Quote from: Ben Buse on August 14, 2017, 09:07:34 AM
Quote from: John Donovan on August 14, 2017, 08:43:04 AM

What aspect of L and H type crystals(?) are you thinking of?
john

Hi John
So the H crystals have poorer resolution compared to the L crystals - due to the smaller rowland circle. Therefore they ideally would be given different energy resolutions

Ben

Hi Ben,
Yeah, that could be done, but I suspect there are many other aspects where the wavescan simulation could be also be improved.

For example, currently I synthesize the wavescan by combining pure element Penepma spectra for pure elements at 1 eV resolution. Then I convolve the summed scans using the Convolg.exe FORTRAN app that comes with Penepma for each crystal type.  At the time I tweaked the convolution resolution to get approximately WDS type peaks using the following equations:

C  ****  Example of FWHM(E) function for a WDS LIF spectrometer (~10 eV at 5898 eV)
      FWHM=0.0000003D0*E**2.

C  ****  Example of FWHM(E) function for a WDS PET spectrometer (~14 eV at 3690 eV)
      FWHM=0.0000008D0*E**2

C  ****  Example of FWHM(E) function for a WDS TAP spectrometer   (~4 eV at 1500 eV)
      FWHM=0.000000000005D0*E**3.9

C  ****  Example of FWHM(E) function for a WDS LDE spectrometer (~10 eV at 512 eV)
      FWHM=0.000000001D0*E**3.7

I'm not saying these are correct, but they give reasonable looking results.  The problem is that WDS resolution is extremely non-linear as a function of energy, and in fact has very high resolution at low energies and very low resolution at high energies, which is of course the opposite of EDS spectra!  See an example of a WDS spectra plotted in eV space:

https://smf.probesoftware.com/index.php?topic=837.msg5476#msg5476

Then there is the question of peak shape. The Convolg.exe FORTRAN app assumes only a Gaussian peak, but in fact the WDS peak is a Gaussian-Lorentenzian, with greatly extended tails due to polygonization effects from the crystal produced during manufacturing as described here:

https://smf.probesoftware.com/index.php?topic=837.msg5592#msg5592

So, there is a lot that can be improved.
john
The only stupid question is the one not asked!

Ben Buse

Quote from: John Donovan on August 09, 2017, 09:19:09 PM

Hi Ben,
Ok, there is no caveat now with the latest version of PFE.

The simulation mode will now always use the composition of the last standard for subsequent unknowns or wavescans, as long as the takeoff, keV and analyzed elements do not change.

If the conditions change and a standard has not been simulated, the program will simulate a random unknown or wavescan composition until a standard has been simulated using the new conditions.

I think this makes it much more intuitive now, try it out please and let me know what you think. And thanks for your help.
john

Hi John,

This works really well - you can even simulate analysis of samples for elements not present in the sample.

Ben

John Donovan

The previous WDS Penepma simulation code I implemented a few months ago utilizes the last standard composition for simulation of wavescans or unknowns.  But when running a bunch of wavescan simulations with different compositions, that method doesn't work very well, as it just utilizes the last standard composition that was run.  So I've modified the Penepma WDS simulation code, so that if the unknown or wavescan has the same *name* as one of the standards in the run, the program will now automatically load that standard composition for simulation.

To facilitate this, I use the Positions window (from the Automate!, then Digitize window), as seen here:



Using the buttons outlined in red, one merely selects the standard positions from the position database and then duplicates them to either unknown or wavescan position samples.  Now one can automate realistic simulations of unknowns and wavescans based on specific standard compositions as seen here:

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

We've made another small improvement in addition to the above simulation of unknowns and wavescans based on the standard composition if the sample name matches that of a standard already in the run.  This latest improvement automatically adds in oxygen by stoichiometry if the Calculate with Stoichiometric Oxygen option is selected in the Analyze! | Calculation Options dialog.

