Donovan and Moy (DAM) backscatter correction

Started by John Donovan, September 19, 2023, 02:02:05 PM

Previous topic - Next topic

John Donovan

Based on our recent paper:

https://academic.oup.com/mam/article-abstract/29/4/1436/7224307

Aurelien Moy and I have a developed a new backscatter correction (which we call the DAM backscatter correction, get it?), which utilizes Z fraction averaging as opposed to traditional mass fraction averaging. This new backscatter correction is especially important for situations with large atomic number effects, particularly when compounds contain elements with different A/Z ratios as discussed in the above paper.

Here is the new ZAF/Phi-rho-z dialog:



We're basically using the Armstrong/Brown absorption correction and replaced the mass based Love/Scott backscatter correction with our new Z fraction based correction. Here's a good example of what this new BSE correction can do, where we've analyzed PbS using FeS2 as a sulfur standard, first using the default Armstrong/Brown/Love-Scott correction:

St  731 Set   4 Galena U.C. #7400, Results in Elemental Weight Percents

ELEM:       Fe       S      Pb
TYPE:     ANAL    ANAL    ANAL
BGDS:      LIN     LIN     EXP
TIME:    60.00   60.00   60.00
BEAM:    29.88   29.88   29.88

ELEM:       Fe       S      Pb   SUM 
   440   -.001  14.229  87.048 101.276
   441    .003  14.258  86.510 100.771
   442    .010  14.303  86.553 100.867
   443    .020  14.216  86.300 100.535
   444    .005  14.307  86.617 100.930

AVER:     .007  14.263  86.606 100.876
SDEV:     .008    .042    .275    .269
SERR:     .004    .019    .123
%RSD:   107.77     .29     .32

PUBL:     n.a.  13.400  86.600 100.000
%VAR:      ---    6.44   (.01)
DIFF:      ---    .863   (.01)
STDS:      730     730     731

STKF:    .4276   .5015   .8698
STCT:   319.56  459.23   72.26

UNKF:    .0001   .1520   .8697
UNCT:      .07  139.21   72.25
UNBG:     4.07    1.28     .90

ZCOR:    .8501   .9383   .9958
KRAW:    .0002   .3031   .9999
PKBG:     1.02  109.75   81.62

Note the 6.4% relative error in the sulfur value when extrapolating from FeS2. Now the same material, using the DAM backscatter correction:

St  731 Set   4 Galena U.C. #7400, Results in Elemental Weight Percents

ELEM:       Fe       S      Pb
TYPE:     ANAL    ANAL    ANAL
BGDS:      LIN     LIN     EXP
TIME:    60.00   60.00   60.00
BEAM:    29.88   29.88   29.88

ELEM:       Fe       S      Pb   SUM 
   440   -.001  13.138  86.989 100.126
   441    .003  13.167  86.467  99.636
   442    .010  13.209  86.515  99.734
   443    .019  13.128  86.257  99.404
   444    .005  13.213  86.577  99.795

AVER:     .007  13.171  86.561  99.739
SDEV:     .008    .039    .268    .263
SERR:     .003    .018    .120
%RSD:   107.77     .30     .31

PUBL:     n.a.  13.400  86.600 100.000
%VAR:      ---   -1.71  (-.05)
DIFF:      ---   -.229  (-.04)
STDS:      730     730     731

STKF:    .4266   .5045   .8526
STCT:   319.56  459.23   72.26

UNKF:    .0001   .1529   .8525
UNCT:      .07  139.21   72.25
UNBG:     4.07    1.28     .90

ZCOR:    .8246   .8612  1.0154
KRAW:    .0002   .3031   .9999
PKBG:     1.02  109.75   81.62

Note that the relative error is now only ~1.7% using this new Z fraction based backscatter correction method.  Now let's look at the Pouchou k-ratio dataset for compounds with larger than 10% atomic number corrections. The traditional Armstrong/Brown/Love-Scott method shows this error distribution for the Armstrong/Brown/Love-Scott correction:



And with the new DAM backscatter correction we obtain this error distribution:



A considerable improvement for large atomic number corrections.

We believe this correction can be further improved by calculating the optimized average Z BSE exponent based on the electron beam energy and we are modeling this now and will implement it soon.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

#1
Re-processing all 826 binary k-ratios in the Pouchou2.Dat file, we obtain this error distribution using the Armstrong/Love-Scott matrix correction:



Not too bad, but now using the Armstrong/Donovan and Moy (BSC) correction we obtain this error distribution:



A significant improvement (take a look at the y axes)!
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

We've also now added the ability to specify a variable exponent for the Donovan and Moy (DAM) backscatter calculation as shown here:



The calculated "optimized" exponent is based on the electron beam energy from PENEPMA Monte Carlo modeling of pure elements and compounds at various electron beam energies.

If all your elements are run at the same beam energy you could just specify the exponent explicitly. The optimized exponent is calculated as shown here by Aurelien Moy:

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

#3
Just a reminder that everyone should change their default matrix correction in Probe for EPMA to option 11:



This can be done by editing the probewin.ini file in the [software] section:

https://smf.probesoftware.com/index.php?topic=8.msg14405#msg14405

Here's an example of analyzing sulfur in galena (PbS) using pyrite (FeS2) as a primary standard:

https://smf.probesoftware.com/index.php?topic=1831.msg14064#msg14064

Note that this improvement in the physics models applies not just to Probe for EPMA but also to our CalcZAF/Standard free software. Though if you haven't updated your Probe Software applications in a while, you may need to update to take advantage of this improvement in accuracy.

