News:

:) If you are a member, please feel free to add your website URL to your forum profile

Main Menu

Testing new multi-channel analyser PHA capability for Cameca SX

Started by Karsten Goemann, March 23, 2014, 09:24:33 PM

Previous topic - Next topic

Karsten Goemann

Hi John,

I tested the new Cameca PHA MCA feature a bit (http://smf.probesoftware.com/index.php?topic=42.msg883#msg883).

It's great, 1 second PHA scans!

However, I noticed that when PHA differential is specified, the x axis range of the PHA window reflects that PHA window, but the PHA scan data is actually for the full window (.56 to 5.5V) so the x axis range is wrong, see attached two images for S Ka on PET, identical conditions but PHA diff switched on for the first one (base .75, window 2.09V) and PHA integral for the second one.

Cheers,

Karsten

John Donovan

Ok, I see this.

Interesting that I got it to work correctly in demo mode!

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

Hi Karsten,
OK, I got it working.  But here's something I don't understand...

So here is a MCA PHA acquired using the new MCA PHA hardware:



It looks just like the display on the PeakSight software. And here is the same x-ray line acquired using the traditional PHA scanning method:



Why are they different?  Where is the Ar escape peak in the MCA PHA acquisition?

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Karsten Goemann

Not sure.

I do normally see the escape peak using PeakSight's MCA and usually it has more separation between the main and the escape peak, but that's obviously energy dependent.

Which x-ray line did you use for that PHA scan?

Gseward

here is an MCA PHA scan and a traditional PHA scan of the Fe ka peak from my SX100.





the peak position appears different, Is the x-axis wrong for one of the plots?

Gseward

Testing the new MCA PHA acquisition I increased the time to 1s with 40 intervals, I get this error:

PHA MCA Test on spectro  1 timed out. This is usually caused by too low an x-ray intensity. Please try the PHA MCA Test again using a material with a higher concentration or uncheck the Use MCA PHA Acquisition checkbox in Acquisition Options dialog and instead utilize the traditional PHA acquisition method.

I did a bit more messing around and discovered that if:

count time * intervals >= ~30

the error occurs. count time * intervals < 30  and all is good.

Gareth

John Donovan

Quote from: Karsten Goemann on March 25, 2014, 08:32:52 PM
Not sure.

I do normally see the escape peak using PeakSight's MCA and usually it has more separation between the main and the escape peak, but that's obviously energy dependent.

Which x-ray line did you use for that PHA scan?
Fe Ka.

Gareth's Fe scan looks more like my traditional one!

I wonder if these MCA PHA electronics ever need to get recalibrated?
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Gseward

John,

If I move to lower gain so that the peak is at a similar V to yours, then I see similar results to you, I don't know why yet either:

MCA


Traditional


BUT note again that the peak has an apparent shift - I have a feeling that the Cameca MCA data is returned as 0-5V not 0.5-5.5V as it is being plotted by PFE.

Gareth

John Donovan

That makes some sense.  I was just going by the PeakSight data display which I thought shows a blue line for the PHA data going from 0.5 to 5.5 and yes:



but I have a pretty old version of PeakSight. What do you see?
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Gseward

Well, what about that - Mine shows 0-5V!



p.s. I have no idea how I managed to turn the background black, and I can't figure out how to make it white again!

EDIT by GSeward: this plot has no smoothing applied

Probeman

You know what this means?

We can't demonstrate the escape peak any longer using this new MCA PHA hardware!
john
The only stupid question is the one not asked!

John Donovan

So I tried PeakSight again with checking the smoothing option off and get a slightly better escape peak:



and now my acquisition looks pretty much the same:



So I guess the MCA method is not quite as good energy resolution...?
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Karsten Goemann

Our SX is in use right now so I can't collect examples for comparison before Monday.

But I seem to remember that the "traditional" PHA scans we collected with PFE on our system were looking extremely jagged and asymmetrical. Back then I thought it was because the electronics can't keep up with the fast baseline/window changes. Or the count rates are just too low, or both. The minimum PHA window setting I needed to get an acceptable PHA peak shape was 0.6V.

Gareth, to change the background colour in PeakSight, double left-click into the background of the graph. A window should open where you can adjust the colour.


Probeman

Quote from: Karsten Goemann on March 27, 2014, 04:03:57 PM
So in my quick testing, I can see we are getting back 256 intensities, but what voltage range do these intensities represent?  I ask because my PeakSight display shows them going from 0.5 to 5.5, but on Gareth's PeakSight display they go from 0 to 5 volts.

The test PHA return packet seems to *always* return 805 and 5637 millivolts for the baseline and window respectively,  but if I set the PHA window *before* the pha acquisition to 3 volts, I can see that the instrument window is 3 volts in the PeakSight display during the test PHA acquisition, but the test PHA function still returns the same exact range of intensities no matter what window setting I specify.
The only stupid question is the one not asked!

John Donovan

Here is what one sees when running a MCA PHA "test" in DebugMode (from the Output menu)...

Setting spectro  2 PHA (BWGBMD):  .5 3 1003 1250 -1 .0000038
Moving to crystal and on-peak spectrometer positions...
SX100MoveMotor: 2 to 27644 at speed 3000
Moving motor  2 from  32295 to  27644
Setting column conditions (TKCS)...
Inserting faraday cup or blanking beam
Faraday cup inserted or beam blanked
Setting operating conditions (TKCS):  40 15 50 0
Setting operating conditions (TKCS):  40 15 50 0
Setting operating conditions (TKCS):  40 15 50 0, getting current conditions...
SX100SetConditions: current conditions (TKCS) =  40 15 51.5915 0
SX100SetConditions: static variables (TKCS) =  40 15 50 0
SX100SetConditions: ForceAnalyticalConditions =  0
SX100SetConditions: OperatingVoltageTolerance =  .002, BeamCurrentTolerance =  .01
SX100SetConditions: beam current error = 51.5915 / 50 =  3.18300 %
Setting operating conditions (TKCS):  40 15 50 0, setting beam current...
SX100SetConditions: set beam current =  50
Setting operating conditions (TKCS):  40 15 50 0, setting beam size...
SX100SetConditions: set beam size =  0
Operating conditions set (TKCS)

Setting beam mode (Analog  Spot)...
Setting beam mode (Analog  Spot)...

Setting magnification:  40000
Setting magnification to  40000...
Magnification set to 40000

Removing faraday cup or unblanking beam
Faraday cup removed or beam unblanked
Acquiring PHA acquisition...
Setting spectro  2 PHA (BWGBMD):  .5 3 1003 1250 -1 .0000038

SX100GetPHADistributionMCA2: Motor= 2, Status= 1, MaxTries= 0
SX100GetPHADistributionMCA2: Motor= 2, Status= 0, MaxTries= 1
SX100GetPHADistributionMCA2: Motor= 2, Acquisition number=1
SX100GetPHADistributionMCA2: Baseline= 804
SX100GetPHADistributionMCA2: Window= 5627
SX100GetPHADistributionMCA2: Alignment= 0
SX100GetPHADistributionMCA2: Index= 256
SX100GetPHADistributionMCA2 (Byte): Min= 0, Max= 246, CountTime= 0.035
SX100GetPHADistributionMCA2 (PHA Sum): Min= 0, Max= 236.5714
SX100GetPHADistributionMCA2 (Averaged): Min= 0, Max= 236.5714
SX100GetPHADistributionMCA2 (Normalized): Min= 0, Max= 6759.184

Note that the returned "baseline" and "window" from the MCA PHA acquisition are 0.804 and 5.627 volts respectively.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"