CAMECA Penning gauge cable

Started by Ben Buse, November 18, 2026, 05:37:02 AM

Previous topic - Next topic

Ben Buse

Hi All,

Anyone have a spare CAMECA Penning gauge cable for connecting Agilent R0310301 penning gauge to Varian Multi-gauge controller,. OR know where one can be purchased for a reasonable price. Seems to be penning gauge is SHV and Varian controller is R232, but also that there are connectors between the two,


sem-geologist

?!? Which SX model is it? Penning Gauge on SX100 I know was Alcatel. The mentioned R0310301 is Inverted Magnetron Gauge (different technology from Penning). As it is standard part, Agilent sells standard cables (Still should, as They still sell same model IMG, and they took over Varian).
I.e: R03113010.

BTW what exactly do you mean by "R232"? I guess you mixed up protocole (RS232) which commonly use D-sub 9 pin? D-sub 9 pin is used for ConvecTorr Agilent low vacuum gauges as these do not require high voltage but is just higher current devices. There is no way D-sub would be used for 4kV connection, I believe you had to mix-up which cable connects to which card of Agilent/Varian Gauge Controller. IMG should be connected with High voltage coaxial cable connectors from both sides of cable ( I am not sure it is SHV-SHV  Varian/Agilent loves custom connectors).

I realy advice against buying used high voltage cables from uncertain sources (i.e. e-bay) as they could be worn out (i.e. used in humid environment and dialectric material partly broken by HV leakage), also clueless random folks tend to pack cables wrongly (these cables have minimal radius for bending - bend more and you damage dialectric - apply High Voltage and a leak is guarrantied). I completely understand that original Agilent cables cost an arm and a leg - but that is high voltage cable - not some USB phone charger cable.

What about spare cable? Why anyone would have it, unless it is spare and leaky (broken, but forgoten that it is broken). If it is only for test purpoises (to test IMG) - again it could result in False False indentification of problem (IMG would not strike because of leak in the cable - this would result in assumtion that IMG is broken, while it would be actually a cable which is broken).


Here I found this on Ideal vacuum:
https://www.idealvac.com/en-us/Varian-Agilent-XGS-600-Multi-Gauge-SenTorr-Gauge-Controller-to-IMG-100-Sensor-CABLE-10-ft/pp/P104125

however standard 3m/or 10 feet Agilent/Varina cable could be too short. If you can confirm that it is SHV - SHV cable - then there are few (budget friendly and safe) options for custom built depending from IF your gauge is bakable, or not (Is it SXFiveFE or SXFive or SX100?). If it is not bakable then standard SHV could work (does not need expensive PEEK isolator in connector).
I.e. there is this www.koax24.de where you could order custom lenght SHV-SHV assembled cable.



If it is bakable, then I would try asking at https://www.hivolt.de/   ... but I guess You will find then Agilent official custom cables less expensive.

Ben Buse

Hi Thanks for your reply. SX-100. We've old agilent heads and an old varina controller, which I wanted to rig up, as to the setup on the CAMECA see attached images, cables don't seem same as what Agilent sell.

sem-geologist

#3
Thanks You Ben - This is indeed very interesting to me as our SX100 is 28 years old and I am also looking into what could be some of vacuum parts replaced with. I was wondering what had Cameca replaced the indestructable Alcatel "PI" low pressure detectors with (these works already for 28 years! never changed), after Alcatel merged into Adixen. So your SX100 was produced at transition into SXFive, and Cameca was on move to Agilent technology. We have also SXFiveFE - so I see the final shape of technological transition, but I was sure there had to be some middle transitional gap, when Cameca had to partially switch to Varian(Agilent) Convector from Alcatel "PI" low vacuum gauges.

What I see in this pictures is a standard Varian (Agilent) gauge controller and it has two cards for ConvecTorr low vacuum gauges with double D-sub 9 connectors for 2 gauges per card. What we see in pictures are 4 identical cables for 4 gauges (jauges in french). These d-sub 9 connectors can't power any High Voltage Gauge none of these: Cold Cathode, Penning or IMG – D-Sub 9 pin connector would spark already at 1kV. 2 D-sub 9 pin connectors are dedicated explicitly to  ConvecTorr gauges (they are similar to Pirani gauges, but much faster and with much larger longevity). We have similar Agilent controller on our SXFiveFE which aditionally to 2 ConvecTorr cards have 2 cards for IMG (there are 2 IMG on SXFiveFE, and there should be 1 IMG on simple SXFive) - the change at SXFive was that vacuum got its own additional 19 inch crate, where such Agilent/Varian controller is exposed at rack to the user. In your SX100 case I see it is put somewhere under column? with user panel hidden from user. What I dont understand is that I see in picuture only 4 slots, while there is supposed to be 5. and only 2 is populated (I see no card for IMG).
BTW 4 cables is exactly same number as there is 4 low vacuum points measured on SX100: 1)backing vacuum of Diffusion Pump 2) backing vacuum of Airlock (and isolation line) 3) WDS vacuum 4) chamber vacuum when at low vacuum.
Missing IMG or Cold Cathode card means that your SX100 for measuring High vacuum of the chamber still uses Alcatel Penning Gauge. If it is Penning Gauge which you have problem with then PM me -I have a ready drop in solution. If it is ConvecTorr - these are quite cheap to find and buy officially from Agilent, –both: ConvecTorr gauges, and cables and they should be interchangable with Varian controller - as it was Varian which invented them.

sem-geologist

Oh, I think I also unraveled your controller:
https://skycraftsurplus.com/products/varian-l8350-301-multi-gauge-controller.html

We dont see the 5th slot - as it is much further upward (out of your picture).