If this flag is in the default Calculate As Elemental option, and the unknown (or wavescan) sample is *not* using a sample name that is the same as a standard already in the run, the program will generate a random composition based on the currently analyzed elements as seen here:

Un    4 test as elemental, Results in Elemental Weight Percents

ELEM:       Si      Fe
BGDS:      LIN     LIN
TIME:    20.00   20.00
BEAM:    30.01   30.01

ELEM:       Si      Fe   SUM 
     8  23.709  76.564 100.273

AVER:   23.709  76.564 100.273
SDEV:     .000    .000    .000
SERR:     .000    .000
%RSD:      .00     .00
STDS:       14      26

However, with the latest code in version 12.1.3, if the Calculate with Stoichiometric Oxygen option is selected as seen here:



the software will now correctly calculate a random *oxide* composition based on the current cation stoichiometry as seen here:

Un   11 test oxide, Results in Elemental Weight Percents

ELEM:       Si      Fe       O
TYPE:     ANAL    ANAL    CALC
BGDS:      LIN     LIN
TIME:    20.00   20.00     ---
BEAM:    30.03   30.03     ---

ELEM:       Si      Fe       O   SUM 
    18  14.537  53.817  31.981 100.334

AVER:   14.537  53.817  31.981 100.334
SDEV:     .000    .000    .000    .000
SERR:     .000    .000    .000
%RSD:      .00     .00     .00
STDS:       14      26     ---

STKF:    .4101   .6548     ---
STCT:   144.18  151.28     ---

UNKF:    .1089   .4903     ---
UNCT:    38.27  113.29     ---
UNBG:      .09     .40     ---

ZCOR:   1.3353  1.0975     ---
KRAW:    .2655   .7489     ---
PKBG:   415.00  283.06     ---

And if the Display as Oxides checkbox is displayed, these results will also be displayed:

Un   11 test oxide, Results in Oxide Weight Percents

ELEM:     SiO2     FeO       O   SUM 
    18  31.099  69.235    .000 100.334

AVER:   31.099  69.235    .000 100.334
SDEV:     .000    .000    .000    .000
SERR:     .000    .000    .000
%RSD:      .00     .00     .00
STDS:       14      26     ---

Of course most of the time you'll probably want to enter an unknown (or wavescan) sample name that is the same as a standard already in the run (as described in the previous post), so that one will obtain a specific composition for teaching purposes.  But the random composition code is there just in case one acquires an unknown (or wavescan) simulation and either the sample name doesn't match an existing standard or there are no standards in the run.

Either way, you'll get a reasonable composition with simulated statistics applied, also suitable for teaching!   :)
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Probeman

As many of you know, Probe for EPMA has a very realistic demo or simulation mode which can be utilized for educational purposes. 

Probe for EPMA utilizes spectral intensities generated from the Penepma Monte Carlo software in two different ways. First, for WDS simulation it utilizes previously calculated pure element spectra from Penepma calculated using 1 eV resolution in 5 keV steps from 5 to 25 keV. For compounds it simply sums the spectra based on the weight fraction of the element (Yes, we should also be performing an absorption correction on each energy bin, but it's on the to-do list. In the meantime it's good enough...) to produce the final WDS spectrum.

Of course the spectrum is subsequently modified for higher Bragg orders, WDS exponential effects and absorption edges.  See the folder C:\UserData\Penepma12\Penepma\pure for these pure element intensity files. If you don't have these files, or you don't have a complete set of these files, you should update your Penepma data files using the Help | Update menu with the Update Penepma Monte Carlo Files Only checkbox checked.

For EDS, Probe for EPMA utilizes the Penepma in "real time", that is the EDS spectra is generated as the Penepma Monte Carlo program runs in the background and displayed in Probe for EPMA.  The photon rate depends on a number of factors but is roughly equivalent to having a beam current of a few nA or so.

In a recent post Donovan showed an example demonstrating some new flags for EDS quantification and data was shown where the Fe concentration was not very accurate compared to the other elements as seen here:

ELEM:       Si      Ca      Fe      Mg      Al      Mn       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC
BGDS:      LIN     LIN     EDS     EDS     EDS
TIME:    20.00   20.00   24.00   24.00   24.00     ---     ---
BEAM:    30.02   30.02   30.02   30.02   30.02     ---     ---

ELEM:       Si      Ca      Fe      Mg      Al      Mn       O   SUM 
     5  21.416  11.062   6.624  11.571   4.886    .077  43.597  99.234

AVER:   21.416  11.062   6.624  11.571   4.886    .077  43.597  99.234
SDEV:     .000    .000    .000    .000    .000    .000    .000    .000
SERR:     .000    .000    .000    .000    .000    .000    .000
%RSD:      .00     .00     .00     .00     .00     .00     .00

PUBL:   21.199  10.899   7.742  11.657   4.906    .077  43.597 100.077
%VAR:     1.02    1.50  -14.44  (-.73)  (-.41)     .00     .00
DIFF:     .217    .163  -1.118  (-.09)  (-.02)    .000    .000
STDS:      162     162     162     160     160     ---     ---

