Optimal autofocus settings
Started by Anette von der Handt, June 19, 2014, 12:10:37 PM
not sure if this forum section or the JEOL one is the better place.
I have been noticing that my JEOL autofocus (JEOL 8900; unknown autofocus type - I guess maximum intensity) and the PFE autofocus (maximum intensity) results do not match and that the PFE autofocus has a constant slight offset, enough to screw with the Si and Al intensities.
Playing around with the settings has not produced the results I am looking for. Therefore, I wondered if anyone could weigh in with their preferred PFE autofocus settings.
Things I have tried so far:
Meddled with the actual brightness settings on the JEOL (OM-AFD) - no effect
Added an automatic offset to the probewin.ini file - no effect, maybe I did it wrong??
Change autofocustype to gaussian - better but results in an error message of 'not enough data points' relatively often.
I would love to play around with it in much more detail and run the debugger but I would rather avoid this now (paying customers). The fact that the JEOL focus does a better job, lets me think that some values in the probewin.ini should be tweaked.
Current settings:
AutoFocusPresent=1
AutoFocusType=1
AutoFocusOffset=0.0
AutoFocusMaxDeviation=30
AutoFocusThresholdFraction=0.33
AutoFocusMinimumPtoB=1.4
AutoFocusRangeFineScan=100
AutoFocusRangeCoarseScan=600
AutoFocusPointsFineScan=200
AutoFocusPointsCoarseScan=100
AutoFocusTimeFineScan=20   ; 1-500 msec
AutoFocusTimeCoarseScan=20
Any input is welcome. Thank you!
Anette
Quote from: Anette von der Handt on June 19, 2014, 12:10:37 PM
Right you are!
It's this setting you want to tweak:
AutoFocusOffset=0.0
so enter the value in plus or minus microns.
Thanks for the quick reply! You know I tried that but in the rush missed the <micron bit, so only moved it an indiscernible 0.002 microns.
So just to be on the safe side, does it matter if I write:
AutoFocusOffset=+20
    or
AutoFocusOffset=20
Much obliged.
Quote from: Anette von der Handt on June 19, 2014, 01:41:38 PM
AutoFocusOffset=+20
    or
AutoFocusOffset=20
Much obliged.
Well, I never tried that but it should "resolve" OK. Try it!
