Does Cancel "work" during PHA scan (JEOL vs. Cameca)
Started by UofO EPMA Lab, December 29, 2013, 01:22:25 PM
UofO EPMA Lab
John Fournelle reports that Philippe Pinard seems to not be able to cancel a PHA scan once it is started on his 8530.
We tested this today on my Sx100 and it seems to cancel properly. Below is the debug mode text of running a PHA scan *prior* to the peaking from the Acquire! window, and canceling after a few points have been acquired.
Note this behavior maybe quite different once I implement the new MCA (multi-channel-analyzer) methods for PHA scanning.
Inserting faraday cup or blanking beam
Faraday cup inserted or beam blanked
Setting operating conditions (TKCS):  40 15 30 5
Setting operating conditions (TKCS):  40 15 30 5, getting current conditions...
SX100SetConditions: current conditions (TKCS) =  40 15 29.9707 5
SX100SetConditions: static variables (TKCS) =  40 15 30 5
SX100SetConditions: ForceAnalyticalConditions =  0
SX100SetConditions: OperatingVoltageTolerance =  .002, BeamCurrentTolerance =  .01
SX100SetConditions: beam current error = 29.9707 / 30 =  -.09767 %
Operating conditions set (TKCS)
Setting beam mode (Analog  Spot)...
Setting magnification:  20000
Setting magnification to  20000...
Magnification set to 20000
Removing faraday cup or unblanking beam
Faraday cup removed or beam unblanked
Getting spectro  5 PHA (BWGBMD):  1.36 .2 700 1812-1 0
Setting spectro  5 PHA (BWGBMD):  .56 4 700 1812 -1 2.97E-06
ti ka Set Column,
ti ka Set Column,
Starting spectrometer peaking procedure for ti ka on spectro 5...
ti ka Move On Peak,
ti ka Move On Peak,
ti ka Init Prior PHA,
Point:  1, Baseline:  .56, window:  .2
Setting spectro  5 PHA (BWGBMD):  .56 .2 700 1812 -1 2.97E-06
Start count spectro  5, counttime  .5, maxcounts  100000000
ti ka Wait Prior PHA,
Point:  2, Baseline:  .76, window:  .2
Setting spectro  5 PHA (BWGBMD):  .76 .2 700 1812 -1 2.97E-06
Point:  3, Baseline:  .96, window:  .2
Setting spectro  5 PHA (BWGBMD):  .96 .2 700 1812 -1 2.97E-06
Point:  4, Baseline:  1.16, window:  .2
Setting spectro  5 PHA (BWGBMD):  1.16 .2 700 1812 -1 2.97E-06
Point:  5, Baseline:  1.36, window:  .2
Setting spectro  5 PHA (BWGBMD):  1.36 .2 700 1812 -1 2.97E-06
Cancel: Ready        <----- cancel button clicked in main log window
Setting magnification:  2524
Setting magnification to  2524...
Magnification set to 2524
All motors stopped
All counters stopped
Acquire: Ready
UofO MicroAnalytical Facility
UofO EPMA Lab
Here's a bias also canceled properly during acquisition (again on the Cameca SX100):
SX100SetConditions: current conditions (TKCS) =  40 15 29.9691 5
SX100SetConditions: beam current error = 29.9691 / 30 =  -.10300 %
ti ka Set Column,
ti ka Set Column,
ti ka Move On Peak,
ti ka Move On Peak,
Point:  1, Bias:  1100
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1100 -1 2.97E-06
Point:  2, Bias:  1120
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1120 -1 2.97E-06
Point:  3, Bias:  1140
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1140 -1 2.97E-06
Point:  4, Bias:  1160
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1160 -1 2.97E-06
Point:  5, Bias:  1180
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1180 -1 2.97E-06
Point:  6, Bias:  1200
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1200 -1 2.97E-06
Point:  7, Bias:  1220
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1220 -1 2.97E-06
Point:  8, Bias:  1240
Setting spectro  5 PHA (BWGBMD):  2 .1 700 1240 -1 2.97E-06
Cancel: Ready         <--------- cancel button clicked here
All motors stopped
Acquire: Ready