This is primarily because of the poor counting statistics resulting from a relatively high energy emission line at 15 keV, only running the Penepma simulation for 12 seconds (it assumed an EDS count time of 12 seconds because of the assumed 50% EDS deadtime), and only a single "measurement".  Here one can see the low count rate from the Fe Ka emission line from Penepma:

On-Peak (off-peak corrected) or EDS (bgd corrected) or MAN On-Peak X-ray Counts (cps/1nA) (and Faraday/Absorbed Currents):
ELEM:    si ka   ca ka   fe ka   mg ka   al ka   BEAM1   BEAM2
BGD:       OFF     OFF     EDS     EDS     EDS
SPEC:        1       2       0       0       0
CRYST:     PET    LPET     EDS     EDS     EDS
ORDER:       1       1       1       1       1
    5G   56.93   33.28    2.82   22.62    9.75  30.016  29.974

AVER:    56.93   33.28    2.82   22.62    9.75  30.016  29.974
SDEV:      .00     .00     .00     .00     .00    .000    .000
1SIG:      .31     .24     .06     .18     .12
SIGR:      .00     .00     .00     .00     .00
SERR:      .00     .00     .00     .00     .00
%RSD:      .00     .00     .00     .00     .00

However, if we increase the WDS counting time to 64 seconds we see a much better result as seen here:

ELEM:       Si      Ca      Fe      Mg      Al      Mn       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC
BGDS:      LIN     LIN     EDS     EDS     EDS
TIME:    24.00   24.00   64.00   64.00   64.00     ---     ---
BEAM:    30.03   30.03   30.03   30.03   30.03     ---     ---

ELEM:       Si      Ca      Fe      Mg      Al      Mn       O   SUM 
    37  21.533  10.589   7.691  11.640   4.903    .077  43.597 100.030

AVER:   21.533  10.589   7.691  11.640   4.903    .077  43.597 100.030
SDEV:     .000    .000    .000    .000    .000    .000    .000    .000
SERR:     .000    .000    .000    .000    .000    .000    .000
%RSD:      .00     .00     .00     .00     .00     .00     .00

PUBL:   21.199  10.899   7.742  11.657   4.906    .077  43.597 100.077
%VAR:     1.57   -2.85    -.66  (-.15)  (-.05)     .00     .00
DIFF:     .334   -.310   -.051  (-.02)   (.00)    .000    .000
STDS:      162     162     162     160     160     ---     ---

In fact, here we got lucky!   Spurious accuracy as they say!

However, instead of increasing the EDS count time, we could simply acquire more data points per standard.  Note that because PFE "re-seeds" the random number generator in Penepma for each spectrum "acquisition", the average will usually be more accurate than a single point.  Of course, even better is to acquire multiple data points and a longer counting time as seen here:

ELEM:       Si      Ca      Fe      Mg      Al      Mn       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    SPEC    SPEC
BGDS:      LIN     LIN     EDS     EDS     EDS
TIME:    24.00   24.00   64.00   64.00   64.00     ---     ---
BEAM:    30.01   30.01   30.01   30.01   30.01     ---     ---

ELEM:       Si      Ca      Fe      Mg      Al      Mn       O   SUM 
    37  21.540  10.588   7.696  11.396   5.202    .077  43.597 100.098
    38  21.430  10.739   7.906  11.129   4.779    .077  43.597  99.657
    39  21.481  10.701   6.680  11.716   5.343    .077  43.597  99.595
    40  21.296  10.845   7.757  11.773   4.451    .077  43.597  99.796
    41  21.675  10.843   7.908  11.969   4.561    .077  43.597 100.630

AVER:   21.485  10.743   7.589  11.597   4.867    .077  43.597  99.955
SDEV:     .140    .107    .517    .333    .392    .000    .000    .424
SERR:     .062    .048    .231    .149    .175    .000    .000
%RSD:      .65    1.00    6.81    2.87    8.05     .00     .00

PUBL:   21.199  10.899   7.742  11.657   4.906    .077  43.597 100.077
%VAR:     1.35   -1.43   -1.97  (-.52)  (-.79)     .00     .00
DIFF:     .286   -.156   -.153  (-.06)  (-.04)    .000    .000
STDS:      162     162     162     160     160     ---     ---

So the Fe is off about 2% which is typical EPMA accuracy. The value here for education is that the students can actually see the effects of increasing count time and/or replicate measurements on their own laptop computers.  With a bit of "luck" of course!   ;)
The only stupid question is the one not asked!

