News:

:) You cannot see "members only" boards if you are not a member, so please join the forum!

Main Menu

An Open Letter to the Microanalysis Community

Started by Probeman, November 18, 2021, 11:36:16 AM

Previous topic - Next topic

crystalgrower

#15
FIRM WRITTEN price quotes are the first step.

Note added next day:  Also required would be the minimum order whether in kilos or $$$.

The "cost to synthesize" does not include the cost of raw materials.  Hf-free Zr will cost a lot more than natural Zr.  Then  there is the cost of making ZrO2 powder, then the recrystallizatio

ZrO2 is usually stabilized with Y2O3: better make sure your spec is precise.  You may not have an option if the transition from low symmetry to cubic is below the melting point.  Charles Taylor had natural ZrO2 with 0, 5%, 10%, 15% and 20% Y2O3.  Below 15% Y2O3 the pieces were crappy polycrystalline dust.   But who knows where all that got to?

One  reason to avoid LaB6, CeB6 etc is the extremely high melting point.  Materials are usually sintered--pressed in a furnace but not melted.   Commercial RSi2 is a mix of R rich and Si rich phases , each phase about 1 micron across.



Probeman

#16
Excellent advice.

Would you be able to help us locate some high purity ZrO2? And also perhaps some single crystal borides?  We only need a few grams to start our initial testing of purity and stoichiometry. Later we will want quotations for 500 to 1000 grams.

If you know of specific sources we would love to hear from you. This is a crowd sourced project, if we don't do it, no one else will.

Please send any written quotes to Will Nachlas:

Will Nachlas
Weeks Hall for Geological Sciences
1215 West Dayton St
Madison WI 53706

Will Nachlas <nachlas@wisc.edu>

or add the quotation information to our Google spreadsheet:

https://docs.google.com/spreadsheets/d/19AeXvxIaP6qvChbE7cxK05B_6rkZSN14T-7ZJ7nkm8M/edit*gid=0

We need all the help we can get, thank-you!
The only stupid question is the one not asked!

wonachlas

#17
I have a lead on potential sources for ZrO2 and YAG (Y3Al5O12). Does anyone have experience with materials from the Shelby Gem Factory? They were a renowned synthesizer of several gemstones and apparently invented a technique for synthesis of CZ (ZrO2) using the Czochralski method. The factory closed in 2019 and all of the remaining stock was sold in bulk. I met a local gem dealer who claims to have obtained "hundreds" of pounds of bulk gemstones from the owner when it closed. I purchased ~7 g pieces of colorless ZrO2 and YAG. I will mount and test them for crystallinity and trace impurities to evaluate if materials from the Shelby factory could be suitable for our purposes.

sem-geologist

Quote from: crystalgrower on November 26, 2021, 02:11:10 PM
One  reason to avoid LaB6, CeB6 etc is the extremely high melting point.  Materials are usually sintered--pressed in a furnace but not melted.   Commercial RSi2 is a mix of R rich and Si rich phases , each phase about 1 micron across.

The high melting point was actually my main Pros to propose these REEB6, as that would be outstandingly stable under the electron beam. However, after doing more research on hexaboron structure, I came to see a much bigger problem for this material to be considered as standards.
I was trying to understand what oxidation state Ce inside CeB6 would be presented (generally REEB6, with REE as +2/+3/+4 and B +3/-5, looks not straight forward to understand). The structure of REEB6 can be described as B cage (B octhahedras connected in cubic fashion providing cells in between for REE) with entrapped REE in formed cells. Ideally stochiometric LaB6 (deep violet) is then all cells are occupied. But there can be vacancies, some sources suggest that in some extreme cases it can be 1/4 of cells not occupied by REE. Generally, it is easy to get slightly out of stochiometry values, and it is not so straight forward to check how much vacancies are there, without relying on other methods. The LaB6 which I have in one of SPI sets, and which looks very clean and homogenos, is clearly out from ideal stochiometry/has vacancies (phew, I at last have an explanation why it would give so systematically slightly different result compared with LaPO4 as standard). So REE-hexaborons are out from considerations. (They would still be excellent beam-shape standards for low voltage EDS, but should not be used as primary standards)

