Dominik

I am measuring small rutile inclusions in mostly titanite. I first did with 20 kV, as we are interested in trace elements. The totals were a little high, but almost OK, but we often got mixed analyses from the adjacent titanite. So I switched to 10 kV, less influence by the titanite, but the TiO2 is up between 2-4 wt% (net Ti cps are up 4-6%) compared to the standard, which always gives a fine ~100 wt% TiO2. The rutile is in the µm range, no idea what the problem is. Some effect due to the grain size?
I use 120 nA current (for both 10 and 20 kV), again to get a good signal of trace elements.
I use a JEOL 8530F-Plus and the standard JEOL software.

Thanks, Dominik

Ben Buse

Hi Dominik

What standard are you using? What's the count rate for Ti on your spectrometer, is it running into deadtime >60k on either standard or TiO2

Ben Buse

If it's worse at 10kV also need to think about are both standard and unknown coated with same thickness of carbon. If a pure metal standard does it have an oxide layer at surface.