Probeman

But what about unknown acquisitions when PFE is in demo or simulation mode?

So if one acquires an unknown sample the software will create a random composition as described previously. However, if the unknown sample is the *same name* as a standard already added to the probe run, it will base the physics on that standard composition, as seen here for the NIST K-412 mineral glass:

ELEM:       Si      Ca      Fe      Mg      Al
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL
BGDS:      LIN     LIN     EDS     EDS     EDS
TIME:    60.00   60.00   64.00   64.00   64.00
BEAM:    30.00   30.00   30.00   30.00   30.00

ELEM:       Si      Ca      Fe      Mg      Al   SUM 
    21  21.376  10.788   8.220  10.257   5.251  55.893
    22  21.171  10.703   7.830   9.829   4.200  53.734
    23  21.220  10.716   9.168   9.776   4.973  55.853

AVER:   21.256  10.736   8.406   9.954   4.808  55.160
SDEV:     .107    .046    .688    .264    .544   1.235
SERR:     .062    .026    .397    .152    .314
%RSD:      .50     .43    8.18    2.65   11.32
STDS:      162     162     162     160     160

But now you may ask: why is the total only 55%?  Well since this is an unknown sample, the software does not know that there is oxygen present. Because of course because oxygen was not an analyzed element!

So if we then specify oxygen calculated by stoichiometry in the Calculation Options dialog we now get this result:

ELEM:       Si      Ca      Fe      Mg      Al       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    CALC
BGDS:      LIN     LIN     EDS     EDS     EDS
TIME:    60.00   60.00   64.00   64.00   64.00     ---
BEAM:    30.00   30.00   30.00   30.00   30.00     ---

ELEM:       Si      Ca      Fe      Mg      Al       O   SUM 
    21  21.216  10.896   8.503  11.568   5.392  43.370 100.946
    22  21.100  10.810   8.100  11.076   4.318  41.808  97.212
    23  21.073  10.834   9.488  10.939   5.103  42.794 100.232

AVER:   21.129  10.847   8.697  11.194   4.938  42.657  99.463
SDEV:     .076    .045    .714    .331    .556    .790   1.982
SERR:     .044    .026    .412    .191    .321    .456
%RSD:      .36     .41    8.21    2.96   11.26    1.85
STDS:      162     162     162     160     160     ---

Note that the iron statistics are still not wonderful, but the other elements are fine.
The only stupid question is the one not asked!

Probeman

Ok, so this is a little silly but it could help students learning to do EPMA in simulation mode from their laptops in class before they get on the actual instrument (as we do in our internship program at UofO). First it was noticed that the code wasn't displaying an escape peak when the emission line energy is high enough to cause one, so that is fixed:



Next the code was modified to respect the PHA gain value, as the user adjusts it, so here:



now gain adjusted down:



and now gain adjusted up (too high in fact!):



as suggested by a couple of students in Julie's class this quarter.
The only stupid question is the one not asked!

JonF

I've been playing around a bit recently with the simulation/teaching mode and I was wondering - as this is a simulation, do we have to run this in real time? Is there a way I could get PfE to simulate wavescans (for example) as fast as the computer will go (whilst still maintaining the counting statistics for example 6s/point)?

I tried having a look around the forums for whether this has been asked before, but I couldn't find an answer.

John Donovan

Quote from: JonF on February 04, 2019, 05:04:42 AM
I've been playing around a bit recently with the simulation/teaching mode and I was wondering - as this is a simulation, do we have to run this in real time? Is there a way I could get PfE to simulate wavescans (for example) as fast as the computer will go (whilst still maintaining the counting statistics for example 6s/point)?

I tried having a look around the forums for whether this has been asked before, but I couldn't find an answer.

Hi Jon,
It's an interesting idea but we cannot think of an easy way to do it.  But it's worth thinking about some more.

The simulation mode was actually meant to run as realistically as possible for several reasons, one being to teach students how the various options affect acquisition time. This means keeping the photon rate "real" if you know what I mean.  One thing you might try is a very high beam current. That might help with the WDS statistics in simulation mode, but the EDS simulation is already constrained by the Monte Carlo electron trajectory rate- which is somewhere around a few nA of beam current for most computers.