Probeman

#19
Quote from: Probeman on November 22, 2021, 12:32:26 PM
...dead time calibrations are probably a major area of inaccuracy. Paul Carpenter has tried to point this out for decades. If anyone out there is still using the "default" dead time calibrations provided to them at the time their instrument was installed, I can promise them that they have large accuracy problems. These gas detectors age quite dramatically over time and one should be re-running these dead time calibrations every year to two at most for reasonable accuracy.  This issue has actually gotten worse over time as both Cameca and JEOL have migrated to larger and larger Bragg crystals with increasing geometric efficiency.

So unless you always run a 10 nA or less, here is a link to Paul Carpenter's dead time spreadsheet which is very nice for performing your own dead time constant calculations using any software:

https://smf.probesoftware.com/index.php?topic=1160.0

It should be emphasized that utilizing standards that are not exactly matrix matched, will result in accuracy errors if ones instrument is not properly calibrated. Paul Carpenter and John Fournelle and been consistently making this point for decades, and before them it was Dan Weill and John Armstrong...  when will we ever learn?

To make this point clearer, I am attaching an abstract and presentation by Paul and John (but not George and Ringo!), from about 15 years ago.

Unfortunately I don't think things have gotten any better on this front since then.  The good news is that the synthetic high accuracy high purity minerals that are being proposed in the topic will help reveal to what extent our instrument require proper re-calibration of detector dead times, PHA settings, effective takeoff, etc., etc.
The only stupid question is the one not asked!

qEd

One possibility for evaluating the synthetic crystals for stoichiometry under consideration for this endeavor is X-ray synchrotron diffraction imaging. I have never performed such a measurement, but imagine there are a number of beam lines that could be used to determine the distribution of defects/inclusions in a multi-mm cm?-sized crystal to verify the structure on a spatially resolved basis. 

crystalgrower

I will not comment on the ZrO2 because it belongs in a separate topic.

There is absolutely no reason for individuals to go and get price quotes because they will not be valid for the body that would be making the purchase according to the plan in the first post. 

Individuals are NOT representatives of the money holding body for the purpose of signing contracts for custom syntheses or purchases.

In any case the price quotes will probably expire  before any formal agreement to spend funds will be reached.

Lab managers who purchase supplies and sometimes instruments should be familiar with the process of funds being approved.  I don;t know how any professional could get an idea that crowdsourcing tactics are going to collect  any valid information towards  a corporate purchase. 

And since you haven't been placing orders for custom materials, please be aware that a sample is not released  until a contract is signed.  A significant part of the total price is required to be paid before ANY work of synthesis begins.


Probeman

#22
Quote from: crystalgrower on December 13, 2021, 12:51:53 PM
I will not comment on the ZrO2 because it belongs in a separate topic.

There is absolutely no reason for individuals to go and get price quotes because they will not be valid for the body that would be making the purchase according to the plan in the first post. 

Individuals are NOT representatives of the money holding body for the purpose of signing contracts for custom syntheses or purchases.

In any case the price quotes will probably expire  before any formal agreement to spend funds will be reached.

Lab managers who purchase supplies and sometimes instruments should be familiar with the process of funds being approved.  I don;t know how any professional could get an idea that crowdsourcing tactics are going to collect  any valid information towards  a corporate purchase. 

And since you haven't been placing orders for custom materials, please be aware that a sample is not released  until a contract is signed.  A significant part of the total price is required to be paid before ANY work of synthesis begins.

I adamantly disagree. The vendor information, availability and (rough) pricing is still valuable information for eventual group purchases.  Remember, we need to purchase small amounts initially to test for purity and stoichiometry.

Even more to the point, we are asking our community to investigate *all* possible sources for these materials, not only commercial sources but also state/government laboratories that may be able to offer our community such materials for *free*.  If we do the leg work.  You can decide for yourself if you would like to join this effort (or not).  :)
The only stupid question is the one not asked!

