructing the k-ratio with both the primary and secondary standards being measured at the same beam current for each beam current measurement.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Probeman

Quote from: Probeman on March 11, 2024, 09:47:38 PM
Quote from: Probing on March 11, 2024, 07:53:19 PM
Quote from: Probeman on March 11, 2024, 03:12:32 PM
Quote from: Probing on March 11, 2024, 03:02:33 PM
Quote from: Probeman on May 17, 2023, 09:26:51 AMWe are pleased to announce the publication of our new paper on improving dead time corrections in WDS EPMA:

John J Donovan and others, A New Method for Dead Time Calibration and a New Expression for Correction of WDS Intensities for Microanalysis, Microscopy and Microanalysis, 2023

https://academic.oup.com/mam/advance-article-abstract/doi/10.1093/micmic/ozad050/7165464

How to get the "best" dead time constant which produces the "best" zero slope "k-ratio v. current" line in your constant k-ratio method? I don't have the PROBESOFTWARE, so can I still perform dead time calibration by applying the constant k-ratio method?

Yes, you can perform the dead time calibration using the constant k-ratio method for general use as described here:

https://smf.probesoftware.com/index.php?topic=1466.msg11102#msg11102

But if you do not have the Probe for EPMA software, you will not be able to take advantage of the new non-linear expressions described in our paper for count rates above 30 to 50 kcps. In other words, you could perform the dead time calibration using the constant k-ratio method using the traditional linear expression, but you probably won't see a huge difference in the dead time value you obtain below 30 to 50 kcps.  Above those count rates, the tradition linear expression fails as shown in the linked post below.

That said, I think the constant k-ratio method is much easier to use and more precise for dead time calibrations whatever dead time expression is being used...

See here also:

https://smf.probesoftware.com/index.php?topic=1466.msg11173#msg11173

I like to know how the ideal dead time constant is produced. In the example of "TiO2/Ti"  shown in your post, when you found the DT constant (1.32 us) was too high, you dropped it  to 1.28 us and got a more "flat" line. So my question is why it is 1.28, but not 1.27 or 1.29. How is the exact number produced, using some algorithm or by repeated iterate and trial? If the later is the case, how do you determine the observed "k-ratio v. current" line, which depends on the exact DT constant, is the best?

It is easy. One simply adjusts the dead time constant (depending on which dead time expression is being utilized (because the dead time constant is actually a *parametric* constant as described in the paper) in order to obtain a trend with a slope close to zero.

Therefore, the goal is to obtain a dead time constant (with a suitable dead time expression) that yields a flat (zero slope) response from low count rates to the highest possible count rates.  One can do this visually by inspection because a zero slope is easy to evaluate.

Obviously at some point (above several hundred kcps) the dead time correction becomes very large and quite sensitive to small changes in the dead time constant.  But we found the logarithmic dead time expression can yield quantitative results from zero to several hundred kcps count rates once the dead time value was properly adjusted.

Don't forget, as described in the paper one can also utilize the same constant k-ratio dataset to check their picoammeter linearity and also the agreement of (simultaneous) k-ratios from one spectrometer to another (and even from one instrument to another given the same materials), which can be used to check one's effective take-off angle for each (WDS and EDS) spectrometer.

It should be pointed out that one can switch simply to the logarithmic WDS dead time correction method and keep using the dead time constant that was originally obtained using a linear extrapolation (limited to 30 to 40 kcps throughput):



Why do I say this?

Because even though you might be slightly over estimating your dead time at very high throughput (> 50 kcps) using the logarithmic expression, your accuracy will still be significantly better than using the old (traditional) linear dead time correction method:



And also, below 50 kcps the linear and logarithmic expressions give essentially identical results.
The only stupid question is the one not asked!

dcoulthard319

#182
Hello,

Last week during our first PFE lab visit, we were getting in the weeds and gathering some data for calibration of dead time constants. I am still in the process of sourcing our standard blocks here at Colorado State, so I'm left to work with the Smithsonian suite of minerals and the JEOL maintenance block. Our channel loadout was initially Ch1 and Ch2 = Si on TAP, Ch3-5 = Zr (la) on PET (PETJ on Ch3 and PETLs on 4 and 5). We programmed a run to last a couple hours. Since we decided to use Si ka to calibrate channels 1 and 2, we focused our work on the NMNH Quartz and San Carlos olivine reference materials; setting the beam to 15 kV, with current ranging between 5- and 200 nA.

