Display wt% with 4 or 5 digits after the decimal point in log window
Started by Donggao Zhao, April 03, 2015, 02:15:21 PM
Donggao Zhao
Currently our PFE log window displays wt% with three digits after the decimal point. For trace element measurement, we would like to see ppm level concentrations. How to display elemental or oxide wt% with 4 or 5 digits after the decimal point in the log window? Thanks.
Quote from: Donggao Zhao on April 03, 2015, 02:15:21 PM
Hi Donggao,
Good question!
From the Analytical | Analysis Options dialog, simply check the Use Automatic Format For Quantitative Results checkbox as seen here:
You'll have two choices, both are useful for trace element analyses.
The first is what you asked for, it gives the maximum number digits possible. The second, formats based on sensitivity and detection limit. So a normal output looks like this:
ELEM:        P       P       P       P       P      Zr      Si       O   SUM
XRAY:     (ka)    (ka)    (ka)    (ka)    (ka)    (la)      ()      ()
   179   -.012    .002    .000    .004    .025  50.088  15.323  34.913 100.342
   180    .003    .001    .001   -.019    .000  49.584  15.323  34.913  99.806
   181    .005    .004   -.001    .002   -.021  49.794  15.323  34.913 100.019
   182   -.006   -.001   -.008   -.001    .001  49.676  15.323  34.913  99.897
   183    .009   -.006    .008    .014   -.005  49.718  15.323  34.913  99.975
AVER:     .000    .000    .000    .000    .000  49.772  15.323  34.913 100.008
SDEV:     .009    .004    .006    .012    .016    .192    .000    .000    .203
SERR:     .004    .002    .003    .005    .007    .086    .000    .000
%RSD:     ----    ----    ----    ----    ----     .39     .00     .00
PUBL:     n.a.    n.a.    n.a.    n.a.    n.a.  49.764  15.323  34.913 100.000
%VAR:      ---     ---     ---     ---     ---   (.02)     .00     .00
DIFF:      ---     ---     ---     ---     ---   (.01)    .000    .000
STDS:      285     285     285     285     285     257       0       0
And the first option outputs like this:
   179 -.01235  .00239 -.00034  .00366  .02467 50.0875 15.3230 34.9130 100.342
   180  .00322  .00124  .00133 -.01930  .00027 49.5837 15.3230 34.9130 99.8065
   181  .00521  .00356 -.00112  .00231 -.02082 49.7941 15.3230 34.9130 100.019
   182 -.00556 -.00143 -.00805 -.00107  .00123 49.6762 15.3230 34.9130 99.8974
   183  .00946 -.00576  .00821  .01434 -.00540 49.7178 15.3230 34.9130 99.9746
AVER:   .00000  .00000  .00001 -.00001 -.00001 49.7719  15.323  34.913 100.008
SDEV:   .00881  .00371  .00581  .01223  .01638  .19210    .000    .000  .20333
SERR:   .00394  .00166  .00260  .00547  .00732  .08591  .00000  .00000
%RSD:     ----    ----    ----    ----    ----  .38596  .00000  .00000
PUBL:     n.a.    n.a.    n.a.    n.a.    n.a. 49.7640 15.3230 34.9130 100.000
DIFF:      ---     ---     ---     ---     ---   (.01)  .00000  .00000
The second option output looks like this:
   179    n.d.    n.d.    n.d.    n.d.     .02    50.1  15.323  34.913   100.3
   180    n.d.    n.d.    n.d.    n.d.    n.d.    49.6  15.323  34.913    99.8
   181    n.d.    n.d.    n.d.    n.d.    n.d.    49.8  15.323  34.913   100.0
   182    n.d.    n.d.    n.d.    n.d.    n.d.    49.7  15.323  34.913    99.9
   183    .009    n.d.    .008     .01    n.d.    49.7  15.323  34.913   100.0
AVER:     n.d.    n.d.    n.d.    n.d.    n.d.    49.8  15.323  34.913  100.01
%RSD:     ----    ----    ----    ----    ----    .386    .000    .000
Hope that helps!