Probeman

#23
I am attaching below (login to see attachments) documents from the global synthetic microanalysis standard project (unofficial):

1. Action plan

These are just draft ideas for how to proceed with this project. Will Nachlas is already moving ahead on a number of these steps and I hope you will participate. Check with the protocols he has specified including making careful instrument calibrations prior to any measurements, especially dead time calibrations on your WDS spectrometers.

2. Resources

Possible resources for obtaining high purity synthetic minerals, though we need your help to find/locate more materials and obtain rough pricing.

3. George Rossman's personal synthetic crystal collection (Cal Tech).

This is a spreadsheet from George Rossman showing what synthetic minerals have been produced in the past, as a guide to what *might* be possible to produce on larger scales for use as global microanalysis standards.
The only stupid question is the one not asked!

crystalgrower

Under the original heading of sourcing  high quality materials:
I  learned a very different way of business with ACS and Mineralogical Society of America.

US Professional associations can incorporate as nonprofits for  their educational activities.  I have not checked to see if either MAS or Microscopy  Society of America have done this.
The benefit of having approval from a nonprofit to solicit donations of materials is that you may be able to solicit a much larger donation in return for a tax receipt.  Which tax receipt can only be issued by the financial officer of the association and their 503(b) registration number must appear. 
In fact a custom synthesis will never be donated without  100% of fair market  value including the cost of starting materials becoming a tax benefit.

Good Biz 101 is  not difficult or obscure. 
I have a specific concern.  The drive to  lowball legitimate businesses started with the forum survey asking users what they would like to pay as opposed to teaching them what it costs to make specialty pure materials.  Now users are being asked to carry out fishing   expeditions.   Should many users  falsely state they represent the MAS then the blowback might be worse than just denial of funds. 

The real intent of this Open Letter may be  to acquire some piles of materials not able to be sold for other reasons.   A pitch to ask for business leftovers  should have been stated up front, rather than all the noise about money possibly  sitting in the bank accounts of nonprofits. 
FYI: all those  pieces that SPI advertises are owned by Astimex Standards registered in Canada (no US tax benefit for donation) and the physical location might be a problem.

Users had better think very carefully where lowlballing small industry might lead.  About 10  years ago I found a small US company with SBIR grand funds, whose business was hydrothermal-grown  crystals of pure and layered materials.  They did not hand out free samples when asked.  They would have been the ideal place to commission Pb-free replacements for RPO4 of R=La-Gd (since all crucible fluxed methods leave contamination).  That commission could then have been donated to the NMNH as Lynn Boatner did with the original TVA materials.  But they  closed sometime before 2017. 

Probeman

Quote from: crystalgrower on December 17, 2021, 11:41:22 AM
The real intent of this Open Letter may be  to acquire some piles of materials not able to be sold for other reasons.   A pitch to ask for business leftovers  should have been stated up front, rather than all the noise about money possibly  sitting in the bank accounts of nonprofits. 

Asking for commercial quotations, seeking state/government sources, asking for "leftovers" and custom synthesis are all activities that can be pursued in parallel, they are not exclusive pursuits.

Nachlas et al. are currently writing grant proposals, and matching funds may be required from "money in bank accounts of non profits", though exactly how that will be leveraged is up to those society directors.
The only stupid question is the one not asked!

Probeman

#26
There has been some discussion in this topic on whether we really require so called "matrix-matched" primary standards for high accuracy EPMA. That is, do our primary standards really need to be similar in composition (and also valence and coordination), to our unknown materials?

So aside from the questions regarding the accuracy of our compositional matrix correction physics, it's a valid question since as we know from multiple studies that in the case of light elements at least, we experience peak shift and shape effects that can result in accuracy problems when not utilizing integrated area scan acquisitions for elements such as oxygen, nitrogen, carbon, boron, etc. Even sulfur k-alpha can have a significant peak position shift (though not a shape change) depending on the oxidation state of the sulfur:

https://smf.probesoftware.com/index.php?topic=127.0

So, even if our compositional matrix corrections were perfect, we would still need to ascertain the magnitude of the peak shift and shape effects due to chemical states. And I think it's still an open question whether we can accurately extrapolate from one material to another when the element (mission transitioning from the valence shell) in our primary standard is in a difference valance state and/or coordination than that of our unknown.

