Using the new plot option of relative percent variance (i.e., accuracy) in Probe for EPMA from the Output | Plot Standard and Unknown XY plots menu dialog:

https://smf.probesoftware.com/index.php?topic=40.msg14388#msg14388

we can quickly evaluate our secondary standard accuracy. Here for Mg Ka in MgAl2O4 using MgO as a primary standard in the FIGMAS mount from Will Nachlas:



All within 1% relative accuracy. But since we're plotting relative percent accuracy, we can plot more than a single compound, here Ti Ka in SrTiO3 and RbTiOPO4 using TiO2 as a primary standard:



And here for Fe ka in a multitude of compounds using Fe metal as the primary standard for the sulfides, and magnetite as a primary standard for the oxides, silicates and glasses:



Note that only the natural chromite is significantly outside the 1% variance accuracy limits.
The only stupid question is the one not asked!

John Donovan

In order to utilize the accuracy checks shown in the above plots, one must run their secondary standards as standard samples. Therefore, some of you have asked: how do I run my standards more than once?

Well first, to re-run ones standard samples a second time, just check the "Acquire Standard samples (again)" checkbox.  Even if you don't have an unknown sample selected, the standards will run a second time.

To run your standard even more replicates, just create an unknown sample and check the Re-Standardization Interval checkbox and enter a short interval (in hours) as seen here:



Say every hour or so. That's easy, right?
John J. Donovan, Pres. 