For a bit of history, the simulation mode in Probe for EPMA was originally designed for one purpose only and that was so we could develop the acquisition software without having a microprobe available. That is also why there is separate simulation code for the places where the JEOL and Cameca instruments behave differently!

But maybe Sandrin's "method development" application is what you are looking for?

https://smf.probesoftware.com/index.php?topic=743.0
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

JonF

Hi John,

  Sandrin's method development tool is good, but I've been using the simulation mode in Probe for EPMA to 'sense check' some of the crazier ideas and setups I've been having regarding unusual samples (i.e. things that arent available in that database). Also helps to avoid glaring errors in background positioning etc - all things to minimise setup time actually on the probe.

I'd been ramping up the (simulated) beam current and decreasing the count time, but after your suggestion I decided to take it to an extreme and fiddle with the aperture settings in the probewin.ini file to create a new aperture with a high beam current condition: 5kV, >1000nA, focused beam and a 0.1s dwell time over 1000 points results in full simulated WDS of a simulated materials in a couple of minutes - that'll do for me (I'll use the time to grab a coffee ;-) ).

To go with this, I've created a simulated standard database that I have been inputting ideal compositions of materials, digitize the position of the material in PfE and then convert the position to a wavescan and its the closest software that I know of (only?) to realistically simulate WDS. There's plenty of EDS simulation packages, so this fills a gap for me at least.

John Donovan

I thought I would share some screen snaps of WDS spectrum simulation in PFE for teaching and/or modeling new compositions and analytical situations.

As you may know PFE synthesizes WDS spectra from previously calculated pure element spectra from Penepma 2012 calculations (EDS spectra are simulated "on-the-fly" during the acquisition). These simulated WDS spectra are summed based on the elemental concentrations, with a crude background simulation, and a rough spectrometer efficiency and absorption edge effects thrown in. Not research grade maybe, but good enough for teaching and/or "what-if" situations.

Starting off with a scan of ZnO at the oxygen emission line position (simulated at 15 keV, 30 nA, 6 sec per point and 300 points):



Note that you can see a little of several Zn La 2nd order reflections. Here is another scan of the same region but on pure Zn:



Now one can see the 2nd order Bragg reflections better.  Here's another scan also at the oxygen emission position but on Al2O3:



Note the 3rd order Al reflections. And again on Al metal with about 1% oxygen:



Now let's switch to the nitrogen emission line position scanning first on TiN:



This is why quantifying nitrogen in the presence of titanium is so difficult...
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

I'm bumping this topic forward again, because with everything going on nowadays, the idea of running the Probe for EPMA software in "simulation" mode for teaching EPMA is maybe not such a bad idea. Here's what I recently mentioned to a colleague at a small university that doesn't have their own microprobe, but still wants her students to learn EPMA:

"We have found that using the Probe for EPMA software in "simulation" mode gives the students a quite realistic sense of running samples on the instrument. In simulation (or demo) mode one can set up elements, peak spectrometers, run PHAs, check backgrounds using wavescans, acquire standards, analyze them as secondary standards and also acquire standards "as unknowns" by using an unknown sample name that matches a standard currently in the run (otherwise it generates a random composition that totals close to 100%!). WDS and EDS spectra are realistic and contain typical artifacts such as higher order lines, etc.

It's what Julie does for her probe class in the classroom. Of course, once the students are comfortable with the physics theory and software practice, then they normally do get some hours on the instrument. But teaching in the classroom in simulation mode works quite well. After Julie's class, I've seen students come in the lab, and once samples are in the instrument, they just fire up the software and are basically comfortable on their own."

So if you already have the Probe for EPMA software, you might want to consider the idea of running PFE in this "simulation" mode, at least for part of your EPMA classes.  Basically every student simply downloads a copy of CalcZAF and Probe for EPMA on their laptop and they can choose JEOL or Cameca simulation mode and then you walk them through the various procedures in the (socially distanced) classroom (or remotely using Zoom or some other social/meeting app).

The simulation mode is not perfect (no satellite peak emissions from double ionized atoms!), but it does a number of interesting stuff, also including interference corrections, MAN, multi-point backgrounds, automation, etc., etc.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Probeman

Hi all,
Hope you're all coping with the present "situation".

