PHA - Integral vs Differential
Started by BenjaminWade, February 20, 2014, 02:25:29 PM
Hi all
I am wanting to draw on everyones wisdom again, and apologies if this has been discussed before. I have used the search function and couldn't find anything on it.
I was just interested in what peoples philosophy is for the two different modes. I know some purely use integral for everything and rely on interference corrections to get rid of their higher order overlaps, and some use a combination of both integral and differential. I am about to set up a monazite method again (using Julien Allaz great pdf). In the past we used differential mode for a lot of the elements in our monazite method, and was wondering if this is still the case for the most part?
There is generally little downside to using differential mode as long as you leave the baseline and window "loose".  On the Cameca instrument that would be 0.5 for the baseline and 4 or so for the window (9 or so for JEOL instruments).
The only advantage of differential mode is a partial reduction of higher order interferences where an interference standard is difficult to obtain, for example 2nd order sodium K-alpha on oxygen, because it's hard to find a standard that contains Na but no oxygen.
If you are analyzing materials with a large difference in count rates, one may need to utilize integral mode but be sure that the PHA peak is fully above the baseline level at the highest count rate you will be observing.