spectrometer limit values

Started by wrigke, August 28, 2013, 03:41:11 PM

Previous topic - Next topic

wrigke

I noticed today that the lower and upper spectrometer limit values are not the same between the Cameca configuration file and the PFE motors.dat file.  I am not sure why they are not the same but it seems that that Cameca limits were set up at the factory and the PFE setting should be made to match the Cameca, not the other way around.  Is that reasonable?

Karen

John Donovan

Yes, you can edit the PFE spectrometer limit config values to match the PeakSight config values.  They are different because every time you re-start the instrument, it does a self test on each spectrometer and will therefore come up with slightly different values which are then saved to the PeakSight config files.

Another philosophy is to set the PFE spectrometer limits to values that will generally always be slightly inside the PeakSight spectrometer limits when they are re-determined during the next instrument re-start.
John J. Donovan, Pres. 