And that is one reason why we selected as our first test sample for the high purity synthetic standard round robin three materials: MgO, Al2O3 and spinel (MgAl2O5). The idea being that these pairs (Mg Ka in MgO to MgAl2O3 and Al Ka in Al2O3 to MgAl2O5), are not only very different in composition, but also somewhat different in their chemical states.  And those materials were readily available as high purity synthetics!  :)

We are still awaiting the results from this first round robin, but I decided to share another test of this concern, that is Si Ka in SiO2 compared to some common silicates.  This was a test I ran recently looking further into the problem of measuring trace Sr and Rb in silicates, but let's ignore those trace results for now and focus on the Si and Al major elements. Unfortunately I didn't have an Al2O3 standard in the standard mount (and wasn't running this test for the Si and Al concentrations as they were only being measured for the interference corrections), but still the Si data might be helpful regarding these major elements accuracy issues.

So using SiO2 as the primary standard for Si (and nepheline as the primary standard for Al), we obtain these results for labradorite:

ELEM:      Sr      Rb      Si      Al      Ca      Na      K      Fe      Mg      O  SUM  
  1379    .055  -.013  24.495  16.458  9.577  2.841    .100    .319    .084  46.823 100.739
  1380    .051  -.001  24.060  16.495  9.577  2.841    .100    .319    .084  46.823 100.350
  1381    .061  -.004  24.038  16.513  9.577  2.841    .100    .319    .084  46.823 100.351
  1382    .070  -.016  23.908  16.544  9.577  2.841    .100    .319    .084  46.823 100.250

AVER:     .059  -.008  24.125  16.502  9.577  2.841    .100    .319    .084  46.823 100.423
SDEV:     .008   .007    .256    .036   .000   .000    .000    .000    .000    .000    .216
SERR:     .004   .003    .128    .018   .000   .000    .000    .000    .000    .000
%RSD:    13.91 -82.23    1.06     .22    .00    .00     .00     .00     .00     .00

PUBL:    n.a.    n.a.  23.957  16.359  9.577  2.841    .100    .319    .084  46.823 100.060
%VAR:     ---     ---     .70     .88    .00    .00     .00     .00     .00     .00
DIFF:     ---     ---    .168    .143   .000    .000   .000    .000    .000    .000
STDS:     251    1023      14     336    ---     ---     ---     ---    ---    ---

So well within 1% relative accuracy on both Si and Al. Now for the nepheline (just looking at Si because this is the primary standard for Al):

ELEM:      Sr      Rb      Si      Al      Na      K      Fe      O      Ca  SUM  
  1383    .009    .030  20.553  17.872  12.552  4.657    .155  44.418    .075 100.322
  1384   -.002    .045  19.924  17.774  12.552  4.657    .155  44.418    .075  99.598
  1385    .006    .029  20.594  17.954  12.552  4.657    .155  44.418    .075 100.440
  1386    .002    .025  20.422  17.857  12.552  4.657    .155  44.418    .075 100.163

AVER:     .004    .032  20.373  17.864  12.552  4.657    .155  44.418    .075 100.131
SDEV:     .005    .009    .308    .074    .000   .000    .000    .000    .000    .373
SERR:     .002    .004    .154    .037    .000   .000    .000    .000    .000
%RSD:   134.64   27.41    1.51    .41     .00     .00     .00     .00     .00

PUBL:     n.a.    n.a.  20.329  17.868  12.552  4.657    .155  44.418    .075 100.054
%VAR:      ---     ---     .22  (-.02)     .00    .00     .00     .00     .00
DIFF:      ---     ---    .044   (.00)    .000   .000    .000    .000    .000
STDS:      251    1023      14     336     ---     ---     ---     ---    ---

Again excellent accuracy extrapolating from SiO2. Now our orthoclase standard:

