Amphibole data reduction
Started by njdeems, November 22, 2013, 11:25:19 AM
njdeems
I'm trying to get rid of some bad points in some EMP data I collected on amphiboles a while ago. I figure there's a cut-off on what one might consider a "1st glance bad point" with regards to total wt. % but I don't know what people generally use. Any suggestions? I also figure I could eliminate any points outside of 2sigma, but I don't know if I should calculate the St. Dev before or after eliminating "first-glance" bad totals. Any suggestions would be greatly appreciated.
Nik Deems
Hi Nik,
Are you asking what is too low an acceptable total for an amphibole because it is a hydrous phase and one will usually get less than 100% summation because we don't analyze for hydrogen (water and hydroxl)?
njdeems
Yes! That's exactly it.
qEd
I am not the right person to respond to this but in broad terms amphiboles have ~1.5 weight percent tied-up in hydroxyl, hence at 98.5% you may have a good analysis (and your cation values on the basis of 24 oxygens look reasonable as well). Others will hopefully give you more detailed guidance regarding limits.
Julien
Hi Nik,
Usually with hydrous phases, you need to recalculate the expected H2O content based on mineral formula (i.e., assuming a fixed atom per formula unit of OH-group). Such a mineral formula recalculation can be also helpful to check the site occupancy and possibly identify "bad" analysis (i.e. how much cation should be is site A, B, T... and compare with the ideal case). I'm sure you can find on the web or ask a colleague. I also have a website you can calculate mineral formula, and there is a specific entry for amphibole (http://cub.geoloweb.ch/index.php?page=mineral_formula&reset=true). You can also output your results in Probe for EPMA as atomic proportion and recalculate these values based on a fixed amount of cation (e.g. 15 cations for amphibole, with 22 oxygens and 2 OH group), but I don't think there is an option there to recalculate the H2O wt-%... I also attach a spreadsheet for amphibole mineral formula calculation. You should also check the literature for amphibole classification (e.g., papers from Hawthorne 2007 [Reviews in Mineralogy and Geochemistry 67, pp. 55-88] or Leake 1997 [American Mineralogist 86, pp. 1019-1037]), I think these papers contain useful info on the variation of amphibole composition, which could be of some help for your issue, along with some help about mineral formula calculation / classification of the amphibole.
Once you have your mineral formula recalculated and you have an estimate of H2O content, you can then check for outliers based on the classical assumption that a "good" analysis should be between 98 and 102%. If data are below 98%, something is wrong. However, the mineral formula recalculation can be tricky for these nasty "garbage" amphibole for numerous reasons. Here are some...
a) F and Cl can substitute for OH (did you measure them?). The amount of F and Cl will have a direct influence on how much H2O could be present. There can also be some O2- present instead of OH-...
b) Fe is both Fe2+ and Fe3+ in amphibole, and you cannot reliably measure the Fe2+/3+ proportion by EMP. The proportion of Fe2+ and Fe3+ will influence slightly your total as there is more oxygen associated to Fe3+...
c) vacancies exist in amphibole. The presence of vacancy is problematic essentially when checking the site occupancy.
d) Are you sure you have analyzed every single element in these amphiboles? Any missing element will induce a slightly lower total, and might screw up your mineral formula recalculation. However, if you did analyze all elements you can see on the EDS, then your total (with H2O recalculated) should be close to 98-100%. For minor / trace element you could do a long WDS scan. However, in most cases, the simple analysis of Si, Al, Fe, Ti, Mn, Mg, Ca, Na, K, F, Cl and possibly Cr, too, should be enough...
e) Totals might be too low due to, for instance, a bad standardization of some elements or the alteration of the amphibole (e.g. chloritization => more OH-group than normally expected...). A way to know if the problem could be sample or instrument / calibration related is to analyze a secondary standard. Any chance you have a Kakanui hornblende or another hornblende standard (reference material) of known composition? If so, try to analyze such a standard, and compare your analysis with the certified values...