I was recently chatting with my lab manager Julie Chouinard, and we were discussing the Fall Industrial Internship program:

https://internship.uoregon.edu/

at the UofO CAMCOR facility:

https://camcor.uoregon.edu/

and how student teaching and lab practicals will be handled with regards to the present social distancing rules.  We usually have 10-15 students per class, so with the present restrictions on the number of people in classrooms and labs (see attached below), this needs to be re-considered somewhat.

Now, in past years Julie has taken advantage of the simulation mode in Probe for EPMA, to teach the physical theory of EPMA and microanalysis as the students are guided through the software on their laptop computers in class. As the previous posts in this topic show, it is a quite realistic simulation for acquisition, automation and quantification of both WDS and EDS on JEOL and Cameca simulated instruments. Following these 6 to 8 weeks in the classroom, they would normally schedule a session on the actual instrument with myself or Julie for "hands on" learning in the lab.

Now with the social distancing guidelines in place, she has decided to again proceed with classroom instruction, again using the simulation mode of Probe for EPMA, but now with the students logging in using Zoom and running Probe for EPMA also on their own laptops. The only difference is that everyone will be remote except Julie.

For actual instrument time with the students, she will utilize our screen sharing capabilities already present on our EPMA/SEM instruments, which we have utilized in the past for checking on our instrument when we are away from the lab, by initially letting the student into the lab, then watching from her office using these screen sharing apps, including a PTZ camera for "looking over their shoulder" as needed.

If something unforeseen happens she'll be just down the hall in her office and can walk over to help, but at least for most of the time, it can probably all be handled remotely using screen sharing.  This seems like a very reasonable plan to me.   What's the plan in your lab for student teaching?

For those labs that already have purchased Probe for EPMA, remember that all your students can download a free copy for running in simulation mode, or for re-processing previously acquired data from your instrument.
The only stupid question is the one not asked!

John Donovan

We made some small changes to the simulation or "demo" mode of Probe for EPMA. This simulation mode can be utilized for teaching students as discussed in this topic above.

In fact our company has also been utilizing this simulation mode of PFE for remote training sessions when, for national or corporate security reasons, we cannot make a direct network connection to the instrument computer:

https://smf.probesoftware.com/index.php?topic=1297.0

However, one WDS artifact that we never simulated properly are the long tails of emission lines in WDS spectra, which are due to the "polygonization" effects produced during the manufacturing of Bragg crystals:

https://smf.probesoftware.com/index.php?topic=79.msg7818#msg7818

However it is worth the effort to simulate these polygonization tails for teaching/training purposes because these tails produce most of the spectral interferences that are observed in WDS analyses.  Here is an example of a *simulated* Si Ka emission line without these polygonization tails:



Note that we are missing the typical extended tails as seen here in an empirical measurement, again in the region of the Si Ka emission lines using a TAP crystal:



In addition to the main Ka peak, we also see the minor Si Kb peak. But if we zoom in a little on the background we can more easily see these extended tails around the peaks:



We also note some satellite emission lines due to doubly ionized Si atoms.

How to simulate these long tails? Well, one idea is to convolve the WDS spectra at a very low resolution (for example, EDS resolution of ~125 eV), then rescale the spectra to a few percent of the original spectra and sum them. If we do that we obtain a simulated spectrum as seen here, again for Si Ka:



and zooming in on the simulated spectrum we see this:



It's not perfect, but pretty good. Now all we are missing are the satellite emission lines as noted here:



Unfortunately the Penepma software which generates these spectra assumes only neutral atoms, so we are out of luck there. But at least now we can better simulate WDS spectral interferences!    :)

IMPORTANT: If you would like to be able to simulate these extended WDS tails in PFE demo mode, you should update *both* your Penepma files *and* your Probe for EPMA files, using the Help | Update Probe for EPMA menu in Probe for EPMA. For updating the Penepma files, check the Update Penema Files Only checkbox.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

John Donovan

The purpose of simulating WDS polygonization tail artifacts is simply to create a more realistic looking spectrum with WDS spectral overlaps from the long tails we observe in our instruments. So yesterday I decided to test how well these simulated tails could used to demonstrate the quantitative spectral interference corrections in Probe for EPMA simulations.

But first let me mention that the simulated WDS spectra generated in Probe for EPMA are synthesized by summing pre-calculated Penepma spectra for *pure elements* at various electron beam energies:

https://smf.probesoftware.com/index.php?topic=202.msg9589#msg9589

This is in contrast to the EDS simulated spectra generated in Probe for EPMA which are created by running Penepma in real time based on the actual compositions. My impression is that running Penepma in real time is roughly equivalent to an actual beam current of a few nano-amps. These simulated EDS spectra can then be quantified (when run long enough for sufficient statistics) for any elements in the spectrum.