The attached spreadsheet contains the data from this initial run as well as the one that followed, which focused on the JEOL reference materials Zirconium and Zirconia. Contained in sheets 1 and 2, respectively. Data gathered using Channel 1 formed k-ratios that formed a linear array that was fairly easy to correct back to constant k-ratios using the methods described here and using the on-board pdf available in PFE. However, when we plotted the k-ratios for Channel 2, we observed a non-linear shape to the line with low calculated k-ratios at low current, rising to a maximum at ~40 nA and decreasing again towards our maximum current setting. Compare the signals from Channels 1 and 2 here:



We decided to take a closer look at the counts from these spectrometers, and they appeared to be linear throughout the session:



After scratching our heads for a little bit, we decided to repeat the entire procedure using PET crystals, which are available on all spectrometers. We again observed a hook-shaped curve in the low-current k-ratios on channel 2 that is not observed on any other channel. Note that the following figure was made after dead time constants were estimated for this spectrometer:



Again, the counts themselves (as cps) appear to be linear:



Now, I recognize this is not a substantial effect. Using the average k-ratios available, I calculated a 0.8% relative effect (max) from our observations. The critical part is that these effects are occurring where a vast majority of our analyses are going to be taking place (i.e., below 50 nA), and I'd like to understand this seemingly unique behavior better. Any input on this matter would be greatly appreciated!

~Danny C

John Donovan

#183
It's amazing how sensitive the constant k-ratio method is for seeing instrumental artifacts. The above plots showing raw counts vs. beam current look linear but of course the k-ratio plots show they are not. You read our paper attached below?

https://smf.probesoftware.com/index.php?topic=1466.msg12465#msg12465

Your data is very interesting.  The artifacts you are seeing are relatively small but clearly significant. 

I might have said that a hump like that could be induced from a non-linearity in the picoammeter, but since it only affects a single spectrometer, it can't be that since that would affect all spectrometers.  Same for the composition of the standards. If they were heterogeneous, you would see that trend on all the spectrometers.

It could be from a PHA adjustment issue, for example  if the PHA depression was worse in one spectrometer than than another.  That is why I suggest making these constant k-ratio measurement using a completely wide open PHA in INTEGRAL mode where the PHA peak is *above* the baseline at the high beam current in the highest concentration material:

https://smf.probesoftware.com/index.php?topic=1466.msg11636#msg11636

As the count rate decreases at lower beam currents the PHA peak will shift up, but if you are in integral mode you will still see all the intensities even as the PHA peak shifts beyond the visible range to the right:

https://smf.probesoftware.com/index.php?topic=1466.msg11450#msg11450

The good news is that your variation between spectrometers is pretty small. Worth trying to figure out, but not anywhere as bad as some instruments:

https://smf.probesoftware.com/index.php?topic=1466.msg11531#msg11531

First, check your PHA adjustments and make sure they are wide open (increase gain until all PHA peaks are above the baseline at the highest current in the highest concentration material) and that you are in INTEGRAL mode and try again.

A difference in the effective takeoff angle between spectrometers would explain the vertical differences between the spectrometers, but again, why one spectrometer's k-ratios would be sloped positive and then negative I can't explain.  It's almost as though the counting electronics has two different pulse processing ranges (dead times).

Would it be possible for your tech to swap the spectrometer electronics with another spectrometer to see if the problem moves with the electronics?
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

#184
As mentioned previously, the constant k-ratio method is surprisingly sensitive to instrumental artifacts. Danny kindly sent me his k-ratio measurement data files and I took a look, and while I can't quite explain everything I see, it is interesting.

First let's start with his Si Ka k-ratio measurements on San Carlos olivine (secondary std) and SiO2 (primary std) from 5 to 200 nA.  I noticed that he had not turned off the standard intensity drift correction, so I did that, though it didn't change things much because he had acquired his primary and secondary standards in "pairs" with the primary standard measured first for every k-ratio pair. Here's the unaveraged data for Si Ka on SP1 (TAP) and SP2 (TAP):



The k-ratios are pretty close, but as Danny mentioned, it is bothersome that the greatest divergence in the k-ratios occurs at beam currents 40 nA and below.

Next I disabled all the primary (SiO2) stds except the first one to look at the picoammeter linearity, because by measuring our primary and secondary stds in pairs, we null out any effects from picoammeter non-linearity. Now we see this:



So the picoammeter isn't quite as linear as we would like, but there's more going on here than just that.

I then wondered if the downward trend in the intensities for sp2 could be due to the dead time constants, because I noticed that all the spectrometer dead times were the JEOL default 1.1 usec.  Adjusting the sp2 dead time from 1.1 usec to 1.0 usec I obtained this plot:



The dead time constant primarily affects the high intensity count rates, so this makes sense.  My best guess at this time is that there is a very small alignment problem in sp2 that is compounded by some non-linearity in the picoammeter.

Interestingly, Danny also sent me some Zr La k-ratio measurements he did on all his PET crystals (all 5 spectrometers), going up to 1000 nA(!) and the results are actually quite (amazingly) good:



This is actually what we want to see for all spectrometers for all Bragg crystals!

Can your instrument perform as well?  Measure some constant k-ratios on your own spectrometers (both EDS and WDS) over a range of beam currents, and find out!
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"