Good luck!
Julien
mac
Julien's reply covers the most ground. Here is my two cents.
In practice, I would keep all analyses with total >96% for amphiboles when O is calculated, then check formula for each point base on O22(OH,Cl,F)2. Amphiboles contain various OH, up to 2.5 wt%? - check "Rock-forming Minerals" for info. Some also contain Li, which is not detectable by epma.
Amphiboles may show elemental zonation within single crystals. Do not eliminate points outside of 2sigma. Those points may be real. Do high-contrast BSE imaging using high current to check zonation.
Have fun!
Chi Ma
In case anyone is interested in amphibole calculations there is an option for site occupancy in amphiboles and biotites in PFE in the Calculation Options dialog as seen here:
These calculations are from Jay Ague and George Brimhall when Jay was a grad student at UC Berkeley. When this output option is run on say Kakanui Hornblende, one gets the following output:
St  468 Set   2 Hornblende (Kakanui) USNM 143965, Results in Oxide Weight Percents
ELEM:     Na2O    SiO2     K2O   Al2O3     MgO     FeO     CaO       S      Cl    TiO2    P2O5       F       O     H2O     MnO   SUM
   141   2.530  40.074   2.069  14.921  12.741  10.924  10.250    .010   -.014   4.796    .069    .005    .328    .938    .090  99.732
   142   2.647  40.163   2.049  14.768  12.885  10.856  10.207    .006   -.035   4.712   -.063    .012    .328    .938    .090  99.564
   143   2.576  40.171   2.080  14.916  12.717  10.844  10.327    .021    .017   4.763    .012    .029    .328    .938    .090  99.830
   144   2.642  40.068   2.111  14.872  12.903  10.823  10.211   -.018   -.026   4.681   -.061   -.013    .328    .938    .090  99.550
   145   2.558  40.279   2.004  14.977  12.863  10.983  10.222    .043   -.010   4.640   -.076   -.012    .328    .938    .090  99.827
AVER:    2.590  40.151   2.063  14.891  12.822  10.886  10.243    .012   -.014   4.719   -.024    .004    .328    .938    .090  99.700
SDEV:     .052    .086    .040    .078    .086    .066    .050    .022    .020    .063    .062    .017    .000    .000    .000    .137
SERR:     .023    .039    .018    .035    .039    .030    .022    .010    .009    .028    .028    .008    .000    .000    .000
%RSD:     2.01     .21    1.93     .52     .67     .61     .48  178.26 -144.96    1.33 -263.26  414.50     .00     .00     .00
PUBL:    2.600  40.372   2.050  14.900  12.800  10.920  10.300    .000    .000   4.721    .000    .000    .328    .938    .090 100.020
%VAR:     -.38    -.55     .61    -.07     .17    -.31    -.55     .00     .00    -.04     .00     .00     .00     .00     .00
DIFF:    -.010   -.221    .012   -.010    .021   -.034   -.056    .000    .000   -.002    .000    .000    .000    .000    .000
STDS:      336     162     374     160     162     162     162     730     285      22     285     284       0       0       0
Amphibole Formula Calculations (from Jay Ague AMPHI.F code)...