ELEM:      Sr      Rb      Si      Al      Fe      K      Na      Ba      O  SUM  
  1387    .005    .120  29.905  8.844  1.461  12.859    .675    .054  45.798  99.721
  1388   -.011    .121  29.736  8.860  1.461  12.859    .675    .054  45.798  99.553
  1389   -.001    .087  30.128  8.792  1.461  12.859    .675    .054  45.798  99.853
  1390   -.010    .117  30.202  8.773  1.461  12.859    .675    .054  45.798  99.929

AVER:    -.004    .111  29.993  8.817  1.461  12.859    .675    .054  45.798  99.764
SDEV:     .008    .016    .212   .041   .000    .000    .000    .000    .000    .165
SERR:     .004    .008    .106   .021   .000    .000    .000    .000    .000
%RSD:  -172.78   14.61     .71    .47    .00     .00     .00     .00     .00

PUBL:     n.a.    .027  30.286  8.849  1.461  12.859    .675    .054  45.798 100.009
%VAR:      ---  311.36    -.97   -.36    .00     .00     .00     .00     .00
DIFF:      ---    .084  -.293  -.032   .000    .000    .000    .000    .000
STDS:      251    1023      14   336     ---     ---     ---     ---     ---

Again within 1% relative accuracy for both.

In other prior work I've seen similar accuracy extrapolating from MgO to other Mg silicates, so I do believe these extrapolations are feasible, though we will see in the case of Al Ka since we already know from work by Fournelle that there are subtle Al peak position shifts in feldspars at least.

By the way, these measurements were performed at 50 nA because the purpose was to look at the trace elements, but the beam was defocused to 10 um to minimize TDI effects. Never the less, some intensity changes over time were observed as shown below, but only for the Si Ka emissions!



The above being a normal exponential TDI fit. A hyper-exponential fit might be worth trying even though it appears to overfit, the intercepts are probably more accurate and that's what we utilized here:



This resulted in a TDI correction of around 1.7% +/- 0.2 for Si Ka and 0.34% +/- 0.1 for Al Ka in the labradorite

Anyway, bottom line is that major element matrix correction extrapolations can be quite accurate even when extrapolating from a pure oxide to a silicate mineral, and the valence and coordination effects seem to be minimal.
The only stupid question is the one not asked!

Probeman

#27
I just wanted to address a question on the purity required for synthetic standards that we utilize as primary standards for major element analysis.  In other words, how pure do these synthetic mineral standards really need to be?

My answer would be that I think we should be initially focused on major element standards. The main reason being to guide analysts away from the seduction of so called "matrix matched" standards of questionable accuracy (and availability).  Instead what we need are high accuracy (and high availability) major element standards, e.g., MgO, Mg2SiO4, MgAl2O4, SiO2, Fe2SiO4, Fe3O4, etc., etc.

Trace element (doped) standards are really not necessary in my opinion. The only thing we need for trace element accuracy are pure metals or simple oxides for primary standards (and their accuracy is not all that important!). What's important for trace elements is the background modeling! And optionally also high purity blank (possibly matrix matched) materials for the application of a blank correction.  Doped trace standards are for dopes!  😁

Back to major elements, our preliminary MgO, Al2O3, MgAl2O4 FIGMAS data reported by Will shows that with accurate dead time calibrations we can obtain better than 1 to 2% accuracy extrapolating from simple compounds to significantly different compositions ( I obtained similar results myself with the test mount).  And also my own long term experience with other high concentration primary standards for other elements, and a recent test using SiO2 as a primary standard here:

https://smf.probesoftware.com/index.php?topic=1415.msg10574#msg10574

shows similar sub 1% accuracy.  The good news is that our modern matrix corrections are pretty damn good, the weak point being the dead time calibrations on our instruments.  And those can be easily fixed in a couple of hours work:

https://smf.probesoftware.com/index.php?topic=1160.0

So our concerns regarding trace elements in these synthetic materials I think should only be relative to the extent to which these traces affect the accuracy of the theoretical (major element) stoichiometry of these synthetic compositions.  So to me that means 99.99% purity is just fine. A 100 ppm variance on our major elements is not a significant concern.
The only stupid question is the one not asked!