Unfortunately we cannot do the same for WDS scans because the scan would vary in statistical significance during the simulated scan "acquisition".  So we resort to pre-calculated pure element simulations performed at 1 eV intervals for WDS resolution. In addition these WDS spectra for each analyzed element are re-normalized for the element in question to generate a quantitative intensity for that element, but other secondary emission lines in the spectra from other elements cannot be considered quantitative since currently, no absorption correction is applied to the spectra when they are summed.  In other words the pure element spectra are quantitative, but not necessarily the compound spectra.

Anywho, I decided to perform a simulation for Ti, V and Cr using pure element standards and a secondary standard of a Ti, Al, V, Cr alloy (SRM 654b) to see just how bad the interference correction is when performed on these simulated spectra with the newly added polygonization artifacts.  For reference the composition of SRM 654b is shown here:

St  654 NBS SRM-654b
TakeOff = 40.0  KiloVolt = 15.0  Density =  5.000  Type = alloy

NBS (NIST), Ti base alloy
Elemental Composition

Average Total Oxygen:         .000     Average Total Weight%:  100.000
Average Calculated Oxygen:    .000     Average Atomic Number:   21.491
Average Excess Oxygen:        .000     Average Atomic Weight:   45.783

ELEM:       Ti      Al       V      Fe      Sn      Cu      Ni      Cr      Si      Mo      Zr
XRAY:      ka      ka      ka      ka      la      ka      ka      ka      ka      la      la
ELWT:   88.974   6.340   4.310    .230    .023    .004    .028    .025    .045    .013    .008
KFAC:    .8825   .0449   .0420   .0021   .0002   .0000   .0003   .0002   .0004   .0001   .0001
ZCOR:   1.0082  1.4110  1.0270  1.0837   .9563  1.0886  1.0434  1.1243  1.2290  1.0746  1.1283
AT% :   85.041  10.758   3.873    .189    .009    .003    .022    .022    .073    .006    .004

First here is the simulated WDS spectra for V Ka (interfered by Ti Kb):



Note that the 4 wt% V emission is quite interfered by the 89 wt%Ti (Kb) emission. The Cr Ka emission is also somewhat interfered by the vanadium Kb emission as seen here:



An example of what I call a "cascade" interference, where A interferes with B, and B interferes with C, which requires an iterative solution for quantitative results (the situation where two primary analytical emission lines *both* interfere with each other (e.g., Ba La and Ti Ka), I call "self-interfering" or "pathological" interferences, which also requires an iterative solution).

Using instrumental measurements on this same alloy produces excellent results as shown many years ago in this paper (see last line in table 1 on page 26):

https://epmalab.uoregon.edu/publ/Improved%20Interference%20(Micro.%20Anal,%201993).pdf

OK, so first is the simulated "analysis" of the alloy *without* an interference correction:

St  654 Set   1 NBS SRM-654b
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 30.0  Beam Size =    0
(Magnification (analytical) =  20000),        Beam Mode = Analog  Spot
(Magnification (default) =      600, Magnification (imaging) =    200)
Image Shift (X,Y):                                         .00,    .00

NBS (NIST), Ti base alloy
Number of Data Lines:   5             Number of 'Good' Data Lines:   5
First/Last Date-Time: 08/11/2021 09:32:18 AM to 08/11/2021 09:35:22 AM

Average Total Oxygen:         .000     Average Total Weight%:  100.360
Average Calculated Oxygen:    .000     Average Atomic Number:   21.492
Average Excess Oxygen:        .000     Average Atomic Weight:   45.810
Average ZAF Iteration:        2.00     Average Quant Iterate:     2.00

St  654 Set   1 NBS SRM-654b, Results in Elemental Weight Percents

ELEM:       Ti       V      Cr      Al      Fe
TYPE:     ANAL    ANAL    ANAL    SPEC    SPEC
BGDS:      LIN     LIN     LIN
TIME:    20.00   20.00   20.00     ---     ---
BEAM:    30.01   30.01   30.01     ---     ---

