Blank Values Outside Min/Max Analytical Total?

Started by Scott B., January 04, 2024, 03:30:41 PM

Previous topic - Next topic

Scott B.

Greetings,

In the CalcImage "Specify Quantitative Parameters" dialogue for setting up a quant map calculation, there's this checkbox.



Where does one input their preferred values for this feature?

Cheers,

Scott B.

John Donovan

#1
The analytical min/max totals are defined in the [software] section of the probewin.ini file as shown here:

lpAppName$ = "Software"
lpKeyName$ = "AnalyticalTotalMinimum"
lpDefault$ = "95"

lpAppName$ = "Software"
lpKeyName$ = "AnalyticalTotalMaximum"
lpDefault$ = "105"

We will also add a "tool tip" popup text to help in the future.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

In the [software] section of the Probewin.ini file (usually found in the C:\ProgramData\Probe Software\Probe for EPMA folder), the keywords are:

AnalyticalTotalMinimum="95"
AnalyticalTotalMaximum="105"

The AnalyticalTotalMinimum value must be greater than 50 and the AnalyticalTotalMaximum value must be less than 150.