Probeman

Attached below is the latest draft of the action plan of global standards (login to see attachments).

If anyone has additional information (e.g., additional suppliers of high purity synthetic single crystals), just send me the information and I will add it to the document.

Thanks!
The only stupid question is the one not asked!

Probeman

#29
I have to say I am very much enjoying the book "Beyond Measure":

https://www.amazon.com/Beyond-Measure-History-Measurement-Constants/dp/1324035854

In the section on the effort to redefine the kilogram, they mention the concept of standards "For All Times, For All Peoples":

https://www.nist.gov/blogs/taking-measure/all-times-all-peoples-how-replacing-kilogram-empowers-industry

Although our standards (as far as I can imagine) will always be physical objects, we can definitely place them on a basis more akin to the "For All Times, For All Peoples" concept.

For example, instead of defining an olivine standard as a crystal that came from a specific outcrop (on a specific date and collected by a specific person), with all it's natural heterogeneity, inclusions and limited availability, we can instead define our olivine standard as a material that contains 2 atoms of Mg, 1 atom of Si and 4 atoms of O.

In other words single crystal Mg2SiO4, which is available commercially in essentially unlimited quantities, with high purity and can therefore be distributed globally and re-sourced and/or reproduced easily in the future.  Unlike every naturally sourced material we currently have in our collections!

Remember also that such a standard can not only be utilized as a highly accurate primary standard for Mg, but also as an accurate blank standard for any trace element, in order to test ones ability to measure zero in an olivine matrix:

https://smf.probesoftware.com/index.php?topic=454.msg11753#msg11753

Now it's true that synthetic high purity Fe2SiO4 is not available commercially that I know of, though I'm sure further efforts could be made to synthesize it (as it has been successfully synthesized at Oak Ridge National Laboratory by Lynn Boatner because I have a nice piece of it!), though I suspect we would also have excellent results using almost any other synthetic Fe oxides (e.g., Fe3O4) for the analysis of Fe in olivines.

Is anyone willing to search their closets and attics for a possible source for such previously grown materials?  I'm looking at you national lab people!

Quote
Treasures in your attic (or more likely your lab cabinets)!

The authors of this open letter believe that one possible source of such stoichiometric materials could come from past efforts by our colleagues in the crystal growth community to create large crystals of stoichiometric compounds. Such efforts might have resulted in archiving some of these materials, perhaps forgotten at the back of a lab cabinet, that could now be repurposed to serve as ideal standards for the microanalysis community. Ideally we are seeking quantities in the 500 to 1000 gram or more quantities, for example:

Mg2SiO4, YAG, RbTiOPO4, KTiPO4, MnO, Fe3O4, NiO, ZnO, LaAlO3, MnPSe3, LiTaO3, ZrSiO4 (zircon), ZrO2 (zirconia), HfSiO4 (hafnon), HfO2 (hafnia), ThSiO4 (tetragonal thorite), ThSiO4 (monoclinic huttonite), Fe2SiO4 (fayalite), Mn2SiO4 (tephroite), CaMgSi2O6 (diopside), Al2SiO5 (sillimanite), NaAlSiO4 (nepheline), KAlSi3O8 (sanidine), KAlSi2O6 (leucite), KAlSi3O8 (orthoclase), NaAlSi3O8 (albite), CaAl2Si2O8 (anorthite), Fe3Al2Si3O12 (almandine), PbSiO3 (alamosite), CaAl2O4 (krotite), CaAl4O7 (grossite), CaAl12O19 (hibonite), CaSiO3 (wollastonite), MgSiO3 (enstatite), FeSiO3 (ferrosilite), sulphides and sulfosalts, etc., etc.

Do you have access to such materials in your laboratory, or know of materials located elsewhere, that you would be willing to share with your colleagues worldwide in this important endeavor? If so, we would welcome your participation and ask that you contact the MAS FIGMAS on Standards with any information that can advance our "quest for fire"!

a plea from Dale Newbury
The only stupid question is the one not asked!