ELEM:       Ti       V      Cr      Al      Fe   SUM 
   116  89.094   4.914    .089   6.340    .230 100.666
   117  88.600   5.037    .108   6.340    .230 100.315
   118  88.645   4.936    .100   6.340    .230 100.251
   119  88.728   5.012    .076   6.340    .230 100.386
   120  88.608   4.952    .052   6.340    .230 100.182

AVER:   88.735   4.970    .085   6.340    .230 100.360
SDEV:     .207    .052    .022    .000    .000    .187
SERR:     .093    .023    .010    .000    .000
%RSD:      .23    1.05   25.80     .00     .00

PUBL:   88.974   4.310    .025   6.340    .230  99.879
%VAR:     -.27   15.32  239.36     .00     .00
DIFF:    -.239    .660    .060    .000    .000
STDS:      522     523     524     ---     ---

STKF:    .9940  1.0000   .9988     ---     ---
STCT:   344.27  340.70  339.74     ---     ---

UNKF:    .8805   .0484   .0008     ---     ---
UNCT:   304.95   16.49     .26     ---     ---
UNBG:      .49     .53     .94     ---     ---

ZCOR:   1.0078  1.0268  1.1233     ---     ---
KRAW:    .8858   .0484   .0008     ---     ---
PKBG:   627.81   31.97    1.28     ---     ---


Note that the V concentration is high by about 0.5 wt% and the Cr is high by about 160 PPM.  Now the same analysis but *with* an interference correction applied:

St  654 Set   1 NBS SRM-654b
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 30.0  Beam Size =    0
(Magnification (analytical) =  20000),        Beam Mode = Analog  Spot
(Magnification (default) =      600, Magnification (imaging) =    200)
Image Shift (X,Y):                                         .00,    .00

NBS (NIST), Ti base alloy
Number of Data Lines:   5             Number of 'Good' Data Lines:   5
First/Last Date-Time: 08/11/2021 09:32:18 AM to 08/11/2021 09:35:22 AM

Average Total Oxygen:         .000     Average Total Weight%:   99.749
Average Calculated Oxygen:    .000     Average Atomic Number:   21.482
Average Excess Oxygen:        .000     Average Atomic Weight:   45.781
Average ZAF Iteration:        2.00     Average Quant Iterate:     4.00

St  654 Set   1 NBS SRM-654b, Results in Elemental Weight Percents

ELEM:       Ti       V      Cr      Al      Fe
TYPE:     ANAL    ANAL    ANAL    SPEC    SPEC
BGDS:      LIN     LIN     LIN
TIME:    20.00   20.00   20.00     ---     ---
BEAM:    30.01   30.01   30.01     ---     ---

ELEM:       Ti       V      Cr      Al      Fe   SUM 
   116  89.113   4.287    .084   6.340    .230 100.053
   117  88.619   4.413    .103   6.340    .230  99.705
   118  88.664   4.312    .095   6.340    .230  99.641
   119  88.747   4.387    .072   6.340    .230  99.775
   120  88.627   4.328    .047   6.340    .230  99.573

AVER:   88.754   4.345    .080   6.340    .230  99.749
SDEV:     .207    .053    .022    .000    .000    .186
SERR:     .093    .024    .010    .000    .000
%RSD:      .23    1.22   27.23     .00     .00

PUBL:   88.974   4.310    .025   6.340    .230  99.879
%VAR:     -.25     .82  221.58     .00     .00
DIFF:    -.220    .035    .055    .000    .000
STDS:      522     523     524     ---     ---

STKF:    .9940  1.0000   .9988     ---     ---
STCT:   344.27  340.70  339.74     ---     ---

UNKF:    .8805   .0423   .0007     ---     ---
UNCT:   304.95   14.42     .24     ---     ---
UNBG:      .49     .53     .94     ---     ---

ZCOR:   1.0080  1.0270  1.1242     ---     ---
KRAW:    .8858   .0423   .0007     ---     ---
PKBG:   627.81   28.07    1.26     ---     ---
INT%:     ----  -12.59   -5.68     ---     ---

Now with the interference applied the V concentration is quite good, but the Cr concentration is under corrected and still a bit high. I'm going to say that this first attempt at adding WDS polygonization artifacts in these simulated WDS spectra is pretty good, but not quite quantitative enough at the 1000 PPM level and below. I will continue to evaluate and let you all know what else I find.

In summary I would say these polygonization simulations are good enough for teaching and training, which of course is the whole purpose of the simulation mode in Probe for EPMA. Please feel free to try some WDS simulations of your own, and share your results here in this topic.
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"