Here is an example of analyzing PbSiO3 using SiO2 as a primary standard using all matrix corrections:

Summary of All Calculated (averaged) Matrix Corrections:
St  386 Set   8 Alamosite (PbSiO3)
LINEMU   Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV

Elemental Weight Percents:
ELEM:       Si       S      Fe      Mg      Pb      Zr      Zn      Hg      Th      Hf      Sb       O   TOTAL
     1  10.322    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.412   Armstrong/Brown/Scott-Love (prZ)
     2   9.168    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.258   Philibert/Duncumb-Reed
     3  10.126    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.216   Heinrich/Duncumb-Reed
     4   9.777    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.867   Love-Scott I
     5   9.827    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.917   Love-Scott II
     6   8.382    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  98.472   Packwood Phi(prZ) (EPQ-91)
     7  10.624    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.714   Bastin (original) (prZ)
     8   9.731    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.821   Bastin PROZA Phi (prZ) (EPQ-91)
     9   9.647    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.737   Pouchou and Pichoir-Full (PAP)
    10   9.560    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.650   Pouchou and Pichoir-Simplified (XPP)
    11   9.823    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.913   Armstrong/Donovan and Moy BSC/BKS (prZ)

AVER:    9.726    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  99.816
SDEV:     .591    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .591
SERR:     .178    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000

MIN:     8.382    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939  98.472
MAX:    10.624    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.714

Which one is most accurate? Here are the relative percent errors for all matrix corrections:

Percent Variances:
ELEM:       Si       S      Fe      Mg      Pb      Zr      Zn      Hg      Th      Hf      Sb       O
PUBL:    9.910    n.a.    n.a.    n.a.  73.151    n.a.    n.a.    n.a.    n.a.    n.a.    n.a.  16.939
STDS:       14     730     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---

ELEM:       Si       S      Fe      Mg      Pb      Zr      Zn      Hg      Th      Hf      Sb       O
     1    4.16     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Armstrong/Brown/Scott-Love (prZ)
     2   -7.49     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Philibert/Duncumb-Reed
     3    2.18     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Heinrich/Duncumb-Reed
     4   -1.34     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Love-Scott I
     5    -.83     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Love-Scott II
     6  -15.42     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Packwood Phi(prZ) (EPQ-91)
     7    7.20     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Bastin (original) (prZ)
     8   -1.80     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Bastin PROZA Phi (prZ) (EPQ-91)
     9   -2.65     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Pouchou and Pichoir-Full (PAP)
    10   -3.53     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Pouchou and Pichoir-Simplified (XPP)
    11    -.87     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---           Armstrong/Donovan and Moy BSC/BKS (prZ)

AVER:    -1.85     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00       
SDEV:     5.97     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00       
SERR:     1.80     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00       

MIN:    -15.42     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00       
MAX:      7.20     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00       

This was at 12 keV.  Here's another PbSiO3 analysis using just the Armstrong phi-rho-z DAM backscatter at 14 keV:
St  386 Set   9 Alamosite (PbSiO3), Results in Elemental Weight Percents
 
ELEM:       Si       S      Fe      Mg      Pb      Zr      Zn      Hg      Th      Hf      Sb       O
TYPE:     ANAL    ANAL    SPEC    SPEC    SPEC    SPEC    SPEC    SPEC    SPEC    SPEC    SPEC    SPEC
BGDS:      LIN     LIN
TIME:    60.00   60.00     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---
BEAM:    19.98   19.98     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---

ELEM:       Si       S      Fe      Mg      Pb      Zr      Zn      Hg      Th      Hf      Sb       O   SUM 
  1167   9.959    .000    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.049
  1168   9.980   -.002    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.068
  1169   9.924    .002    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.016
  1170   9.949   -.008    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.032
  1171   9.951   -.002    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.040
  1172  10.063   -.006    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.147

AVER:    9.971   -.003    .000    .000  73.151    .000    .000    .000    .000    .000    .000  16.939 100.059
SDEV:     .048    .003    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000    .047
SERR:     .020    .001    .000    .000    .000    .000    .000    .000    .000    .000    .000    .000
%RSD:      .49 -133.59     .00     .00     .00     .00     .00     .00     .00     .00     .00     .00

PUBL:    9.910    n.a.    n.a.    n.a.  73.151    n.a.    n.a.    n.a.    n.a.    n.a.    n.a.  16.939 100.000
%VAR:      .62     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---
DIFF:     .061     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---
STDS:       14     730     ---     ---     ---     ---     ---     ---     ---     ---     ---     ---

Note that these analyses were using the Henke MAC for Si ka in Pb as the FFAST value seems to be problematic. Here are the various tabulated MACs showing the FFAST value seems to be an outlier:

MAC value for Si Ka in Pb =    1882.24  (LINEMU   Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV)
MAC value for Si Ka in Pb =    1929.00  (CITZMU   Heinrich (1966) and Henke and Ebisu (1974))
MAC value for Si Ka in Pb =    1799.37  (MCMASTER McMaster (LLL, 1969) (modified by Rivers))
MAC value for Si Ka in Pb =    1915.70  (MAC30    Heinrich (Fit to Goldstein tables, 1987))
MAC value for Si Ka in Pb =    1929.40  (MACJTA   Armstrong (FRAME equations, 1992))
MAC value for Si Ka in Pb =    1584.08  (FFAST    Chantler (NIST v 2.1, 2005))
MAC value for Si Ka in Pb =    1917.00  (USERMAC  User Defined MAC Table)

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"