       ALL FE2  NORM 1  NORM 2  NORM 3  NORM 4  NORM 5
    Si  5.8886  5.9916  5.8292  5.7065  5.5664  5.8646
    Ti   .5205   .5296   .5152   .5044   .4920   .5183
    Al  2.5744  2.6194  2.5484  2.4948  2.4336  2.5639
  Fe3+   .0000  -.8040   .4644  1.4234  2.5172   .1883
  Fe2+  1.3352 -1.0000   .8573 -1.0000 -1.0000  1.1415
    Mg  2.8025  2.8515  2.7742  2.7158  2.6492  2.7911
    Mn   .0112   .0114   .0111   .0109   .0106   .0112
    Ca  1.6097  1.6379  1.5935  1.5599  1.5217  1.6032
    Na   .7366   .7495   .7292   .7139   .6963   .7336
     K   .3859   .3927   .3821   .3740   .3648   .3844
     F   .0019   .0020   .0019   .0019   .0018   .0019
    Cl  -.0034  -.0034  -.0033  -.0033  -.0032  -.0034
      NORM 1: TOTAL-(NA+K)=15   NORM 2: TOTAL-(NA+CA+K)=13
      NORM 3: TOTAL-K=15        NORM 4: SI+AL=8.0
      NORM 5: TOTAL=15.8
      STRUCTURAL FORMULA:
        SI     5.8589
        ALIV   2.1411        ALVI   .4203
        TI      .5178
        FE3+    .2322
        FE2+   1.0963
        MG     2.7884
        MN      .0112
        CA     1.6016
        NA      .3322        NA     .4007
        K       .3840
        F       .0019
        CL     -.0034
        OH     2.0015
MOLE FRACTIONS AND LOGARITHMS OF ATOMIC RATIOS:
      X-FE2+=     .157      X-MG=       .398
      X-FE3+=     .033      X-ALVI=     .060
      X-MN=      .0016      X-TI=      .0740
      X-CA=       .229      X-NAM4=    .0475
      MG / (MG + FE2+) =      .718
      FE2+/(FE2+ + FE3+)=     .825
      MG/(MG+FE2+ + FE3+)=    .677
      X-OH=                        1.001    X-F=                   .001    X-CL=            -.002
      LOG(X-MG/X-FE2+)=             .405    LOG(X-F/X-CL)=         .000    LOG(X-F/X-OH)=  -3.023
      LOG(X-MG/(X-FE2+ + FE3+))=    .322    LOG(X-FE2+/X-FE3+)=    .674
      A-SITE=                       .785    TOTAL VI=             7.000
      ALVI+2TI+A-SITE+FE3+=        2.473    ALVI+2TI+FE3+=        1.688
      Schmidt (1992) Pressure (All FE2+):     9.03 KBar,     (FE2+ -FE3+):     8.96 KBar
The output is also saved to a text file in your user data folder (containing your MDB probe database), called AMPHI.OUT in a "machine readable" format as seen here:
Sample "467"
41.6150    6.05719    5.93890   Si
1.39705    .152929    .149900   Ti
15.4665    2.65373    2.60190    2.06110    .540800   Al    9.42825    9.16756
.000000    .000000    .898900   Fe+3
11.4627    1.39535    .469200   Fe+2
14.2514    3.09144    3.03110   Mg
.149784    .018467    .018100   Mn
11.5703    1.80451    1.76930   Ca
1.90354    .537216    .526700    .122700    .404000   Na
.219696    .040797    .040000   K
-.00216    -.00099    -.00100   F
-.01291    -.00319    -.00310    2.00410   Cl
-.49136    .810242    .865954    .067029    .021414    .077257    .002586    .000000    2.18350    1.73950    2.60190    .444000    .342957    5.93890
.122700    .404000    7.00000    1.76930          0
Sample "468"
40.1509    5.88858    5.85890   Si
4.71855    .520451    .517800   Ti
14.8907    2.57440    2.56140    2.14110    .420300   Al    9.02922    8.96384
.000000    .000000    .232200   Fe+3
10.8859    1.33522    1.09630   Fe+2
12.8218    2.80249    2.78840   Mg
.090387    .011229    .011200   Mn
10.2433    1.60972    1.60160   Ca
2.59042    .736635    .732900    .332200    .400700   Na
2.06269    .385949    .384000   K
.004190    .001943    .001900   F
-.01359    -.00338    -.00340    2.00150   Cl
.000000    .405426    .717790    .156614    .073971    .060043    .001600    -3.0226    2.47280    1.68810    2.56140    .784700    .825216    5.85890
.332200    .400700    7.00000    1.60160          0
If anyone needs the source code for these calculations, please let me know.